Checking pH meter calibration HCL >37%

  • Thread starter rwooduk
  • Start date
711
45
Hi,

I've calibrated our pH meter using 4.01 and 7.01 buffer solution, but I'm not sure if the probe has dried out at some point. So I want to check the calibration, I have some HCl > 37% that I can use. Just a quick question, since pH is on a log scale so I have to make solutions every ten unit?

i.e. 1ml HCl + 100mL DI water
1ml HCL + 1000mL DI water
1ml HCL + 10000mL DI water

For large volumes of water would my pH meter even detect a change? Please could someone suggest a realistic range of diluted HCl values with which to check the calibration is correct or is the above correct?

Thanks for any advice
 

Borek

Mentor
28,013
2,518
You can't properly check the calibration with diluted acid, you need buffers for that. Diluted acid pH is very approximate (not to mention the fact it won't work at all at pH 7.00).
 
711
45
Ok thanks again Borek! I think we have more buffer solutions other than those used to calibrate, I will attempt to make use of them
 

Borek

Mentor
28,013
2,518
You can't calibrate the pH electrode too often :wink:

Seriously, I suggest checking it once a day. It may turn out the readings are quite stable, but its in no way guaranteed.
 

Want to reply to this thread?

"Checking pH meter calibration HCL >37%" You must log in or register to reply here.

Related Threads for: Checking pH meter calibration HCL >37%

  • Posted
Replies
2
Views
2K
  • Posted
Replies
11
Views
980
  • Posted
Replies
8
Views
2K
  • Posted
Replies
9
Views
10K
Replies
3
Views
11K

Physics Forums Values

We Value Quality
• Topics based on mainstream science
• Proper English grammar and spelling
We Value Civility
• Positive and compassionate attitudes
• Patience while debating
We Value Productivity
• Disciplined to remain on-topic
• Recognition of own weaknesses
• Solo and co-op problem solving

Hot Threads

Top