Checking pH meter calibration HCL >37%

Click For Summary

Discussion Overview

The discussion revolves around the calibration of a pH meter, specifically in relation to using hydrochloric acid (HCl) solutions greater than 37% for this purpose. Participants explore the appropriateness of using diluted HCl for calibration checks and the frequency of calibration.

Discussion Character

  • Technical explanation, Debate/contested, Experimental/applied

Main Points Raised

  • One participant suggests using diluted HCl solutions to check the calibration of a pH meter and questions the necessary dilution ratios for accurate readings.
  • Another participant argues that diluted acid cannot properly check calibration and emphasizes the need for buffer solutions instead, noting that diluted acid pH is approximate and ineffective at pH 7.00.
  • A later reply indicates that the original poster may have access to additional buffer solutions and plans to utilize them for calibration checks.
  • One participant humorously notes that pH electrodes should not be calibrated too frequently but suggests checking them once a day, while acknowledging that stability in readings is not guaranteed.

Areas of Agreement / Disagreement

Participants express differing views on the use of diluted HCl for calibration, with some advocating for buffer solutions instead. The discussion remains unresolved regarding the best method for checking calibration.

Contextual Notes

There are limitations regarding the effectiveness of diluted acid for calibration, as well as the potential variability in pH meter readings over time.

rwooduk
Messages
757
Reaction score
59
Hi,

I've calibrated our pH meter using 4.01 and 7.01 buffer solution, but I'm not sure if the probe has dried out at some point. So I want to check the calibration, I have some HCl > 37% that I can use. Just a quick question, since pH is on a log scale so I have to make solutions every ten unit?

i.e. 1ml HCl + 100mL DI water
1ml HCL + 1000mL DI water
1ml HCL + 10000mL DI water

For large volumes of water would my pH meter even detect a change? Please could someone suggest a realistic range of diluted HCl values with which to check the calibration is correct or is the above correct?

Thanks for any advice
 
Chemistry news on Phys.org
You can't properly check the calibration with diluted acid, you need buffers for that. Diluted acid pH is very approximate (not to mention the fact it won't work at all at pH 7.00).
 
  • Like
Likes   Reactions: rwooduk
Ok thanks again Borek! I think we have more buffer solutions other than those used to calibrate, I will attempt to make use of them
 
You can't calibrate the pH electrode too often :wink:

Seriously, I suggest checking it once a day. It may turn out the readings are quite stable, but its in no way guaranteed.
 
  • Like
Likes   Reactions: epenguin and rwooduk

Similar threads

Replies
5
Views
6K
  • · Replies 21 ·
Replies
21
Views
5K
Replies
1
Views
3K
  • · Replies 10 ·
Replies
10
Views
11K
  • · Replies 3 ·
Replies
3
Views
14K
  • · Replies 5 ·
Replies
5
Views
2K
  • · Replies 9 ·
Replies
9
Views
3K
Replies
11
Views
4K
  • · Replies 1 ·
Replies
1
Views
11K
  • · Replies 1 ·
Replies
1
Views
3K