Decreasing the power density of a laser

Click For Summary
SUMMARY

The discussion focuses on reducing the power density of a 488 nm continuous wave laser, set at 1 watt, which is causing burn deposits on quartz sample holders. Participants suggest various methods to decrease power density, including using beam splitters, additional pinholes, and defocusing the lens. They also recommend using neutral density filters and microscope slides to attenuate the laser intensity effectively. Safety precautions are emphasized due to the high power of the laser and potential hazards from stray reflections.

PREREQUISITES
  • Understanding of laser optics and beam manipulation
  • Familiarity with 488 nm continuous wave laser technology
  • Knowledge of optical components such as pinholes and beam splitters
  • Awareness of laboratory safety protocols when working with lasers
NEXT STEPS
  • Research the use of beam splitters for laser power reduction
  • Learn about neutral density filters and their application in laser setups
  • Investigate the effects of defocusing lenses on laser power density
  • Explore the use of microscope slides as inexpensive optical attenuators
USEFUL FOR

Laser technicians, optical engineers, and researchers working with high-power lasers who need to manage power density effectively in their experiments.

Ella Reid
Messages
10
Reaction score
0
Hello! I have an issue with the power density of a 488 nm continuous wave laser. I'm using this laser to excite my sample. The problem is that even at low power, the laser is causing the samples to form a burn deposit on the quartz sample holder. Between the laser source and the sample, there are a series of dialectric mirrors, pinholes, and a focussing lens. What are the ways in which power density can be decreased at the focal point?
 
Science news on Phys.org
There are various cheap filters that can reduce the laser intensity. Adding another pinhole at the right place could work as well, simply by blocking a part of the laser light.

You'll have to figure out if that disturbs the measurement.
 
  • Like
Likes   Reactions: Ella Reid
How about defocusing the lens a little?
 
  • Like
Likes   Reactions: Ella Reid
mfb said:
There are various cheap filters that can reduce the laser intensity. Adding another pinhole at the right place could work as well, simply by blocking a part of the laser light.

You'll have to figure out if that disturbs the measurement.
Thanks for the reply, I have tried that but my sample still decomposes.
 
pixel said:
How about defocusing the lens a little?
Thanks for the reply, the focussing lenses have knobs to adjust the focus in the x,y, and z direction. Do you think it matters which direction is defocus? Also, does it matter where in the focussing lens the laser beam hits? Thanks a lot.
 
Ella Reid said:
Hello! I have an issue with the power density of a 488 nm continuous wave laser. I'm using this laser to excite my sample. The problem is that even at low power, the laser is causing the samples to form a burn deposit on the quartz sample holder. Between the laser source and the sample, there are a series of dialectric mirrors, pinholes, and a focussing lens. What are the ways in which power density can be decreased at the focal point?

good grief ! what is the power of the laser ??
 
  • Like
Likes   Reactions: Ella Reid
Ella Reid said:
Thanks for the reply, I have tried that but my sample still decomposes.
A smaller pinhole, at a point where the laser beam is wider? A pinhole off-center?
There must be some way to make the beam weaker if you block a part of it.

I don't know the requirements on the beam parameters, they might change with an additional pinhole.
 
  • Like
Likes   Reactions: Ella Reid and davenn
  • #10
davenn said:
good grief ! what is the power of the laser ??
Thanks for the question, the laser is set to 1 watt.
 
  • #11
mfb said:
A smaller pinhole, at a point where the laser beam is wider? A pinhole off-center?
There must be some way to make the beam weaker if you block a part of it.

I don't know the requirements on the beam parameters, they might change with an additional pinhole.
Thanks for the reply.
 
  • #12
Ella Reid said:
Thanks for the reply, the focussing lenses have knobs to adjust the focus in the x,y, and z direction. Do you think it matters which direction is defocus? Also, does it matter where in the focussing lens the laser beam hits? Thanks a lot.

I don't know how your x, y and z directions are defined, but the relevant one is the optical axis along the laser beam. Let's call that the z-direction. I'm not sure what you mean by changing the focus in the x- and y-direction. Anyhow. the focusing lens is presumably bringing the laser beam to tight beam waist, thus concentrating its power in a small volume. If you move the focusing lens either toward or away from the laser, the beam waist at the sample will expand, thus lessening the power density.

The laser beam should go through the center of the focusing lens, otherwise it will be displaced in the x- or y-direction.
 
  • Like
Likes   Reactions: Ella Reid
  • #13
pixel said:
I don't know how your x, y and z directions are defined, but the relevant one is the optical axis along the laser beam. Let's call that the z-direction. I'm not sure what you mean by changing the focus in the x- and y-direction. Anyhow. the focusing lens is presumably bringing the laser beam to tight beam waist, thus concentrating its power in a small volume. If you move the focusing lens either toward or away from the laser, the beam waist at the sample will expand, thus lessening the power density.

The laser beam should go through the center of the focusing lens, otherwise it will be displaced in the x- or y-direction.

Thanks for the reply, our focussing lens are mounted in an assembly like this https://www.newport.com/p/77259. That is what I meant.
 
  • #14
This might be naive on my part, but what about a beam splitter. A simple splitter between the laser and the beginning of the optics would reduce the power by 50% and might not require any adjustment to the optics. And I think there are some splitters that allow for adjusting the split ratio.
 
Last edited:
  • Like
Likes   Reactions: Nik_2213 and Ella Reid
  • #15
websterling said:
This might be naive on my part, but what about a beam splitter. A simple splitter between the laser and the beginning of the optics would reduce the power by 50% and might not require any adjustment to the optics. And I think there are some splitters that allow for adjusting the split ratio.
Thanks for the reply, I will look into this.
 
  • #16
If interference effects are not deadly to you just add a bunch of microscope slides (with air spaces) somewhere in collimated laser path at some angle (approx. 45 degree). Each reflection will attenuate about 10 % of light if the incidence is s-polarized light (simply just rotate the slide and check for biggest reflection). Avoid direct back reflection as it will make laser unstable! Its cheap and easy.
 
  • Like
Likes   Reactions: Ella Reid
  • #17
Is the laser too powerful to use neutral density filters? Or pop out some lenses from cheap sunglasses?
 
  • Like
Likes   Reactions: Ella Reid
  • #18
Domullus said:
If interference effects are not deadly to you just add a bunch of microscope slides (with air spaces) somewhere in collimated laser path at some angle (approx. 45 degree). Each reflection will attenuate about 10 % of light if the incidence is s-polarized light (simply just rotate the slide and check for biggest reflection). Avoid direct back reflection as it will make laser unstable! Its cheap and easy.
Thanks for the reply, I'll look into this.
 
  • #19
Khashishi said:
Is the laser too powerful to use neutral density filters? Or pop out some lenses from cheap sunglasses?
Thanks for the advice.
 
  • #20
1 W is a very powerful laser. Please be careful of stray reflections (off the pinhole diaphram or ND filter, for instance), and use proper lab safety procedures (no public access to room when laser is running, warning signs, laser-filtering safety goggles for you and other lab personnel, etc.). Also be careful of attenuating filters. If they are placed in a position where the beam is focused, they can crack due to heating from the absorption. It is common to defocus the beam, attenuate it, and then refocus.
 
  • Like
Likes   Reactions: Nik_2213

Similar threads

  • · Replies 10 ·
Replies
10
Views
3K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 6 ·
Replies
6
Views
3K
  • · Replies 5 ·
Replies
5
Views
3K
  • · Replies 8 ·
Replies
8
Views
2K
  • · Replies 33 ·
2
Replies
33
Views
3K
  • · Replies 5 ·
Replies
5
Views
3K
  • · Replies 3 ·
Replies
3
Views
4K
  • · Replies 8 ·
Replies
8
Views
935
  • · Replies 8 ·
Replies
8
Views
6K