Does the Mass of an Accelerometer Affect Vibration Frequency?

Click For Summary

Discussion Overview

The discussion revolves around the variability in acceleration measurements obtained from different accelerometers when measuring vibrations from a shaker. Participants explore the reliability of different instruments, calibration methods, and the potential influence of accelerometer mass on vibration frequency.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested
  • Mathematical reasoning

Main Points Raised

  • Some participants note that it is normal for different accelerometers to yield different acceleration measurements, particularly if they are not calibrated properly.
  • Others suggest that calibration should be done using known values from other reliable methods rather than using accelerometers to calibrate each other.
  • A participant proposes using a turntable to compare readings of accelerometers against calculated centripetal acceleration, emphasizing the importance of knowing the sensor's exact position.
  • There is a discussion about the challenges of measuring small accelerations and the difficulty in finding manufacturers' specifications for calibration.
  • One participant mentions a significant difference in measurements obtained from a laser telemeter and an accelerometer, highlighting a 27% relative difference between two measurement methods.
  • Another participant raises the question of whether the mass of the accelerometer affects the frequency of the vibrations being measured.
  • Concerns are expressed about the need for clarity regarding the experimental setup, including the type of shaker used and the specific measurements being taken.

Areas of Agreement / Disagreement

Participants generally agree that discrepancies in measurements are expected, but there is no consensus on the best calibration methods or the impact of accelerometer mass on vibration frequency. The discussion remains unresolved with multiple competing views on these topics.

Contextual Notes

Participants mention various uncertainties, including the accuracy required for measurements, the specific conditions under which measurements are taken, and the lack of detailed information about the experimental setup. There are also unresolved questions about the definitions of terms like "ecart type" and the context of frequency measurements.

Who May Find This Useful

This discussion may be useful for researchers and practitioners involved in vibration analysis, sensor calibration, and those interested in the technical nuances of accelerometer measurements in experimental setups.

Thien
Messages
9
Reaction score
0
Hi everyone,

Is it normal if the acceleration measured is different with different instrumentation?
I have measured the acceleration of shaker with three different accelerometer and I didn't obtain the same result.

Thanks for your help!
 
Physics news on Phys.org
It is normal for different instruments to disagree - the same instrument, if it is sensitive, will disagree with itself when repeating the same measurement - if they are properly calibrated, though, they should agree to within their uncertainty limits.
If all measurements agree, it is usually suspicious.
 
Simon Bridge said:
It is normal for different instruments to disagree - the same instrument, if it is sensitive, will disagree with itself when repeating the same measurement - if they are properly calibrated, though, they should agree to within their uncertainty limits.
If all measurements agree, it is usually suspicious.

Thank you for your response!

So, how can we know which intrument is reliable to calibrate the others?

Sincerely
 
You don't use an accelerometer to calibrate other accelerometers. You should use the manufacturers specifications for calibration.
In general, you calibrate instruments using known values that you obtain by other means ... i.e. use known accelerations (measured using a stopwatch and an accurate ruler say). It helps to decide how accurate you need the measurements to be.

If you knew that one accelerometer was the most accurate - wouldn't you'd just use that one and not bother with the others?
I'm guessing you are using the accelerometers in a device like a phone - and you have different kinds of phone... is that correct?

Note: the most reliable accelerometer is the one that returns the same value most consistently each time for the same test... reliability is not always helpful: an accelerometer that has two readouts "accelerating" and "not accelerating" will probably be very reliable... ("reliable" can also mean "robust").
 
  • Like
Likes   Reactions: Thien
Simon Bridge said:
You don't use an accelerometer to calibrate other accelerometers. You should use the manufacturers specifications for calibration.
In general, you calibrate instruments using known values that you obtain by other means ... i.e. use known accelerations (measured using a stopwatch and an accurate ruler say). It helps to decide how accurate you need the measurements to be.

If you knew that one accelerometer was the most accurate - wouldn't you'd just use that one and not bother with the others?
I'm guessing you are using the accelerometers in a device like a phone - and you have different kinds of phone... is that correct?

Note: the most reliable accelerometer is the one that returns the same value most consistently each time for the same test... reliability is not always helpful: an accelerometer that has two readouts "accelerating" and "not accelerating" will probably be very reliable... ("reliable" can also mean "robust").

I use an MEMS accelerometer of Analog Device, a laser telemeter and an PCB piezo accelerometer to measure at harmonic regime.

Moreover, the acceleration level is so small about hundreds "mg". It's difficult to calibrate accurately. And I can't find the manufacturers specifications for calibration.
 
Thien said:
So, how can we know which intrument is reliable to calibrate the others?
I would use a turn table to spin them at some precisely known radius and angular velocity, then compare their reading with the calculated centripetal acceleration. But note that you have to know the exact position of the actual sensor for this.
 
  • Like
Likes   Reactions: Thien
A.T. said:
I would use a turn table to spin them at some precisely known radius and angular velocity, then compare their reading with the calculated centripetal acceleration. But note that you have to know the exact position of the actual sensor for this.
Thank you for your response!

Do you mean that the value measured is depend on the position of each accelerometer?
I'm not clear about radius and angular velocity which you mentioned, can they be measured?
 
Thien said:
I'm not clear about radius and angular velocity which you mentioned, can they be measured?
I think that measuring the radius and angular velocity in a rotating setup is more reliable than accelerating the sensors along a linear track, to calibrate them.
 
  • Like
Likes   Reactions: Thien
How did you measure acceleration with a laser telemeter? Recorded position versus time and integrated to get acceleration?
 
  • Like
Likes   Reactions: Thien
  • #10
nasu said:
How did you measure acceleration with a laser telemeter? Recorded position versus time and integrated to get acceleration?
You would have to differentiate it twice, which for noisy data usually means fitting some smooth function.
 
  • #11
Sure, you are right.
We had some discussion with measuring acceleration and wanting to get position by integration and remained in the "integration mode". :)
 
  • #12
There is still the question of how accurate the measurements need to be ... for that matter, how different are they?
Are the measurements consistent within the manufacturers uncertainty limits?
Perhaps if we had an example of the data with a description of the measurement process?

However - the controlled situations suggested above should help calibrate the equipment.
 
  • Like
Likes   Reactions: Thien
  • #13
nasu said:
How did you measure acceleration with a laser telemeter? Recorded position versus time and integrated to get acceleration?
Hi nasu!
I carried out two measurements with the same laser telemeter: 1/ displacement measurement then integrated to get acceleration. 2/ acceleration measurement directly. There is a difference between two measurements and this ecart type relative is 27%.
 
  • #14
A.T. said:
I think that measuring the radius and angular velocity in a rotating setup is more reliable than accelerating the sensors along a linear track, to calibrate them.
I agree with you but I don't have equipment to do that..
Note: the acceleration in Z-axis is useful for my test.
 
  • #15
Simon Bridge said:
There is still the question of how accurate the measurements need to be ... for that matter, how different are they?
Are the measurements consistent within the manufacturers uncertainty limits?
Perhaps if we had an example of the data with a description of the measurement process?

However - the controlled situations suggested above should help calibrate the equipment.
Hi Simon!
The shaker is excited with a function generator and then the vibrating part is measured with three accelerometer. The ecart type depends on frequency and tension excitation. For example, the Analog device and the laser telemeter has 19% and 200% ecart type at 30Hz and 100Hz, respectively.
 
  • #16
Thien said:
Hi Simon!
The shaker is excited with a function generator and then the vibrating part is measured with three accelerometer. The ecart type depends on frequency and tension excitation. For example, the Analog device and the laser telemeter has 19% and 200% ecart type at 30Hz and 100Hz, respectively.
You just left out a great deal of information - assume I have not looked at your setup and do not know what you are doing or what equipment you have.
It sounds like you have a machine for shaking something - what? how? What sort of shaking do you expect?
Are you trying to measure the acceleration in three different axes directions at the same time or just at random orientations about the thing being shaken or what?
19% and 200% of what? 30Hz and 100Hz what?
You still have not said how accurate you need them to be...

I think we've gone as far as we can without knowing what you are trying to do.
 
  • Like
Likes   Reactions: Thien
  • #17
Is the mass of the accelerometer changing the frequency of the vibration?
 
  • Like
Likes   Reactions: Thien
  • #18
Simon Bridge said:
You just left out a great deal of information - assume I have not looked at your setup and do not know what you are doing or what equipment you have.
It sounds like you have a machine for shaking something - what? how? What sort of shaking do you expect?
Are you trying to measure the acceleration in three different axes directions at the same time or just at random orientations about the thing being shaken or what?
19% and 200% of what? 30Hz and 100Hz what?
You still have not said how accurate you need them to be...

I think we've gone as far as we can without knowing what you are trying to do.
The acceleration is recorded in z-axis at the same time. To be more clearly, I want to measure precisely acceleration in z-axis at low frequency with a shaker. The accuracy could be less 10%.
 
  • #19
CWatters said:
Is the mass of the accelerometer changing the frequency of the vibration?
No, it doesn't change the frequency but it changes altitude of the acceleration.
But the measurement condition is unchanged.
 

Similar threads

Replies
4
Views
2K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 10 ·
Replies
10
Views
3K
  • · Replies 1 ·
Replies
1
Views
13K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 12 ·
Replies
12
Views
4K
  • · Replies 1 ·
Replies
1
Views
1K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 1 ·
Replies
1
Views
2K