Finding Propagation of Uncertainty?
The discussion centers on calculating the propagation of uncertainty for the density of a cylinder based on measurements of mass, length, and diameter. The user initially calculates the uncertainty but receives an incorrect grade from the online system. It is clarified that the user needs to multiply the square root of the summed squared uncertainties by the calculated density value. The exponents in the uncertainty formula correspond to the variables' roles in the density calculation. Ultimately, applying this correction resolves the grading issue with the online system.
Monkey618
You measure the mass of the cylinder to be m = 584.9 +- 0.5 grams, and you measure the length of the cylinder to be L = 18.195 +- 0.003 cm. Just like in the lab you performed, you now measure the diameter in eight different places and obtain the following results.
Diameter (cm)
2.125
2.090
2.065
2.240
2.110
2.100
2.080
2.240
This gives an average of 2.131 +- 0.0695
This makes the density = 9.01 g/cm^3 +- propagation of uncertainty
Trying to calculate this, I have: sqrt( ((1*0.5) / 584.9)^2 + ((-2 * 0,0695) / 2.131)^2 + ((-1 * 0.003) / 18.195)^2 ) = 0.0652
However the online grading system say's that I'm wrong. So where have I gone wrong with the uncertainty of the density. All of the other values have been graded and marked correct, so where did I mess up with the uncertainty.
I'm not really clear where the "1", "-2", or "-1" came from in the formula above, I was basing it on my notes from class.
Diameter (cm)
2.125
2.090
2.065
2.240
2.110
2.100
2.080
2.240
It looks like you forgot to multiply the "sqrt" by the calculated value of the function being evaluated (the density value).
They are values that depend upon the exponent of the variable in the function. Write out the function being evaluated in one line (promote the variables in the denominator to the numerator and adjust exponents accordingly):
$$f(m,d,L) = \frac{4}{\pi}M^1 d^{-2} L^{-1}$$
You can pick out the values as the exponents of the variables. Thus, for example, the value "-2" is associated with the diameter variable d.
You calculated the relative uncertainty, and both the formula and the result look good. Similar to gneill, I think the online grading system wants the absolute uncertainty.
I've been meaning to get back here to say Thank You! Multiplying by the density was exactly what I needed to do for WebAssign (the online grading system) to accept the answer.
11K