Gas conductance through tube at low pressure

Click For Summary

Discussion Overview

The discussion revolves around designing a tube for a vacuum chamber to achieve a constant rate of pressure change during pumpdown, specifically from atmospheric pressure to approximately 0.5 mbar. Participants explore the application of gas laws and fluid dynamics principles, particularly Poiseuille's law, in the context of varying pressure conditions and flow regimes.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested
  • Mathematical reasoning

Main Points Raised

  • One participant suggests using the ideal gas law (PV=nRT) and Poiseuille's law to determine the necessary tube radius for a desired pressure change rate of 2.5 mbar/sec, but questions the validity of this approach as pressure decreases.
  • Another participant points out that Poiseuille's law applies to incompressible laminar flow and that air behaves as a Newtonian fluid, emphasizing the need to consider the Reynolds number to assess flow conditions.
  • Concerns are raised about achieving a constant mass flow rate due to changing pressure and density in the chamber, suggesting that a fixed restriction may not suffice for maintaining the desired pressure change rate.
  • A suggestion is made for implementing a control system involving a pressure transmitter, ramp controller, and airflow control valve to achieve the desired pressure change.
  • One participant references a resource on compressible fluid mechanics, indicating a potential avenue for further exploration.

Areas of Agreement / Disagreement

Participants express differing views on the applicability of Poiseuille's law in this scenario, with some agreeing on the need for a control system while others focus on the theoretical calculations. The discussion remains unresolved regarding the best approach to calculate the required tube radius under low pressure conditions.

Contextual Notes

Participants note the limitations of applying certain equations under varying flow conditions and the complexities introduced by non-linear behavior in the control system. There is also uncertainty about the linearity of pressure change over time as chamber pressure decreases.

soothsayer
Messages
422
Reaction score
5
I have a vacuum chamber of volume V connected to a vacuum pump via a tube of radius R and length L, and I want to design this "tube" such that R can be controlled in order to provide a constant rate of pressure change inside of the chamber of 2.5 mbar/sec, from atmosphere down to ~0.5mbar. Assume that pumping speed is infinite, so my flow rate is entirely limited by the conductance of my adjustable tube.

At the beginning of chamber pumpdown, I feel like I should be able to apply a combination of PV=nRT and Poiseuille's law to convert my 2.5mbar rate of pressure change into a desired flow rate, and then into an effective tube radius. When I do this, I get a result that feels pretty intuitive based on my experience. (Does this seem like a valid calculation to anyone else or have I already made an error here?)

But nearing the end of the pumpdown process, I suspect I can no longer apply Poiseuille's law in the same way since I no longer have laminar gas flow. When I try to apply the same equations the same way I get a resulting effective radius that is much smaller than would make sense. Is there a different equation I can use to calculate a required tube radius in this low pressure regime? Does it even make sense to be trying to calculate this when my desired change in pressure (2.5mbar/s) is larger than my chamber pressure (0.5mbar)? It doesn't feel like I can treat pressure over time as linear here.

Any advice for tackling this problem would be appreciated!
 
Engineering news on Phys.org
This sketch is my understanding of your problem:
Vacuum.jpg

The Poiseuille equation applies to the incompressible laminar flow of a Newtonian fluid. Air is Newtonian, the Reynolds number needs to calculated to find if the flow is laminar, and this flow is definitely not incompressible.

Your goal is to get a constant rate of pressure decrease in the vacuum chamber. That requires a constant mass flow rate out of the chamber. At the start, the pressure (and the density) in the chamber is high, and the pressure drop is high. Near the end, the pressure (and the density) in the chamber is low, and the pressure drop is low. Regardless of whether the flow in the tube is laminar or turbulent, the mass flow rate will be a function of chamber gas density and pressure drop in the tube. The mass flow rate will be high at the beginning, and decrease as the pressure in the vacuum chamber decreases.

A fixed restriction, whether a tube or an orifice, will not get you a constant rate of pressure decrease in the vacuum chamber. Unless somebody has a better idea, I think you will need some sort of control system. A pressure transmitter, ramp controller, and air flow control valve would do what you want.
 
  • Like
Likes   Reactions: DaveE
soothsayer said:
I have a vacuum chamber of volume V connected to a vacuum pump via a tube of radius R and length L, and I want to design this "tube" such that R can be controlled in order to provide a constant rate of pressure change inside of the chamber of 2.5 mbar/sec, from atmosphere down to ~0.5mbar. Assume that pumping speed is infinite, so my flow rate is entirely limited by the conductance of my adjustable tube.

At the beginning of chamber pumpdown, I feel like I should be able to apply a combination of PV=nRT and Poiseuille's law to convert my 2.5mbar rate of pressure change into a desired flow rate, and then into an effective tube radius. When I do this, I get a result that feels pretty intuitive based on my experience. (Does this seem like a valid calculation to anyone else or have I already made an error here?)

But nearing the end of the pumpdown process, I suspect I can no longer apply Poiseuille's law in the same way since I no longer have laminar gas flow. When I try to apply the same equations the same way I get a resulting effective radius that is much smaller than would make sense. Is there a different equation I can use to calculate a required tube radius in this low pressure regime? Does it even make sense to be trying to calculate this when my desired change in pressure (2.5mbar/s) is larger than my chamber pressure (0.5mbar)? It doesn't feel like I can treat pressure over time as linear here.

Any advice for tackling this problem would be appreciated!
Hi There!

This is a super fun problem. Are you still looking for help?

Check out the PDF by Genick Bar-Meir, "Fundamentals of Compressible Fluid Mechanics"

Let me know if you need help.
 
  • Informative
Likes   Reactions: berkeman
jrmichler said:
I think you will need some sort of control system. A pressure transmitter, ramp controller, and air flow control valve would do what you want.
Yep. This!
You will absolutely have to have a some flow control "valve" that can be adjusted. You will also need a pressure sensor to measure what you ultimately want. The rest is just a (nonlinear) control system to make it work.
 

Similar threads

  • · Replies 6 ·
Replies
6
Views
3K
  • · Replies 9 ·
Replies
9
Views
2K
Replies
5
Views
2K
  • · Replies 3 ·
Replies
3
Views
3K
Replies
8
Views
2K
  • · Replies 3 ·
Replies
3
Views
2K
Replies
7
Views
2K
  • · Replies 10 ·
Replies
10
Views
6K
  • · Replies 3 ·
Replies
3
Views
3K
Replies
7
Views
3K