How PMMC and MI Instruments Read Average Values for DC Waveforms

  • Thread starter Thread starter cnh1995
  • Start date Start date
  • Tags Tags
    Instruments
Click For Summary

Discussion Overview

The discussion revolves around how PMMC (Permanent Magnet Moving Coil) and MI (Moving Iron) instruments read average values for DC waveforms, including pure DC and pulsating DC. Participants explore the mechanics behind the readings of these instruments, particularly in relation to full wave rectified outputs and square waves, while questioning the underlying physics principles and the impact of mechanical inertia.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested
  • Mathematical reasoning

Main Points Raised

  • Some participants inquire how PMMC instruments read average values from pulsating DC and whether the pointer can stabilize at the average value despite the pulsating nature of the current.
  • One participant suggests that mechanical inertia plays a role in how the pointer comes to rest, referencing the concept of a flywheel and magnetic damping effects.
  • Another participant questions the relationship between mechanical inertia and the average value, seeking clarification on the mathematical principles involved.
  • There is a discussion about the behavior of the meter when fed a square wave with a specific duty cycle, with one participant hypothesizing that the meter would read zero during periods of no current.
  • Some participants discuss the difference between average and RMS values, noting that test meters may report RMS values while responding to average values, particularly in the context of AC measurements.
  • Concerns are raised about the accuracy of meters when measuring AC on DC scales, with participants sharing insights on how meters are calibrated to display RMS values for sine waves.

Areas of Agreement / Disagreement

Participants express varying levels of understanding regarding the mechanics of PMMC and MI instruments, with some agreeing on the role of inertia and damping while others challenge the interpretations of how these factors affect readings. The discussion remains unresolved on several technical points, particularly regarding the mathematical relationships and the behavior of meters under different conditions.

Contextual Notes

Participants reference various physical principles and equations related to torque and mechanical balance, but there are unresolved assumptions about the specific conditions under which the meters operate. The discussion also highlights potential limitations in understanding the behavior of meters with different waveforms and frequencies.

Who May Find This Useful

This discussion may be useful for students and professionals in electrical engineering, physics, and instrumentation who are interested in the operational principles of measurement devices and the nuances of reading average versus RMS values in different waveform contexts.

cnh1995
Homework Helper
Gold Member
Messages
3,489
Reaction score
1,163
How does a PMMC instrument read average value of a dc waveform? Is it true for pure dc or pulsating dc as well? I mean what will the meter show if I fed a full wave rectifier's output to it? Will it show its average vaue 2Vm/π? If yes, I don't understand how. Say the meter is an ammeter. So, the current will be pulsating and hence the torque will also be pulsating. How will the pointer come to rest at the average value on the dial? Is there any specific arrangement inside the meter or am I missing some key physics principle here? Same goes for a moving iron instrument. I know that the torque is proportional to the square of current but the current is alternating;hence the torque should also be pulsating. How does the pointer come to rest at the rms value on the dial?
Thanks a lot in advance..
 
Engineering news on Phys.org
cnh1995 said:
How does the pointer come to rest at the average value on the dial?...am I missing some key physics principle here?...

simple mechanical inertia. Think flywheel.
That's why most DMM's have an averaging filter before the ADC.There's also some magnetic damping by eddy current in the brass or aluminum coil former. It makes a single shorted turn.
Some fine pocket compasses use that same principle rather than liquid fill.
 
jim hardy said:
simple mechanical inertia.
But how exactly on the average value on the dial? Why can't it be anywhere else? I mean for full wave rectificatied output, how does the mechanical inertia make the pointer stop at 2Im/π? What is the math involved here which relates this inertia to average value?
 
cnh1995 said:
But how exactly on the average value on the dial? Why can't it be anywhere else?
I don't know your background
are you taking a physics class?

Answer to your question is
inertia only keeps it from instantaneous movement all over the scale.
Where it comes to rest is determined by Force balance, sum of torques = zero
current causes clockwise torque
that's opposed by a coiled hairspring giving counterclockwise torque proportional to angular displacement
when the two are equal the needle stops
try a search on d'arsonval movement
http://www.tpub.com/neets/book3/7b.htm
32NE0219.GIF


and on T(orque)=NIABsinθ
F29_20.jpg

http://helios.augustana.edu/~dr/203/203-s02/img/F29_20.jpg

^there's your electromagnetic torque proportional to current, NIABsinθ
and red spiral in other picture is the opposing hairspring that makes mechanical torque kX , where k is spring constant and x is angular displscement
NIABsinθ = kX ,
If you add inertia it looks like an invitation to oscillation, doesn't it ? Displacing force plus restoring force with no mention of damping? Harmonic motion?

Needle increases both its moment of inertia and damping from air friction.
The metal band on which coil is wound is a shorted turn for magnetic damping. To visualize magnetic damping look at videos of dropping a magnet down a copper tube.get yourself a cheap analog meter and play with it, they're $9.95 at Walmart or ~$2 in junkshops.above all have fun

old jim
 
  • Like
Likes   Reactions: cnh1995
Ok. Suppose I fed a square wave of frequency 1Hz and duty cycle 25% to a PMMC voltmeter. Now, for 0.25s, there will be a current through the coil of the voltmeter and for the next 0.75s , there won't be any current, hence no torque. What will the meter read in this case? I guess it will read 0 since torque is 0 for 0.75s.
 
jim hardy said:
I don't know your background
are you taking a physics class?
I am studying EE..
 
  • Like
Likes   Reactions: jim hardy
cnh1995 said:
Ok. Suppose I fed a square wave of frequency 1Hz and duty cycle 25% to a PMMC voltmeter. Now, for 0.25s, there will be a current through the coil of the voltmeter and for the next 0.75s , there won't be any current, hence no torque. What will the meter read in this case? I guess it will read 0 since torque is 0 for 0.75s.

Have you taken statics and dynamics yet ? ∑forces = 0 in statics, ∑forces = ma in dynamics
You've asked a Dynamics problem : sum of the forces = moment of inertia X angular acceleration

You'd see the meter swing upscale then back down toward zero once per second.
A good high sensitivity test meter with long needle would never reach equilibrium.
Because its coil only gets about fifty microamps the needle moves slowly. Just not enough current to make much torque.
A less sensitive panel meter whose coil gets 1000 microamps might very well follow your 1 hz signal. You'd notice it's way less damped.
As you increased frequency the meter would appear to come to rest at position corresponding to average value.
 
  • Like
Likes   Reactions: cnh1995
be aware that test meters respond to average but they indicate RMS.
That's achieved by lying. At each point on the scale they paint the number for RMS volts that'd cause that deflection instead of the number for average volts actually measured. So they only report the truth when measuring a sine wave . Surely you've studied RMS versus average for sine wave? I think it's .707/.636 ...
If you want to prove to yourself they lie - measure DC on the AC scale.

Many DMM's do the same thing.. check specs for words "average responding" or "true RMS"
 
  • Like
Likes   Reactions: cnh1995
jim hardy said:
As you increased frequency the meter would appear to come to rest at position corresponding to average value.
Aha..That's what I too was thinking!:smile: Thanks a lot! It was very helpful..
 
  • #10
jim hardy said:
Surely you've studied RMS versus average for sine wave? I think it's .707/.636 ...
Yes, it's roughly 1.11. So, they actually measure average value and paint the numbers 1.11 times the average values?? And that is for pulsating dc right? Because for ac, average is 0.
 
  • #11
cnh1995 said:
And that is for pulsating dc right? Because for ac, average is 0.

Yes. Measure AC on the DC scale and it reads zero. You should try it with a real meter.
Be sure you're on a DC volt scale high enough for the AC you're testing . Those little wires of the meter coil will burn up at less than an amp.
 

Similar threads

  • · Replies 8 ·
Replies
8
Views
8K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 6 ·
Replies
6
Views
3K
  • · Replies 1 ·
Replies
1
Views
2K
Replies
1
Views
5K
  • · Replies 5 ·
Replies
5
Views
3K
  • · Replies 7 ·
Replies
7
Views
10K
  • · Replies 6 ·
Replies
6
Views
7K
  • · Replies 3 ·
Replies
3
Views
4K
  • · Replies 6 ·
Replies
6
Views
4K