If I were to Calibrate a Graduated Cylinder

  • Thread starter Thread starter sponsoredwalk
  • Start date Start date
  • Tags Tags
    Cylinder
Click For Summary
SUMMARY

This discussion focuses on the calibration of a graduated cylinder, specifically achieving accuracy at the 50ml mark. The process involves estimating the volume, weighing the water, and applying the formulas for mass and volume based on density, which varies with temperature. An alternative method is presented, using a scale to measure 50g of room-temperature water directly, which simplifies the calibration process. The distinction between "To Contain" (TC) and "To Deliver" (TD) calibrations is emphasized, with TC being the standard for most graduated cylinders used in laboratories.

PREREQUISITES
  • Understanding of basic fluid mechanics and density concepts
  • Familiarity with the principles of mass and weight measurement
  • Knowledge of temperature's effect on water density
  • Experience with laboratory equipment such as graduated cylinders and scales
NEXT STEPS
  • Research the differences between "To Contain" (TC) and "To Deliver" (TD) calibrations in laboratory settings
  • Explore methods for accurately measuring liquid volumes using volumetric pipettes
  • Learn about the impact of temperature on liquid density and its implications for volumetric measurements
  • Investigate best practices for calibrating laboratory glassware to ensure precision in experiments
USEFUL FOR

Laboratory technicians, chemists, and anyone involved in precise liquid measurements and calibrations will benefit from this discussion.

sponsoredwalk
Messages
531
Reaction score
5
If I were to calibrate a graduated cylinder, I would;

1. Guesstimate the 50ml mark for a graduated cylinder & fill with water of a known temperature.

2. Drain the water out & weigh it, If you weigh it in a separate container you must know the weight of the separate container first & subtract this from the final result.

3. Using the formula [mass=weight/gravity] we find the waters mass & then use

[volume = Density/mass] - density being known as a function of temperature - to find out how

close the volume was to 50ml.

4. Repeat, adding or removing water of the same temperature until you've arrived at 50ml.



Of course there are easier methods but this is one possible way of getting a reasonably accurate graduated cylinder, right?




:biggrin: Gracias :biggrin:
 
Chemistry news on Phys.org
sponsoredwalk said:
If I were to calibrate a graduated cylinder, I would;

1. Guesstimate the 50ml mark for a graduated cylinder & fill with water of a known temperature.

2. Drain the water out & weigh it, If you weigh it in a separate container you must know the weight of the separate container first & subtract this from the final result.

3. Using the formula [mass=weight/gravity] we find the waters mass & then use

[volume = Density/mass] - density being known as a function of temperature - to find out how

close the volume was to 50ml.

4. Repeat, adding or removing water of the same temperature until you've arrived at 50ml.



Of course there are easier methods but this is one possible way of getting a reasonably accurate graduated cylinder, right?




:biggrin: Gracias :biggrin:

Dont' use a 2nd cylinder...
 
Step 4 would be unnecessary if you actually filled exactly to the 50ml mark in step 1.
 
Put the un-marked cylinder on a scale and tare it. Add room-temperature water until the scale reads 50g. Scribe at the meniscus level. Done.
 
turbo-1 said:
Put the un-marked cylinder on a scale and tare it. Add room-temperature water until the scale reads 50g. Scribe at the meniscus level. Done.

This will give you a calibration known as 'To Contain" or TC. If you want to use it to deliver 50 mL you will need the second vessel. That calibration would give you something known as a "To Deliver" or TD calibration. Every grad cyl I have ever used is of the TC type.
 
chemisttree said:
Every grad cyl I have ever used is of the TC type.
Why is that? No one performs a reaction in a grad cyl. Wouldn't it make more sense for labs to stock TD calibrated grad cylinders if you're going to pour the stuff out? Or is it that it rarely makes enough of a difference to matter?
 
Yes, Gokul. All your precise work in making up standard solutions are made up using precise balances for solids and volumetric pipettes of some type of the TD type. Volumetric vessels are invariably used as TC since they are designed to be added to until a certain volume is attained. The resulting solution is precise but no one would ever use one of those volumetric flasks or grad cyls to accurately deliver known volumes. In cases where they are used to measure reagents or solvents, they are exact enough for the purpose at hand, of course.
 

Similar threads

Replies
7
Views
4K
  • · Replies 4 ·
Replies
4
Views
4K
  • · Replies 25 ·
Replies
25
Views
4K
  • · Replies 14 ·
Replies
14
Views
1K
  • · Replies 2 ·
Replies
2
Views
1K
  • · Replies 20 ·
Replies
20
Views
3K
  • · Replies 8 ·
Replies
8
Views
3K
  • · Replies 3 ·
Replies
3
Views
5K
  • · Replies 5 ·
Replies
5
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K