Instrument Accuracy: Does Error Remain % or Change to +-V?

  • Thread starter Thread starter ineedmunchies
  • Start date Start date
  • Tags Tags
    Accuracy Instrument
Join the discussion
Registration is free. Ask a follow-up in this thread, or start your own.
3 replies · 12K views
ineedmunchies
Messages
42
Reaction score
0
If an electromechanical instrument's (say a galvanometer) accuracy is given as a percentage of its full scale deflection, what would its accuracy be at anything lower than full scale deflection?

My thoughts:
If say full scale deflection is 100V
The error is +-1% which is +-1V,
Does that mean at a reading of 50V it would still be +-1V which would be be +-2%?

Or does it remain as a percentage error, and the error would be +-0.5 V at +-1% still?
 
Engineering news on Phys.org
Galvanometers are linear so expect 1% error throughout its range.
 
Sorry i meant to type voltmeter in there.
 
If the tolerances are given by the manufactures, that should be valid throughout its range.