Interpreting +/- digits of accuracy in meter

  • Thread starter Thread starter jeffamm
  • Start date Start date
  • Tags Tags
    Accuracy Meter
Click For Summary

Discussion Overview

The discussion revolves around the interpretation of the accuracy specifications of a capacitance meter, specifically focusing on the meaning of "±0.5% ±20 digits" for the 200pf scale. Participants explore the implications of these specifications for measurements, particularly in the context of measuring capacitors of varying values.

Discussion Character

  • Technical explanation
  • Debate/contested
  • Exploratory

Main Points Raised

  • One participant interprets the ±20 digits as indicating a range of ±2pf for a 100pf measurement, leading to a total error of ±2.5pf.
  • Another participant questions the feasibility of the ±20 digits specification, suggesting it may render the last two digits meaningless and proposes it could be a misprint.
  • A participant shares a link to the capacitance meter and expresses agreement with the initial interpretation, while also questioning whether the ±0.5% accuracy refers to the reading or full scale.
  • Concerns are raised about the accuracy when measuring low capacitance values, with one participant noting that the error could be significant (20-30%) for a 10pf capacitor on the 200pf scale.
  • Another participant speculates that the specification might have been miscommunicated, suggesting it could have originally indicated "PLUS 2, MINUS ZERO DIGITS."
  • Discussion includes considerations of alternative meters that might offer better specifications or features, such as zeroing out stray capacitance.

Areas of Agreement / Disagreement

Participants express differing interpretations of the ±20 digits specification, with no consensus reached on its meaning or implications for accuracy. There is also uncertainty regarding whether the ±0.5% accuracy applies to the reading or full scale.

Contextual Notes

Limitations include potential misinterpretations of the specifications and the lack of clarity on how accuracy is defined in relation to different measurement scales.

jeffamm
Messages
26
Reaction score
0
I'm using a capacitance meter that has 3 1/2 digits of display and lists the accuracy of its 200pf scale as follows:

±0.5% ±20 digit

The ±0.5% is simple enough but I've not found something on how exactly to interpret +/- 20 digits of accuracy. I think it means as follows - can anyone confirm?

If I'm measuring a capacitor of 100pf on the 200pf scale my error is as follows:

+/- .5% = +/- .5pf

For digits, since the 3 1/2 digit meter will read 100.0, the meaning of "20 digits" is 098.0 to 102.0 or +/- 2pf

Total error is +/- 2.5pf. Is that correct?
 
Engineering news on Phys.org
Do you have a link to the data for this capacitance meter?

You may be right, but the figures don't really seem possible.

0.5% suggests at least a fairly useful instrument, but +/- 20 digits means that the last two digits are basically meaningless. Could it be a misprint or interpretation error?

I looked at some typical digital capacitance meters and it was common for them to be claimed as + 1 digit (never minus). I wouldn't want one that gave 40 possible readings for the same capacitor.
 
vk6,

Here's the link

http://cgi.ebay.com/Digital-Capacitance-Meter-Tester-Multimeter-6013L-/260711350530?pt=LH_DefaultDomain_0&hash=item3cb39b8102

I'm glad to see that you're interpreting it the same way that I am.

I'm also wondering if the +/- .5% is typically of the reading or full scale?
 
Last edited by a moderator:
That looks like a nice meter.

Why don't you send them an email and ask what that +/- 20 digit means?

They claim they can read to 0.1 pF so they couldn't do that with the last two numbers being almost random.
 
The calculation by jeffamm indicate a maximum error of 2.5pf when measuring a 100pf capacitor. This is 2.5% maximum error. That's not bad using a $21.00 meter.
 
Carl,

That's what I was thinking. But I think the problem comes up when you're reading a low value on any particular scale. For example, measuring a 10pf capacitor on the 200pf scale would ideally read 010.0 so with +/- 20 digits the range would be 8.0 to 12.0. I'm assuming that the +/- .5% is of the reading, so negligible here. If it's of the full scale it's another +/- 1pf of error. That's 20-30% of error. And I guess overall that's what I would expect from an inexpensive meter, which would be great for most applications. I'm interested in doing some comparative measurements of guitar cable capacitance and this error in the low measurements may be a problem for me. The more expensive meters have a +1 digit error as mentioned earlier.
 
They claim an accuracy of 1 part in 1999, (eg 0.1 pF in 199.9 pF) so there must be something wrong with this specification.

I wondered if the original document may have said "PLUS 2, MINUS ZERO DIGITS" and someone has tried to put it as +/- 2,0 then this was typeset without the comma.

It would be worth clarifying this before you buy one of these meters.

The meter looks OK for the price, although a better meter would possibly have a control for zeroing out the stray capacitance of the leads.

Instead of having a lot of boxes that all do different things, maybe you could look for a multimeter that does capacitance, inductance, temperature and probably other things as well as all the normal voltage and current measurements.

Failing that, there are RLC (resistance, inductance, capacitance) meters that specialize in measuring just these things.
 

Similar threads

  • · Replies 2 ·
Replies
2
Views
752
  • · Replies 1 ·
Replies
1
Views
833
Replies
4
Views
2K
  • · Replies 14 ·
Replies
14
Views
6K
Replies
3
Views
1K
  • · Replies 5 ·
Replies
5
Views
2K
  • · Replies 53 ·
2
Replies
53
Views
8K
  • · Replies 1 ·
Replies
1
Views
1K
  • · Replies 3 ·
Replies
3
Views
12K
  • · Replies 8 ·
Replies
8
Views
7K