Load Cell Application Q&A: Single Cell for Loss in Weight Feeder

Click For Summary

Discussion Overview

The discussion revolves around the application of a single load cell in a loss in weight feeder system. Participants explore the challenges of achieving repeatable measurements when using a single load cell mounted on a hinged hopper, and the implications of design constraints on the setup.

Discussion Character

  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • One participant notes that using a single load cell may lead to issues with x-y components of weight measurement and suggests using two load cells for better accuracy.
  • Another participant questions the consistency of the hinged apparatus and suggests that a perfect hinge joint is necessary for repeatable measurements.
  • Concerns are raised about the potential sideways load on the hinge if it is not level, which could affect measurement accuracy.
  • One participant mentions that the stiffness of the hopper could introduce uncertainties in the measurements due to deflection under load.
  • A suggestion is made to install an additional load cell temporarily to test the weight distribution between the hinge and the load cell.
  • Participants discuss the importance of ensuring that the center of gravity remains stable during operation to achieve reliable readings.

Areas of Agreement / Disagreement

Participants express varying opinions on the effectiveness of using a single load cell versus multiple load cells. There is no consensus on the best approach, and several competing views on the design and mechanics of the setup remain unresolved.

Contextual Notes

Participants highlight limitations related to the design constraints of the hopper and the potential impact of hinge alignment and bearing type on measurement accuracy. The discussion does not resolve these issues.

ghaulk
Messages
3
Reaction score
0
I have a question regarding a load cell application I am working on. I am designing a loss in weight feeder system and would like to use a single load cell for my hopper weight. Upon preliminary trials, I mounted the load cell on one side of the hopper and hinged the other. This does not give me repeatable measurements. I would hang the hopper, but I can't because of design constraints. I am doing this in an effort to save $. I have the load cell connected to a signal conditioner which sends info. to a PLC. I know I can put 3 load cells on the hopper with a summing board, but I was wondering if anyone had any suggestions on how to use one load cell for this application?
Thanks
 
Engineering news on Phys.org
I noticed that there has been no replies to my question. I'm new in this forum. Is this the right place to ask these types of questions?
 
Your problem will always be the x-y components of the weight as measured by a load cell. Typically, this sort of measurement is done with a single ball swivel and two load cells, one at x and one at y.

If your hinge joint were perfect, you should get repeatable measurements; are you using a ground shaft and bearings?

Further, how are you transmitting load to the load cell; sounds like you might be getting some tensile component at one edge. A smooth, domed button (no screws please) works best.
 
Welcome to PF, Ghaulk.
This might be better-suited for Engineering than here, but I'm not sure. Anyhow, your question is a bit vague. For instance, why do your design constraints forbid suspending the hopper? That would definitely be my first approach to the problem. I'm also a bit confused about why your hinged apparatus isn't consistent. It sounds like a good idea to me. Are your hinges well lubricated?

edit: Hi, Tvp. You sneaked in while I was composing.
 
Yes, I am using a polished shaft and a linear bearing, but the fit is not perfect (it was ordered by someone else). While waiting on a response I did more experimenting and found that I may be able to hang the hopper. This would definitely be the best way to go since the center of gravity would remain constant. I really appreciate the advice. Thanks.
 
Good luck with it, then. Keep us posted as to your results. :smile:
 
Danger said:
I'm also a bit confused about why your hinged apparatus isn't consistent. It sounds like a good idea to me.
Problem is that if the hinge is slightly off level then depending where the c of g of the load in the hopper is, you will have some sideways load on the hinge.
You could do it with 2 point contact pivots and a single load cell, or as TVP45 said with a single pivot point and 2 load cells.

Are your hinges well lubricated?
I propose that as the engineering forum secret greeting!
 
ghaulk said:
Yes, I am using a polished shaft and a linear bearing,

Ach, I missed that. By chance do you have a plastic or composite linear bearing rather than recirculating steel balls? If so, and if your load is fairly small, you may be getting significant stiction.
 
mgb_phys said:
Problem is that if the hinge is slightly off level then depending where the c of g of the load in the hopper is, you will have some sideways load on the hinge.
Good point. I had assumed an ideal setup, which is unlikely.

mgb_phys said:
I propose that as the engineering forum secret greeting!

:smile:
 
  • #10
One other thing you may want to look at is the stiffness of the hopper. Depending on the loads, you may be deflecting the hopper enough to also introduce uncertainties.
 
  • #11
It does not make much sense to dig into structural details before the very basic mechanics is examined - The weight of hopper and material inside the hopper is shared between the hinge and the load cell. If you can make sure the distribution of wieght between the hinge and load cell is even all the time, you get reproducible result by multiplication of load cell reading. If hte distribution of weight between the hinge and load cell is sort of ramdom, you get random result.

As material is discharge from the hopper continuously, suppose the centre of gravity of hopper + material cannot stay stable. Suggest to make a simple test - install another load cell - just for test purpose - to measure the weight going the to hinged side, and compare reading from both load cell to determine whether distribution between the hinge and the working load cell is consistently even.
 
Last edited:

Similar threads

  • · Replies 3 ·
Replies
3
Views
2K
Replies
11
Views
4K
  • · Replies 2 ·
Replies
2
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 16 ·
Replies
16
Views
5K
  • · Replies 9 ·
Replies
9
Views
2K
  • · Replies 1 ·
Replies
1
Views
4K
  • · Replies 2 ·
Replies
2
Views
5K
  • · Replies 2 ·
Replies
2
Views
3K