Oscilloscope probe resonantly picking up 50Hz

Click For Summary
SUMMARY

The discussion centers on the phenomenon of oscilloscope probes picking up noise from AC power cables, specifically at 50 Hz. Harry observed that connecting an antenna to the probe significantly increased voltage readings, prompting a deeper investigation into the mechanisms at play. Hman clarified that the observed coupling is primarily due to near-field effects, particularly capacitive coupling, which can lead to substantial voltage readings even with minimal coupling capacitance. The calculations provided indicate that a mere 0.5pF coupling can result in a coupled voltage of approximately 380 mV, demonstrating the sensitivity of oscilloscope probes to electromagnetic interference.

PREREQUISITES
  • Understanding of near-field coupling principles
  • Familiarity with oscilloscope probe characteristics
  • Basic knowledge of capacitive coupling and impedance
  • Experience with AC power systems and frequency effects
NEXT STEPS
  • Explore the principles of near-field electromagnetic coupling
  • Study the effects of capacitive coupling in electronic measurements
  • Learn about oscilloscope probe specifications and their impact on measurements
  • Investigate methods to mitigate noise pickup in oscilloscope measurements
USEFUL FOR

Electronics engineers, oscilloscope users, and anyone involved in troubleshooting noise issues in electronic measurements will benefit from this discussion.

H_man
Messages
143
Reaction score
0
Hi all,

I've noted for years that oscilloscope probes tend to pick up noise from AC power cables, but just recently I tried connecting an antenna (piece of wire) to the probe and placing it near some power cables and voltage recorded was off the scale.

I have calculated that the power radiated from a few meters of normal everyday cable with say an amp or two flowing at 50 Hz should be of the order of 10^-10 Watts, so what is the probe picking up? After all, if I'm not mistaken to build a resonant circuit at 50 Hz you either need a lot capacitance or a lot of inductance and I don't think the probe is big enough to contain either relatively large components.

Anyone dealt with this before in any detail?

Thanks
Harry
 
Engineering news on Phys.org
Hi Hman, any time you're looking at 50/60 Hz pickup you're going to be much closer than a wavelength so it's pretty much always near field coupling (mutual inductance in the case of the H field and capacitive coupling in the case of the E field).

The scope probe is like a little dipole, so you're looking at near field E coupling here - basically just capacitive coupling.

Say for example you had just 0.5pF coupling from the power line to the probe, then impedance at 50Hz is about 6300 Meg. Your using 50Hz so I'll assume 240 volts, the scope input is typically about 10M, so the coupled voltage is about 240*10/6300 = 380 mV. That's about 1 volt peak to peak already. Add some wire to increase the capacitance and you'll get even more.
 
Last edited:
Thanks uart, this is now clear to me :-)
 

Similar threads

  • · Replies 7 ·
Replies
7
Views
3K
  • · Replies 12 ·
Replies
12
Views
6K
Replies
20
Views
2K
  • · Replies 7 ·
Replies
7
Views
4K
  • · Replies 24 ·
Replies
24
Views
5K
  • · Replies 4 ·
Replies
4
Views
3K
  • · Replies 2 ·
Replies
2
Views
2K
  • · Replies 7 ·
Replies
7
Views
2K
  • · Replies 10 ·
Replies
10
Views
7K
Replies
3
Views
2K