When does the calibration uncertainty contribute to a measurement uncertainty?

AI Thread Summary
Calibration certificates typically include both the error from the calibration process and the associated uncertainty, yet many uncertainty calculations do not account for calibration uncertainty as a source of measurement uncertainty. This omission may stem from the common requirement for a Test Accuracy/Uncertainty Ratio greater than 4:1, suggesting that calibration uncertainty is often negligible compared to the instrument's uncertainty. However, when evaluating measurement uncertainty, all sources, including calibration uncertainty, should be considered, and methods like root sum of squares can be used to combine these uncertainties. The discussion highlights the importance of understanding the calibration data's context and potential systematic changes that could affect accuracy. Ultimately, while it is sensible to include calibration uncertainty in calculations, caution is advised unless comprehensive calibration data is available.
fonz
Messages
151
Reaction score
5
Often a calibration certificate for an instrument has the error found during the calibration as well as the uncertainty associated calibration itself.

I'm reasearching uncertainty calculations using the GUM 1995 method and I haven't found one yet that includes the uncertainty of the calibration result as a source of measurement uncertainty for a particular instrument. Only uncertainty derived from the error found by the calibration process is used. Is there a reason for this? Is it likely because in most cases there is a Test Accuracy / Uncertainty Ratio greater than 4:1?
 
Engineering news on Phys.org
I suspect it will depend on where you are looking and what you are calibrating. The uncertainty associated with the calibration "experiment" itself will in many cases (pretty much all electrical instruments) be way lower (in some cases a couple of orders of magnitude) than the uncertainty associated with the instrument being calibrated.
That said, if you look up some papers from journals such as Metrologia you will find cases where every part of the error budget is taken into account.
 
f95toli said:
I suspect it will depend on where you are looking and what you are calibrating. The uncertainty associated with the calibration "experiment" itself will in many cases (pretty much all electrical instruments) be way lower (in some cases a couple of orders of magnitude) than the uncertainty associated with the instrument being calibrated.
That said, if you look up some papers from journals such as Metrologia you will find cases where every part of the error budget is taken into account.

Hi thanks for the reply. That is essentially my understanding and I think that the common requirement for a Test Accuracy / Uncertainty Ratio of 4:1 or even 10:1 validates your statement.

It has got me thinking though. When considering the uncertainty of a measurement made by an instrument with a known calibration. Is is correct to assume the error quoted by a calibration certificate can be considered a standard uncertainty or expanded uncertainty and if so, at what level of confidence? For example, I have shown an image of a typical calibration chart for a pressure sensor. The calibration tolerance is shown in blue, and the calibration points highlighted in green along with the error bars representing the uncertainty of the calibration itself. Is it correct to say that the contribution of uncertainty due to the accuracy of this device on any measurement made by it is within +/-2% of span? If so, at what level of confidence?

Calibration Uncertainty.PNG
 
The short answer is "always." All sources of uncertainty always contribute to the total uncertainty of a measurement, including calibration uncertainty. The question, as you've already discussed, is whether that calibration uncertainty is negligible compared to the rest of the sources. When you do a root sum of squares to add uncertainties, small contributions fade away pretty quickly.
 
  • Like
Likes hutchphd
Would it be appropriate to assume a rectangular probability distribution for the calibration tolerance (blue line in the previous figure) and combine it with the expanded uncertainty of the calibration (green error bars) using the root sum squares method to obtain a standard uncertainty for the device at calibration?

If the 'as-found' calibration error is much less than the calibration tolerance then I suppose that is quite conservative so potentially the largest calibration error could be used and assumed to be a rectangular probability distribution. Combining that with the standard uncertainty of the calibration itself using RSS would then derive a standard uncertainty taking into account the calibration of the device and the uncertainty of the calibration itself. It seems sensible to me but it's not made clear in the GUM.
 
fonz said:
If the 'as-found' calibration error is much less than the calibration tolerance then I suppose that is quite conservative so potentially the largest calibration error could be used and assumed to be a rectangular probability distribution. Combining that with the standard uncertainty of the calibration itself using RSS would then derive a standard uncertainty taking into account the calibration of the device and the uncertainty of the calibration itself. It seems sensible to me but it's not made clear in the GUM.
I think this is not a good strategy unless you have more data about the calibration. For instance there might be a systematic change (say perhaps Temperature in your example) that would shift that green line to the other side of the calibration tolerance. Unless you know that the calibration data includes the entire calibration parameter space this is not a good practice.
If you need numbers better than the 2% band, it looks as though the instrument is capable of higher precision, but you would need to supply a correction factor using "controls" i.e. known outcomes. This is often done for medical diagnostic devices
 
Hi all, I have a question. So from the derivation of the Isentropic process relationship PV^gamma = constant, there is a step dW = PdV, which can only be said for quasi-equilibrium (or reversible) processes. As such I believe PV^gamma = constant (and the family of equations) should not be applicable to just adiabatic processes? Ie, it should be applicable only for adiabatic + reversible = isentropic processes? However, I've seen couple of online notes/books, and...
Thread 'How can I find the cleanout for my building drain?'
I am a long distance truck driver, but I recently completed a plumbing program with Stratford Career Institute. In the chapter of my textbook Repairing DWV Systems, the author says that if there is a clog in the building drain, one can clear out the clog by using a snake augur or maybe some other type of tool into the cleanout for the building drain. The author said that the cleanout for the building drain is usually near the stack. I live in a duplex townhouse. Just out of curiosity, I...
I have an engine that uses a dry sump oiling system. The oil collection pan has three AN fittings to use for scavenging. Two of the fittings are approximately on the same level, the third is about 1/2 to 3/4 inch higher than the other two. The system ran for years with no problem using a three stage pump (one pressure and two scavenge stages). The two scavenge stages were connected at times to any two of the three AN fittings on the tank. Recently I tried an upgrade to a four stage pump...
Back
Top