Why are the S11 and S22 dips different in my microstrip resonator measurement?

  • Thread starter Thread starter uWave_Matt
  • Start date Start date
  • Tags Tags
    Variation
Click For Summary
SUMMARY

The discussion centers on the discrepancies observed in S11 and S22 measurements of a half-wave microstrip resonator designed for 10GHz operation. The S22 dip occurs at 9.7GHz with a more negative magnitude compared to the S11 dip at 10.5GHz, raising concerns about potential issues with the Vector Network Analyzer (VNA) HP8510, feed line length, or coupling gap variations. Suggestions include using "stepped" mode instead of "swept" mode for measurements and verifying the calibration process, as well as switching ports to ensure consistent results. The resonator's high Q factor, estimated to be over 10,000, indicates significant performance potential, but the observed measurement differences suggest underlying issues that need to be addressed.

PREREQUISITES
  • Understanding of microstrip resonator design and operation
  • Familiarity with S-parameter measurements and Vector Network Analyzers (VNA)
  • Knowledge of calibration techniques for RF measurement equipment
  • Basic principles of high-temperature superconductors (HTSC) and their properties
NEXT STEPS
  • Investigate the use of "stepped" mode versus "swept" mode on the HP8510 VNA
  • Learn about the impact of feed line length on S-parameter measurements
  • Research calibration methods for VNAs, particularly in relation to frequency and magnitude
  • Explore the characteristics and measurement techniques for high-Q resonators
USEFUL FOR

Engineers and researchers working with RF design, particularly those involved in microstrip resonator fabrication and measurement, as well as professionals focused on high-temperature superconductors and their applications in microwave technology.

uWave_Matt
Messages
21
Reaction score
0
I've fabricated a simple half-wave microstrip resonator with gap-coupled microstrip feed lines. It's designed to operate around 10GHz, but due to slight over etching, the half-wave portion is 5-10um shorter than I intended. What might be more significant is that the gap, which was originally designed to be 5um, is now closer to 10um.

I'm not sure if any of the aforementioned is relevant to my question...when I characterize the device on the "trusty" HP8510, my S22 dip (9.7GHz) occurs at a slightly lower frequency than my S11 dip (10.5GHz) and has a significantly greater (more negative) magnitude.

So far, I've double checked all of my calibrations.

I suspect feed line length (where my contact touch the feed line), coupling gap variation, or perhaps a problem with the 8510. Does anyone have any other thoughts?

I've posted this question on another forum and it was suggested that I use "stepped" mode instead of "swept" mode. Does anyone know where these settings are?
 
Engineering news on Phys.org
If the resonator is symmetric S11 should be indentical to S22, meaning your result doesn't make physical sense which in turn would point to a problem with the VNA.
Have you tried switching ports? I.e. connected port 1 to "port B" of the resonator and vice versa. The result of a reflection measurement should obviously not depend on which port of the VNA you use.

Also, it can't be a calibration issue as such; when you calibrate a VNA you calibrate the magnitude and not the frequency (although if the calibration if way off the position of the peak might appear to move; simply becaue the magnitude is wrong).

What is the Q (roughly) of your resonator?
 
f95toli said:
If the resonator is symmetric S11 should be indentical to S22, meaning your result doesn't make physical sense which in turn would point to a problem with the VNA.
Have you tried switching ports? I.e. connected port 1 to "port B" of the resonator and vice versa. The result of a reflection measurement should obviously not depend on which port of the VNA you use.

I've rotated the jig that the resonator is in which effectively causes S11 and S22 to switch. This is just the room temperature measurement. I'm dealing with HTSC material so the S-parameters should be significantly different after I cool the chamber.


f95toli said:
Also, it can't be a calibration issue as such; when you calibrate a VNA you calibrate the magnitude and not the frequency (although if the calibration if way off the position of the peak might appear to move; simply becaue the magnitude is wrong).

But you have to calibrate the magnitude over a certain frequency range.

f95toli said:
What is the Q (roughly) of your resonator?

Since it's an HTSC, the Q-should be just about humongous. Another HTSC resonator fabricated in this lab had Q>10000 and I've calculated the Q of this one to be close to that.

I should also mention that one of the coaxial lines from the VNA to the jig is 4-5 time longer than the other, although I'd assume calibration would take care of this.
 