X-ray crystallography; intensity peaks vary with voltage

Click For Summary
SUMMARY

The forum discussion centers on an experiment using a diffractometer to measure the lattice constant of NaCl at electron voltages of 20 kV and 30 kV. The results indicate that the characteristic peaks (Kα and Kβ) appear at slightly smaller angles (0.5 - 1 degrees) for 20 kV compared to 30 kV, which contradicts theoretical expectations. Participants debate potential causes for this discrepancy, including mechanical issues in the experimental setup and crystal impurities, while referencing Bragg's law for further analysis.

PREREQUISITES
  • Understanding of Bragg's law (nλ = 2d sin Θ)
  • Familiarity with X-ray diffraction techniques
  • Knowledge of lattice constants and crystal structures
  • Experience with diffractometer operation
NEXT STEPS
  • Research the impact of electron voltage on X-ray diffraction patterns
  • Explore methods for minimizing experimental errors in crystallography
  • Investigate the effects of crystal impurities on diffraction results
  • Learn about advanced diffractometer calibration techniques
USEFUL FOR

Students and researchers in materials science, crystallography, and physics, particularly those conducting experiments involving X-ray diffraction and lattice measurements.

ngc2024
Messages
15
Reaction score
0

Homework Statement


I have done a simple lab experiment to find the lattice constant for NaCl. Using a diffractometer, radioactive counts were measured as a function of crystal orientation angle for electron voltages of 20 and 30 kV. The average lattice constant found wasclose to the accepted value, but closer for 20kV.

The issue is that the graphs for the two voltage levels are very similar, but the characteristic peaks (Kα and Kβ) consistently occur at slightly smaller (about 0.5 - 1 degrees) for 20 kV than for 30kV. My understanding is that the peaks should theoretically be in the same position, and I am not able to suggest any likely reason for this apparent systematic error.

Homework Equations


Bragg's law
nλ = 2d sin Θ

The Attempt at a Solution


I have been trying to figure out if this is a known phenomena or just due to some experimental flaw such as crystal impurities. Several sources say the peaks should be in the same position (e.g. http://pd.chem.ucl.ac.uk/pdnn/inst1/xrays.htm), but I have also found claims for the opposite (e.g page 6 in https://books.google.no/books?id=i_...ce=gbs_ge_summary_r&cad=0#v=onepage&q&f=false) .

Either way, I don't understand the underlying mechanism, and help would be greatly appreciated!
 
Physics news on Phys.org
ngc2024 said:
but I have also found claims for the opposite (e.g page 6 in https://books.google.no/books?id=i_...ce=gbs_ge_summary_r&cad=0#v=onepage&q&f=false) .
The angle discussed there is the x-rax emission relative to the incoming electron beam, it has nothing to do with the refraction angle you measured.

I would expect some mechanical issue - the crystal tilted a bit, or some other parts of the setup moved a bit relative to each other. For an ideal setup the peak should be nearly* exactly at the same angle.
ngc2024 said:
radioactive counts
There is no radioactivity involved.

* a different background spectrum has a small influence on the precise peak position (unless you determine it with a fit to background and peak), but for narrow and large peaks this effect is negligible.
 
Okay, I see. Thanks!
 