XRD/XRR v. AFM for Film Thickness

  • Thread starter Thread starter IntentKnown
  • Start date Start date
  • Tags Tags
    Film Thickness
IntentKnown
Messages
3
Reaction score
0
Currently in our lab we have a WEDGE shaped sample that we would like to be able to know the thickness at various points along it. Our sample is about 5mm x 5mm. The wedge SHOULD extend from one side to the other. At the highest point, the wedge is 50nm, and then drops down, hopefully at a constant rate, to 0nm at the other side.

**Special Note**
The wedge is made of PZT and is between two layers of LSMO, with the layer on top of the wedge measuring a constant 5nm.

We were thinking about using the XRR/XRD to find the thickness, but I've recently been told to try the Atomic Force Microscope because of our unique shape, and even perhaps the Scanning Electron Microscope.

I have no personal experience using either the AFM or the SEM and have no idea how they would compare to using the XRR

Thoughts/ Concerns from the forum please.
 
Physics news on Phys.org
Hi,

I think all three of the techniques you mention could answer your question:

XRR: thickness interference fringes in XRR can be modeled with the Parratt formalism (e.g. using Parratt32.exe), and give very reliable thickness and roughness values (~0.1 nm precision for good data and fits). As your sample shape is non-standard, you would have to constrict your x-ray beam to ~0.1 mm sideways, so as to not sample the whole wedge but just a part of it. This is doable on a good reflectometer, but you sacrifice a lot of flux, which makes it challenging

AFM: I do not know what the abbreviations of your materials stand for, but if you can make a scratch down to the substrate you can scan the scratch then and (taking care to look beyond the bulge) extract film thickness. z-resolution should be up to 0.5 nm.

SEM / TEM: You could break your sample an look at the cross section. IN SEM one would expect a resolution of 10nm, TEM where sample preparation would be much more involved can get much more precise.

I would try AFM first (simplest), then XRR.
 
From the BCS theory of superconductivity is well known that the superfluid density smoothly decreases with increasing temperature. Annihilated superfluid carriers become normal and lose their momenta on lattice atoms. So if we induce a persistent supercurrent in a ring below Tc and after that slowly increase the temperature, we must observe a decrease in the actual supercurrent, because the density of electron pairs and total supercurrent momentum decrease. However, this supercurrent...