Hello, I am confused as to how to determine the accuracy of a measurement of an AC signal from a Data Acquistion Board.
I went to NI's website to look at the the specifications of the board I was using (a USB-6009) here, and I don't know whether the "accuracy at full scale" is referring to the scale setting of the board or the scale of the signal that the board is actually receiving. If someone could clear this up for me, I would really appreciate it. Thanks! |
| May29-11, 01:08 PM | #2 |
|
|
It depends on the range you selects.
For measure wrt GND, range it's +/-10 V It means that you have to divide 20V by 2^n, n length of the output register. |
| May29-11, 04:48 PM | #3 |
|
|
If so, Quinzios equation is for the accuracy of the A/D but you need to also consider your input sensing device. There will be error there that will propagate through your DAQ board. |
| May29-11, 06:34 PM | #4 |
|
|
Accuracy of measurements from DAQ boardSo what I am asking is how can I determine the error in the measurement I am taking? I'm pretty ignorant on how to use these things, so if you could avoid using acronyms that would be helpful. I don't know what an A/D is for instance... |
| May29-11, 10:43 PM | #5 |
|
|
In order to find the error you need to know the error/uncertainty of the function generator for a particular setting. (see your function generators specifications) e.g. with a sine wave frequency setting of 100 hertz and a amplitude setting of 1 volt, there will be an uncertainty in the frequency and an uncertainty in the amplitude. Then there's the error/uncertainty of the A/D. The error of the function generator will get propagated through the A/D so it's not necessarily as simple as adding the errors to get a final accuracy. This is called Propagation of uncertainty. The article has some example formulas with two real variables A & B with standard deviations σA & σB. |
| May30-11, 03:19 PM | #6 |
|
|
All right, I think I have got it figured out now.
The resolution of the Analog to Digital conversion that the DAQ makes can be calculated using the formulas here: http://en.wikipedia.org/wiki/Analog-...ter#Resolution Given 14 bits, that means that the resolution that the board can produce at a ±10V would be 1.221mV. However, since there are a number of other factors that could effect the measurement produced, it would probably be better to use the values on NI's website for the range that the measurement is made at. I'm assuming the table titled "Absolute accuracy at full scale, differential" is what I need to use. So for ±10V the actual level of inaccuracy is 7.73mV. However, now I can't figure out what the uncertainty is with respect to the signal produced by the function generator. The Specifications state that the "accuracy" and "stability" are both ±20ppm, but it seems like 10^-6 Hz and Volts are awfully small. Does it actually have something to do with the resolution? I can't find any guides that will tell me what all these numbers really mean.... |
| May30-11, 08:51 PM | #7 |
|
|
|
| New Reply |
| Tags |
| daq, function generator, uncertainty, waves |
| Thread Tools | |
Similar Threads for: Accuracy of measurements from DAQ board
|
||||
| Thread | Forum | Replies | ||
| PCI Express (PCIe) Cable for board to board link | Electrical Engineering | 1 | ||
| PCI express for board-to-board link. | Electrical Engineering | 3 | ||
| Accuracy of Current and Voltage Measurements | Introductory Physics Homework | 3 | ||
| Re: QED accuracy | General Physics | 0 | ||