[Biomedical engineering] Question about ECG signal electrodes

AI Thread Summary
The ground electrode in ECG systems helps minimize interference by measuring and inverting the interference signal, which is then fed back through the ground lead, reducing the net interference through superposition. However, directly connecting a patient to ground can be dangerous as it may allow current to flow through the body along a low resistance path, posing a risk of electric shock. The discussion highlights the importance of understanding the capacitance between power lines and the patient, which can introduce significant noise into the ECG signal. A specific example illustrates that the mains potential can create a noise input that exceeds the actual ECG signal, complicating the interpretation of readings. Overall, proper grounding techniques are crucial for effective ECG signal acquisition while ensuring patient safety.
Lajka
Messages
68
Reaction score
0
Hiyo,

Could someone please explain to me how exactly, or why, does the ground electrode in ECG minimize the interference?

And, also, why is it said that the patient could be at risk because of the ground electrode?

I really don't get it, so if there's anyone here who could clarify this for me, that would be awesome!

Much appreciated,
Lajka
 
Engineering news on Phys.org
Humans act as antennas, picking up interference from the mains at 50/60Hz depending on the country you're in. ECG signals are drowned out in the presence of this interference, so what's usually done is that the interference is measured by the circuit, flipped, and fed back in a small amount through the ground lead. Superposition of the two opposite signals means the net amount of interference is reduced.

Simply connecting someone to ground is dangerous as it invites current to flow through the body along a relatively low resistance path.
 
Hi, man, thank you for your answer!

Would you happen to know some litterature/website where I could learn more about this? It sounds very interesting, and I am pretty new in all this.

Thanks!
 
(this is my another account, can't remember the password at the moment :D)

Whoa, that looks awesome! Thank you very much for that!

I'll have a closer look when I come to the office. In the meantime, I tried to make a small drawing (before I saw this post) to see if I can graphically explain to myself what you have written in your first post.

You can see it here:
http://ge.tt/4WQvuTe/v/0
(sorry for the smallish text, but it should be readable if you zoom in a little)

Maybe you can have a look and tell me if that is/isn't what you meant?

Again, thank you very much for your help!

P.S. Okay, just one question, I threw a glance at the paper before I get out. It says:
The ECG voltage V is not the only signal found at the input of the amplifier; one major
source of interference is the electrical power system. Capacitance between power lines in the
wall, floor and ceiling and nearby equipment couples current into the patient, wires and
machine. This current flows through the skin−electrode impedances on the way to ground.
The capacitance to these power line sources varies with proximity but is of the order of 50pF
which corresponds to an impedance of 64M Ω at 50Hz. If the right leg is connected to the
common ground of the amplifier through an electrode with contact impedance of, say, 5k Ω ,
the mains potential of 240V will appear as a 20mV noise input.
This value is well in excess
of the ECG signal itself.

Well, if we connect the right leg to the ground, shouldn't noise disappear because we short-circuited that capacitance between the human and the ground (SCENARIO 1 in my sketch above)?
 
Last edited:
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I have recently moved into a new (rather ancient) house and had a few trips of my Residual Current breaker. I dug out my old Socket tester which tell me the three pins are correct. But then the Red warning light tells me my socket(s) fail the loop test. I never had this before but my last house had an overhead supply with no Earth from the company. The tester said "get this checked" and the man said the (high but not ridiculous) earth resistance was acceptable. I stuck a new copper earth...
Thread 'Beauty of old electrical and measuring things, etc.'
Even as a kid, I saw beauty in old devices. That made me want to understand how they worked. I had lots of old things that I keep and now reviving. Old things need to work to see the beauty. Here's what I've done so far. Two views of the gadgets shelves and my small work space: Here's a close up look at the meters, gauges and other measuring things: This is what I think of as surface-mount electrical components and wiring. The components are very old and shows how...
Back
Top