Do I Need a Terminator for Transmission Lines?

Click For Summary
SUMMARY

Termination of transmission lines is essential to prevent signal reflections and ringing. In this discussion, the focus is on using a 50 Ohm terminator with RG58 coaxial cable, which also has a characteristic impedance of 50 Ohm. The oscilloscope's input impedance must be verified to ensure accurate measurements; if it is indeed 50 Ohm, no additional terminator is necessary. However, if the oscilloscope's termination feature is malfunctioning, adding a terminator will halve the signal amplitude due to parallel resistance effects.

PREREQUISITES
  • Understanding of transmission line theory and impedance matching
  • Familiarity with RG58 coaxial cable specifications
  • Knowledge of oscilloscope operation and input impedance settings
  • Basic concepts of RC time constant and its implications in signal integrity
NEXT STEPS
  • Learn about the implications of impedance mismatches in transmission lines
  • Research the effects of termination on signal integrity in high-frequency applications
  • Explore the use of digital multimeters (DVM) for verifying oscilloscope input settings
  • Investigate the relationship between cable length, capacitance, and RC time constant in transmission lines
USEFUL FOR

Electrical engineers, technicians working with transmission lines, and anyone involved in signal measurement and integrity in electronic systems.

gareth
Messages
188
Reaction score
0
...no, not the cool one, the one found at the end of transmission lines (sorry).

Well here's my question;

I believe you need to terminate transmission lines with the same impedence as the characteristic impedence of the line itself in order to prevent reflections from the end and 'ringing' in the line.

I'm using an oscilloscope to measure electrical signals, the line is standard RG58 co-ax which has a 50 Ohm characteristic impedence. The imput impedence of the scope is 50 Ohm (or so it says on the outside).

In this case do I need a terminator at all?

A technician recently told me you need to put a 50Ohm terminator on the end of the line (a T connector to the scope input, with a terminator on one end and the input on the other) because the impedence usually marked on the scope isn't correct.

But this confuses me, surley then you're measuring the signal across a 25 Ohm resistor rather than a 50 Ohm because the two resistors are in paralell.

The other thing that bothers me is the RC time constant for this system, if you have miles of cable you're increasing the stray capacitance in the cable (~15pF per foot I think), so you get a larger RC time constant with longer cables, agreed? But how does the termination effect this? Can you eliminate the RC time constant to a minimum using termination techniques?


Any info on termination/terminators would be very welcome, searched around the net but nothing very concise is available.

Thanks
Gareth
 
Engineering news on Phys.org
If your oscilloscope has a 50 Ohm input option, then that will work fine for terminating a 50 Ohm transmission line. Just be sure to turn it on.

The wikipedia.org page on transmission lines is a reasonable intro:

http://en.wikipedia.org/wiki/Transmission_line


.
 
that's what I thought, but the lab technician didn't seem to trust the scope. when I put the other terminator on it decreased the signal amplitude by 1/2 as you would expect from putting another 50 ohm resistor in parallel with 50 ohm.
 
I have had cases where the 'scope 50 Ohm input selection stopped working, but that's rare. You can just verify it with a DVM, since it's a DC 50 Ohm termination that gets switched in. Show that to your lab technician.
 

Similar threads

  • · Replies 6 ·
Replies
6
Views
1K
Replies
1
Views
2K
  • · Replies 7 ·
Replies
7
Views
2K
  • · Replies 10 ·
Replies
10
Views
4K
  • · Replies 39 ·
2
Replies
39
Views
9K
  • · Replies 15 ·
Replies
15
Views
5K
  • · Replies 7 ·
Replies
7
Views
3K
  • · Replies 11 ·
Replies
11
Views
3K
  • · Replies 3 ·
Replies
3
Views
4K
  • · Replies 11 ·
Replies
11
Views
5K