Hameg HM203-5 scope retrace fault

In summary: If that doesn't work, I'll take it to an electronic specialist.2. General cleanup and inspection of soldering and connectors.It sounds like you'll want to do a general cleanup and inspection of the soldering and connectors. It's always a good idea to check for poor connections and worn out parts.3. Recap.In summary, the scope is displaying a retrace line when handling a sine wave. It's only visible on the left 1/4 of the screen. The fault doesn't appear when the square wave cal is used. It's possible the blanking signal is failing, but it's also possible the scope is damaged by using it. If the fault persists after the service,
  • #1
Guineafowl
762
366
This circa 1990 scope has started to display a retrace line when handling a sine wave. It’s only visible on the left 1/4 of the screen. The fault doesn’t appear when the square wave cal is used.

Looking at the block diagram, and from initial thoughts, I’m guessing the blanking signal must be present, but is fading out too quickly and letting the beam through just before the start of the next sweep.

Since the scope is quite old, I think I’ll give it a service soon, including:
1. Contact cleaner for all switches.
2. General cleanup and inspection of soldering and connectors.
3. Recap.

If the fault persists after the service, I’ll start looking into it then. Is this a sensible approach, or does someone know straightaway what might be wrong? I don’t know how common a fault this is. Could the scope be damaged by using it in the meantime?
 
Engineering news on Phys.org
  • #2
Best bet is another scope to troubleshoot the one you have. Looks to me like an excellent opportunity to gain some troubleshooting skills. I doubt you will damage anything running it this way. Does it make a difference what the sweep rate is?
 
  • #3
Averagesupernova said:
Best bet is another scope to troubleshoot the one you have. Looks to me like an excellent opportunity to gain some troubleshooting skills. I doubt you will damage anything running it this way. Does it make a difference what the sweep rate is?
Yes, the retrace line alters according to the sweep rate - shorter on a faster sweep. At some speeds, there’s a vertical line at the start of the sine, as if the sweep is jamming at the start.
 
  • #4
That vertical line will likely change based on how you have the trigger set. It is supposed to hold until the trigger initiates a sweep. If it is not blanking you will see something vertically because you have a signal in the vertical section.
 
  • Like
Likes Guineafowl
  • #5
Guineafowl said:
1. Contact cleaner for all switches.
Be cautious about using contact cleaner at the input circuitry. Tektronix specifically warns not to use contact cleaner there because it degrades the PC board material used.
 
  • Like
Likes Guineafowl
  • #6
Tom.G said:
Be cautious about using contact cleaner at the input circuitry. Tektronix specifically warns not to use contact cleaner there because it degrades the PC board material used.
I tend to go: If it ain't broke, don't fix it. In other words, if none of the switches seem noisy, leave them alone.
 
  • Like
Likes Guineafowl
  • #7
OK, thanks. The switches are a bit finicky, especially the component tester toggle and on/off. I was going to spray in a tiny amount, under guard of a paper towel, and work the switch.
 

1. What is a "Hameg HM203-5 scope retrace fault"?

A "Hameg HM203-5 scope retrace fault" refers to a specific issue that can occur with the Hameg HM203-5 oscilloscope model. It involves the trace on the oscilloscope screen not properly returning to the horizontal center after each sweep, resulting in a distorted display.

2. What causes the "Hameg HM203-5 scope retrace fault"?

The most common cause of the "Hameg HM203-5 scope retrace fault" is a malfunctioning or faulty horizontal sweep circuit. This can be due to aging components, dust or debris buildup, or physical damage to the oscilloscope. It can also be caused by incorrect settings or adjustments.

3. How can I fix the "Hameg HM203-5 scope retrace fault"?

To fix the "Hameg HM203-5 scope retrace fault", you will need to troubleshoot and identify the specific cause of the issue. This may involve checking and cleaning the horizontal sweep circuit, adjusting settings, or replacing faulty components. It is recommended to consult the oscilloscope's manual or seek assistance from a professional if you are not familiar with troubleshooting electronic equipment.

4. Can the "Hameg HM203-5 scope retrace fault" be prevented?

While the "Hameg HM203-5 scope retrace fault" may occur due to natural wear and tear, it can be prevented by regularly maintaining and cleaning the oscilloscope, using proper settings and adjustments, and avoiding physical damage to the device. It is also recommended to have the oscilloscope serviced by a professional every few years to ensure it is functioning properly.

5. Can the "Hameg HM203-5 scope retrace fault" affect the accuracy of my measurements?

Yes, the "Hameg HM203-5 scope retrace fault" can affect the accuracy of your measurements as it can distort the display and make it difficult to read or interpret the waveform. It is important to address this issue as soon as possible to ensure accurate measurements and proper functioning of the oscilloscope.

Back
Top