The discussion revolves around the challenges of imaging the bottom of a small plastic container filled with liquid using an optical system. Participants explore issues related to illumination, optical distortion caused by the liquid surface, and potential solutions for achieving even illumination across the image.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • One participant describes a setup involving a camera and a lens with 1x magnification, noting that the illumination works well for an empty container but becomes uneven when liquid is added.
  • Another participant suggests that adding liquid alters the path length and may require re-adjustment of the optical setup.
  • Concerns are raised about the meniscus effect of the liquid surface, which appears to create optical distortion, particularly at the edges of the container.
  • Some participants propose filling the container to the top to flatten the meniscus as a potential temporary solution for testing.
  • There are discussions about the possibility of using a flat coverslip or a corrective lens to mitigate the lensing effect of the liquid surface.
  • One participant mentions the optical density differences between air and water, suggesting that flooding the tube's surroundings with water could help reduce distortion.
  • Another participant raises the idea of using post-processing techniques, such as Fourier transform options, to correct image issues.
  • Several participants express uncertainty about how to measure or calculate the optical effects caused by the liquid surface.

Areas of Agreement / Disagreement

Participants generally agree that the meniscus is causing optical issues, but there is no consensus on the best method to mitigate these effects or how to measure the resulting distortion. Multiple competing views on potential solutions remain present.

Contextual Notes

Participants note limitations in their ability to control the liquid level and the need for the system to function within specific operational parameters. There are unresolved questions regarding the optical properties of the liquid surface and the best corrective measures.

HomeTinker
Messages
14
Reaction score
1
TL;DR
I am trying to solve the issue of the uneven illumination of the imaging surface created by the lens formed from the stop surface of the liquid.
HI

I am currently playing with imaging device that would take pictures of the content of the bottom of the small plastic container. The container is just a tube of 7mm in diameter. When I image an empty tube I get a proper high contrast image of the dust on the bottom of it. However the problem is that I have to take pictures of targets in liquid.

The device has a camera, a lens with 1x magnification (sufficient for the resolution and the target I am trying to image. The problem is the illuminations. I have build a 4f Koehler illumination for the this purpose and it seems to be working fine. When I test the beam of light coming out of it I get mostly parallel beam of light, slightly diverging but not much. When I use that system to image the bottom of the plastic container I get a proper, decent high contrast image of the dust on the bottom of it.

The problem starts when I add liquid into it. The content of the container is no longer evenly illuminated, even though the light outside of it seems to be ok. The effect is that the edges of the container (which is circle) are darker and seems to be slightly out of focus in comparison to the center. The effect is magnified if there is very little liquid added and gradually disappears the more liquid I add (however never getting even close to the picture of the empty container). Clearly the top surface of the water is acting as a lens due to it's meniscus. I tried all sorts of ways to try to mitigate that but with no success and sadly for a good pictures I need even illumination across the entire picture. Small variations I guess would be fine but not more than few percent.

Any ideas how to mitigate this issue in the optical design? How to even measure if this lens made by the water surface is going to be convex or concave?

Any help is very much appreciated.
Many thanks.
 
Science news on Phys.org
Hi. I don't know much about optics but I would imagine that adding liquid into the light path will alter the path length so 'everything' or anything could need re-adjusting. (???)
Perhaps going through the set up procedure with just the tube of liquid first.
 
  • Like
Likes   Reactions: BillTre
Welcome to PF.

HomeTinker said:
I am currently playing with imaging device that would take pictures of the content of the bottom of the small plastic container. The container is just a tube of 7mm in diameter.
Can you post a sketch of the setup, including the position of your illumination sources? Use the "Attach files" link below the Edit window to upload PDF or JPEG files.
 
  • Like
Likes   Reactions: BillTre
I have attached the diagram of the illumination. Please let me know if that helps.
To clarify the illumination area is much larger than the diameter of the container. The illumination can access areas outside of the container, however the issue is that the centre of the container liquid is nice and bright but the edges are dark, as if the illumination was not as bright there or the image of the illumination generated by the liquid surface working as lens was too small for the imaging system.
Hence the question - how can I measure, calculate that? Or alternatively what other lens is needed to mitigate the lens generated by the liquid surface....

Many thanks for help
Illumination.drawio.png
 
Last edited by a moderator:
How much control do you have over the depth of the liquid? If the problem is the meniscus, fill the tube to the top so that the meniscus flattens out.

If that isn't practical in "production" use you could at least use it as a test to confirm that the meniscus is your problem.
 
  • Like
Likes   Reactions: BillTre and Bystander
The meniscus it the problem, I tried to check that by gradually adding more liquid to the container and I can see the effect being drastic with small amount of liquid and reducing when I fill the container to the top. Sadly I have no control over the amount of liquid as the system has to be able to work in a range between 50 and 250ul of liquid in the container. The question is how to compensate for that? Will the distance between illumination system and the surface of the liquid make a difference?
 
HomeTinker said:
I have attached the diagram of the illumination.
I'm sure you could improve on the resolution of the version of your diagram that's been posted.
HomeTinker said:
Or alternatively what other lens is needed to mitigate the lens generated by the liquid surface....
When I have used a microscope I have always used a 'slip' (very thin square of glass) to give a flat upper face to the slide. Perhaps a similar 'flattener' on the top surface of the liquid could work; finding a suitable 'Correcting Lens' could be a problem.
 
  • Informative
  • Like
Likes   Reactions: BillTre and berkeman
@berkeman that was just an idea! Don't rely on it being "informative". it could be BS for all I know. I am a specialist in bodgery in the things I make.
 
  • #10
I tried looking at your diagram, but the printing was too small to make out.

It seems that the direction of light is down, so it is an inverted scope?
If that is true then your drawing appears to show the water container with a non-flat bottom.

If that's the case, it will act as a lens when it has water in it, but not without. Adding more water in the tube could end up reducing the overall difference in OD between neighboring paths.
Most plates made for inverted scopes have flat bottoms to avoid this problem.

The water has a significantly different optical density from that of air, so the light bends.
This might be correctable by flooding the tube's surroundings with water. This might be doable by placing the tube in a water containing petri plate on the microscope stage (depending on how you are doing things). This should remove a lot of the difference in optical densities between paths near the edge, by replaceing air with water.
 
  • Informative
  • Like
Likes   Reactions: berkeman and sophiecentaur
  • #12
Apologies for the poor quality of the image, seems like upload to the forum downscaled it. I tried to increase the font to make it more readable:

Illumination.drawio (1).png


To summarise, yes this is in essence inverted microscope. The illumination is from the top, camera is located below and focusing on the bottom of the container. The bottom of the container is mostly flat, as much as plastic moulding allows. The lensing effect seems to be done by the top surface of the liquid. To be fair I am not even sure if that surface is working as concave or convex lens at this point.

Sadly I cannot immerse the container in any other liquid due to environmental constrains. I somehow need to illuminate the edges of the bottom of the container.
 
  • Like
Likes   Reactions: BillTre
  • #13
Gleb1964 said:
4f Koehler transmission illumination I think picture is illustrated 4f Köhler illumination described
Yes that is pretty much what I am using here for illumination and I can prove that the beam of light is mostly parallel when it comes out from the illumination system. Then the light seems to be bent when it goes from air to liquid inside the container.
 
  • #14
Much better picture.

Do you have to use water?
Maybe an oil of something instead, with less of a meniscus.

Is this some kind of biological prep?

Could you flood the tube and put a coverslip on top with no bubbles? This would make an optically good surface.
 
  • Like
Likes   Reactions: sophiecentaur
  • #15
I really wish it was that simple.
Liquid is water based with some medium for microbiology and cannot be changed.
Sadly it cannot be filled in to the top and covered either for handling.
The working volume is about 3/4 of the height of the container.

The solution has to be down to optics, corrective lens I am afraid.
Not sure how to calculate what is needed.
 
  • #16
HomeTinker said:
Liquid is water based with some medium for microbiology and cannot be changed.
Sadly it cannot be filled in to the top and covered either for handling.
The working volume is about 3/4 of the height of the container.

The solution has to be down to optics, corrective lens I am afraid.
Or post-processing of the image. Have you looked at Fourier transform options for post-processing?
 
  • #17
You might be able to do something like a flat-field correction. It is commonly used in photography and astro-photography to correct for uneven illumination/brightness. Do you have any image processing software? Send me a PM and I'll try to figure out something with you.
 
  • Like
Likes   Reactions: berkeman
  • #18
HomeTinker said:
Sadly it cannot be filled in to the top and covered either for handling.
It may be that there's not a 'simple' solution. A major change in design may be needed.

A suitable correcting lens would be difficult to make. Hard enough to correct if the air / water interface were spherical but very hard for this situation with a very non-spherical shape to correct..
Use a shorter tube?
Use a wider tube with a ring to obscure the curved part of the surface?
Cover the water with a thin layer of oil?

Other workers are sure to come across this problem. Some more reading round, perhaps.
 
  • #19
Many thanks for the answers, I assume post processing may help to certain degree but I think there must also be some additional update to the optics.
I am looking to use something like this:
https://en.wikipedia.org/wiki/Meniscus_corrector

But how do I calculate the curvature I need?
 
  • #20
HomeTinker said:
I am looking to use something like this:
https://en.wikipedia.org/wiki/Meniscus_corrector
A meniscus corrector corrects for spherical aberration, not uneven illumination. It acts as a slightly negative lens and will possibly make the problem worse. Though I am not an optical engineer, so I can't say for certain.
 
  • Like
Likes   Reactions: vanhees71
  • #21
While been an optical engineer, there is not a good description of the problem to see the exact "trade of" of the optical design.
The perihelial of field seems to be vignetted close to the tube walls, the involved factors are meniscus shape of the liquid surface, the depth of the liquid, the inner diameter of tube and the length of the tube walls, preventing the illumination light cone to be accepted by the microscope lens.
The meniscus lens can change the cross section of illumination beam, it does produce the non-zero angular magnification, the spherical aberration is a secondary and not significant factor there, but I don't see directly where it should be placed.
 
  • #22
Drakkith said:
A meniscus corrector corrects for spherical aberration,
Afaics, the name "meniscus" is only used because the lens 'looks a bit like one'. It's designed for spherical lenses, which the meniscus on top of the liquid is not.

Just a thought: the meniscus could be corrected a bit by altering the inner surface of the tube with a suitable paint / varnish or even some soap.
 
  • #23
Think about reverting the optical layout. You have an imaging system, collecting light, so the illumination should match your collection system. If you want even background illumination, the illumination system is supposed to fill the acceptance cone of the imaging lens at every point of the illuminated field. I can assume that the imaging system is telecentric at the object space (where you have a sample). That means that the aperture stop is located at infinity or far enough, to consider that acceptance light cones are parallel at the sample field. Propagating the acceptance light cones backwards, they should hit the light source. But the light cones at the margin of the field would find the walls of the cuvette, where the part of the cone would expire internal reflection, changing its direction somewhere else, missing the source.
 
  • #24
Drakkith said:
It is commonly used in photography and astro-photography to correct for uneven illumination/brightness.
It's common to use 'greys'. These are shots of a flat grey background which are then subtracted / added to (or is it multiplied by?) the wanted image to give an aesthetically pleasing look.
 
  • #25
Many thanks for some good replies. To put things into context please see that image below taken from the side of the container showing how the liquid bends inside.

Untitled.png

The illumination is from the top and imaging is below.
The system is telecentric, the center of illumination is at the center of the container, imaging lens and the camera.

And here are example images with 200 ul (showing the issue) and 400ul of liquid which is the container filled up to the top.
200ul.png
400ul.png


All parameters of the system are this same other then the level of liquid in the container.

Question to Gleb1964:
Are you saying that essentially I see the "shadow" of the container walls and there is no way I can get the light into this areas? Apologies if I misunderstood your comments.

Many thanks for help.
 
  • #26
Interesting. It looks to me like the meniscus is, essentially, a negative lens. The surface forms small angles near the center and larger angles as we move outwards, meaning the light is entering close to normal and hardly refracting at all near the center but being very sharply refracted near the edges. That should explain why the illumination is so uneven. The 400 ml sample has a much shallower meniscus and is roughly twice as far away from the bottom, which would help even out the illumination.

I'd guess that placing a positive lens somewhere before the meniscus-surface would counteract the effect somewhat. I'm unsure if it can be eliminated entirely.

I also wonder if a diffuse light source instead of a collimated source would help. Though that might affect the quality of the dust particles in the image.

Again, just my non-professional thoughts.
 
  • Like
Likes   Reactions: Gleb1964
  • #27
That is what I am thinking, there should be a way to minimise the effect optically (I am sure completely removing it may be difficult). The question is, how to calculate the lens needed.

I also tried the version with just the LED and the diffusor before making it Koehler and the issue was still present (I would say even more prominent).

But here is what I cannot understand.
If the liquid surface is bending that way, the light rays from the illumination system should be bent outwards towards the sides of the container when they cross from air to water via the liquid surface. If anything that should make the opposite effect than what is on the pictures.

Any ideas what am I missing here?
 
  • #28
HomeTinker said:
If the liquid surface is bending that way, the light rays from the illumination system should be bent outwards towards the sides of the container when they cross from air to water via the liquid surface. If anything that should make the opposite effect than what is on the pictures.
The light near the center is less strongly refracted than light entering the water on the outer band of the meniscus. The center 10% is barely refracted at all, and so is barely dimmed. But the outer 10% is refracted entirely out of the container through the glass (I don't know the exact percentage, that's just an example). The light that actually makes it to the outer 10% of the sample came from somewhere in the middle band of the meniscus, not from the outer radius.

Also note that the center of the meniscus is much closer to the bottom of the sample than the outer portions. So light entering from the center of the meniscus has less distance in which to diverge than light entering from the outer bands of the meniscus. It simply can't spread out as much before hitting the bottom of the sample.
 
  • #29
What if you place a plano-covex lens at the bottom of container, flat side towards bottom and observe the cuvette through the convex side of the lens? The aim of such lens is to form sort of Galilean telescopic system in combination with the negative water meniscus. In that case the illumination system woud function as it should.
 
  • Informative
Likes   Reactions: berkeman
  • #30
Surely if I did that the image of the elements on the bottom of the container that I am trying to take will be distorted, right?

By the way - is there any software that could be used to model this?
 
Last edited:

Similar threads

  • · Replies 24 ·
Replies
24
Views
3K
  • · Replies 12 ·
Replies
12
Views
2K
  • · Replies 2 ·
Replies
2
Views
2K
  • · Replies 19 ·
Replies
19
Views
5K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 4 ·
Replies
4
Views
6K
  • · Replies 7 ·
Replies
7
Views
2K
  • · Replies 6 ·
Replies
6
Views
3K
  • · Replies 0 ·
Replies
0
Views
724