Interpretation of Linear Variable Differential Transformer (LVDT) voltage output

Click For Summary
SUMMARY

The discussion centers on the interpretation of voltage output from a Solartron DC50 Linear Variable Differential Transformer (LVDT) with a sensitivity of 6.158 mV/V/mm at 10 VDC. The user observed a polynomial relationship rather than a linear one when plotting LVDT voltage output against height measurements, leading to discrepancies in expected values. The consensus among forum members suggests that the user can rely on the fitted polynomial curve for accuracy, given that the measured accuracy of 2.5% is acceptable for their application, negating the need for recalibration.

PREREQUISITES
  • Understanding of Linear Variable Differential Transformers (LVDTs)
  • Familiarity with data acquisition systems, specifically dataTaker
  • Knowledge of polynomial curve fitting techniques
  • Basic principles of voltage output measurement and calibration
NEXT STEPS
  • Research polynomial regression analysis techniques for better data interpretation
  • Explore advanced calibration methods for LVDTs
  • Investigate the effects of power supply variations on LVDT performance
  • Learn about dataTaker configuration for optimal data collection
USEFUL FOR

Engineers, physicists, and technicians involved in precision measurement and data acquisition using LVDTs, particularly those interested in optimizing measurement accuracy and understanding voltage output characteristics.

ca2n
Messages
10
Reaction score
0
Good day to all Physics Forums members,

I am currently undertaking a project which involves the use of a Linear Variable Differential Transformer (LVDT) to measure the distance an object makes between two points. My setup is as follows:

  1. LVDT Type: Solartron DC50 with sensitivity of 6.158 mV/V/mm at 10 VDC
  2. Power supply to LVDT: 12.12 VDC (const. and cannot be varied)
  3. Voltage output from LVDT fed to dataTaker data acquisition system

In an attempt at verifying the calibrated sensitivity of the LVDT, I performed a number of height measurements using gauge blocks. I took 6 equally-spaced height intervals between the maximums of the LVDT range and plotted the LVDT voltage output vs. height. Here is what I encountered:

  1. The curve plotted was NOT linear as I anticipated it to be. As a matter of fact, the curve fit a polynomial equation of degree five with R2 = 1
  2. When comparing the LVDT output voltage with the LVDT calibrated sensitivity values, discrepancies occurred, e.g. a 20 mm difference between two gauge block heights returned a 20.5 mm difference calculated from the voltage output of the LVDT. The discrepancies grew as the ends of the LVDT were approached.

My questions are:

  1. Could the fitted curve be beneficial for my use seeing as I've determined the end-to-end output values of the LVDT, OR,
  2. Should I, instead, just get the LVDT calibrated and use the obtained sensitivity value.

To be frank, I would highly prefer Option 1, if it is in fact a justifiable solution.

Any help is much appreciated. Thanks. :smile:
 
Engineering news on Phys.org
From your quoted numbers, the accuracy is 2.5% of reading. Considering that your supply voltage is over 20% high, that isn't bad at all. If 2.5% is within your needed accuracy, use it as is.

I see no reason to "...get the LVDT calibrated...", you just did when you measured the gauge blocks. Use those numbers if you need the accuracy.

Cheers,
Tom
 

Similar threads

  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 4 ·
Replies
4
Views
3K
Replies
1
Views
9K
  • · Replies 2 ·
Replies
2
Views
2K
  • · Replies 4 ·
Replies
4
Views
3K
Replies
7
Views
3K
Replies
4
Views
9K
  • · Replies 1 ·
Replies
1
Views
3K