Discussion Overview
The discussion revolves around the feasibility of detecting the energy of a single ion traveling at approximately 10 keV in real time without altering its trajectory. Participants explore various methods, including the use of wire chambers, Doppler shift measurements, Penning traps, and avalanche counters, while considering the challenges associated with signal strength and noise levels.
Discussion Character
- Exploratory
- Technical explanation
- Debate/contested
- Mathematical reasoning
Main Points Raised
- Some participants propose measuring the energy of a passing ion in situ, while others suggest that time-of-flight (ToF) is the standard method for such measurements.
- One participant estimates that a wire at 500nm from the ion would induce a 100 microvolt EMF, which may be too weak to detect in a single pass.
- There is a suggestion to use a Doppler shift measurement with an intense probe beam, although concerns are raised about the time available for photon absorption.
- Participants discuss the potential of using a Penning trap for measuring ion energy, noting that while it allows for measurement, it may alter the ion's trajectory.
- Another participant mentions the PS200 experiment, which successfully captured antiprotons in a Penning trap, but questions whether this method aligns with the original intent of non-destructive measurement.
- Concerns are raised about the signal-to-noise ratio and the need for a low thermal noise detection system.
- One participant calculates the induced current in a Penning trap setup and suggests that it could be measured in real time, but questions the assumptions made in this approach.
- Another participant challenges the claim of measuring currents as low as 10-15 A, emphasizing the importance of integration time in such measurements.
- There is confusion regarding the implications of turning off the magnetic field in a Penning trap and its effect on ion motion.
Areas of Agreement / Disagreement
Participants express a range of views on the feasibility of different measurement techniques, with no consensus reached on the best approach or the implications of various methods. The discussion remains unresolved regarding the optimal strategy for real-time energy detection of a single ion.
Contextual Notes
Participants highlight limitations related to signal strength, noise levels, and the impact of measurement methods on ion trajectories. The discussion includes various assumptions about the conditions under which measurements could be made.