Understanding 1/f Noise in DC Voltmeters: Causes and Solutions

  • Thread starter Thread starter fluidistic
  • Start date Start date
  • Tags Tags
    Dc Noise
Click For Summary

Discussion Overview

The discussion revolves around the challenges and considerations of measuring very low voltages (around 1 nV) with DC voltmeters, particularly in the context of 1/f noise. Participants explore the implications of noise types, measurement techniques, and the physical principles involved in achieving accurate readings in low-voltage scenarios.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested

Main Points Raised

  • One participant notes that 1/f noise becomes problematic at very low voltage measurements, such as 1 nV, while stable readings can be obtained at higher voltages, like 1 V.
  • Another participant requests clarification on the concepts of pink noise, 1/f noise, and thermal noise, and inquires about the source impedance of the voltage being measured.
  • A participant mentions the limitations of their equipment, stating they cannot measure in the 1 nV range, but reference a voltmeter with a peak-to-peak DC noise of 6 nV.
  • Questions are raised about the importance of bandwidth in assessing noise performance, suggesting that it is a critical factor in measurements.
  • One participant proposes a hypothetical scenario involving a material with a resistance of about 1 ohm and a 4-point probe method to measure voltage differences when using a DC current of 1 A, indicating that 1/f noise complicates the use of DC voltmeters.
  • Another participant emphasizes that noise power is dependent on system bandwidth and discusses the implications of operating near absolute zero temperature, mentioning thermal noise power and the noise figure of amplification stages.

Areas of Agreement / Disagreement

Participants express varying levels of understanding and concern regarding the impact of 1/f noise on low-voltage measurements. There is no consensus on the best approach to mitigate this noise or the specific conditions under which it becomes problematic.

Contextual Notes

Participants acknowledge the complexity of noise types and measurement conditions, including the dependence on bandwidth and source impedance, but do not resolve these issues or provide definitive solutions.

fluidistic
Gold Member
Messages
3,934
Reaction score
286
I've been told that the 1/f noise for DC voltmeters is a problem when they measure a very low voltage (i.e. around 1 nV). Some readings can take weeks or even months. However, a DC voltmeter would have no problem to return a stable reading of a simple resistive circuit, of the order of the volt.
Physically I do not understand what's going on and why there's a difference between the 1 nV and the 1 V case regarding the pink noise. I'd like to see equations that would tell me whether the 1/f noise would be a problem, for a DC reading.

Note that DC corresponds to f = 0 Hz, so infinite noise in principle.
 
Engineering news on Phys.org
Could you say more about Pink noise, 1/f noise and thermal noise in the context of your measurement? What is the source impedance of the voltage that you are wanting to measure so accurately?

EDIT/ADD -- BTW, I don't have anything in my lab that can measure in the 1nV range. I do have a pA meter that we use for specialized measurements...
 
  • Like
  • Informative
Likes   Reactions: sophiecentaur and dlgoff
berkeman said:
BTW, I don't have anything in my lab that can measure in the 1nV range.
I just did a google search and found this really good voltmeter but if I understand the specs correctly, the best it can do is 6nV. i.e. the peak to peak dc noise. Here's the pdf for the Keithley 2182A.
 
berkeman said:
What is the source impedance of the voltage that you are wanting to measure so accurately?
Also, what bandwidth is involved? Do you not need to know that in order to asses the noise performance of a device?
 
I am sorry, I am not measuring anything in particular, I ask only to know more.
Let's forget about thermal noise by considering a material near 0 K. I did not want to involve that noise at all.
Let's say the material has a resistance of about 1 ohm and we perform a 4-points probe method on it, to read out its voltage when we use a DC of 1 A. By reversing the current (but still a DC), we should in principle be able to measure a different voltage. The difference in the voltages when the current passes in 1 direction compared to the other might be in the nV range due to some physical effects I do not want to go into here. Apparently using a DC voltmeter (the Keithley is a good example I think) is problematic due to the 1/f noise.
On the other hand, using a "high" (not too high though) frequency AC is better suited to get that small nV voltage difference, which can be done in seconds.
I have no idea about bandwidth. The bandwidth of what exactly?
 
Like the Poor and Taxes, Noise is always with us and the noise power is dependent on the system bandwidth. To discuss operating near 0K, every item in the circuit (before the first stage of amplification) has to be that cold.Thermal noise power is kTB where B is the measurement bandwidth. You have to add to that the Noise Figure of your amplifying stage. That's always a consideration in communications (receivers) but it applies everywhere. This link can give you some information.
 

Similar threads

  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 15 ·
Replies
15
Views
5K
Replies
6
Views
4K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 6 ·
Replies
6
Views
2K
  • · Replies 18 ·
Replies
18
Views
8K
Replies
9
Views
4K
  • · Replies 4 ·
Replies
4
Views
3K
  • · Replies 1 ·
Replies
1
Views
4K
  • · Replies 21 ·
Replies
21
Views
13K