Valid PointWhy Should Zero Point Be Excluded from Calibration Curves?

  • Thread starter Thread starter ShawnD
  • Start date Start date
  • Tags Tags
    Point Zero
Click For Summary
Including the zero point on calibration curves is often debated, with arguments against it focusing on issues of precision and linearity. Zero is typically an outlier unless the instrument is specifically calibrated to recognize it, leading to misleading conclusions about measurement accuracy. Calibration curves should reflect the actual data without artificially assuming a valid zero point, as this can distort the y-intercept and slope, increasing errors at both low and high concentrations. The inclusion of zero can misrepresent the dynamic range and sensitivity of the instrument, making it irrelevant for accurate analysis. Overall, excluding the zero point is recommended to maintain the integrity of calibration data.
ShawnD
Science Advisor
Messages
715
Reaction score
2
I keep getting into arguments with my teachers about including zero as a point on my calibration curves. Their reasoning is that it should theoretically be valid since you won't get a signal when you do nothing (sound reasoning). My argument is that if you plot it, it is usually an outlier, unless you specifically tell your instrument "this is zero", and that only seems to apply to spectroscopy instruments.

Precision
When you do a calibration curve, you are stating "I can accurately determine values that fall within this range", which is why you need to adjust your sample concentrations until their readings fall within your calibration curve. If I include a zero point, I'm saying I can solve for values that are between zero and whatever the next point is. If I have points on my graph for 0 and 1ppm, I am basically stating that my machine is capable of accurate readings to infinite orders of magnitude between 0 and 1. You don't have this same false precision between points like 1 and 2 because 1.000001, 1.1, and 2 are on the same order of magnitude.

Linearity
Reactions do not follow a linear trend through all points. This is especially true at very high and very low concentrations. Most reactions curve off or follow a lazy S shape. Curve-off at low concentrations is generally caused by a lack of instrument sensitivity. Curve-off at high concentrations is either caused by self-absorbance in spectroscopy, or exceeding the linear range of the detector itself for whatever the instrument may be. Lazy-S curves are usually due to logarithmic relations to concentration; cell potential for example. Nobody includes the high concentration curve-off points as part of their linear calibration curve, so why would anybody include the zero point which has all these low concentration linearity problems?

Why are we doing multipoint?
Multipoint calibration is done to confirm linearity as well as correct for y-intercepts. If you assume the point at 0,0 is valid, you've completely missed the boat on why you are doing this. Did you follow the data to see if it should intersect at 0,0? No, you lead the data by saying it's a valid point. That is not science. You do not lead the data like that, ever.

Invalid check standards
From what I've seen, the y-intercept for response vs concentration is usually positive. When the zero point is part of the calibration, it typically decreases the y-intercept and increases the slope (think of rotating the best fit line counterclockwise). This keeps the best fit line in a similar position for the middle part of the calibration, but it greatly increases the errors at low and high ends of the calibration curve. Since your check standard is usually somewhere near the middle calibration point, your check standard will be ok, but whatever you are analyzing may fall closer to the ends of the calibration, and the resulting error is enormous for seemingly no reasaon. I've had this happen to me many times before. Each time I go back, remove the zero point, and redo the calculations, both the check standard and sample being analyzed are closer to what they should be.



In conclusion, including the point 0,0 should almot never be done.
 
Last edited:
Chemistry news on Phys.org
'Tain't clear what your question is: sensitivity, dynamic range, absolute uncertainty, percentage uncertainty, and response function look to have been all mixed together --- not necessarily by you, but somewhere in your education.

0,0? Sure. All the time in thermocouple calibrations; one variable isn't strictly 0. Chemistry and "0,0"? It's a logarithmic universe, so you've got zero log values showing up all the time, but no zeroes. Physical chemistry and 0,0? "log of this vs. one over that?" Same game.

Help?
 
Seems to me that unless the device can be actually calibrated to return a value of zero when you zero it (some do, some don't), you shouldn't use it as a calibration point because, like you said, it gives a false precision.

But it would depend on the device.
 
From the labs that I've done so far, I haven't included zero in the calibration curves, what you're concerned is with the dynamic range of the data and as long as you have subtracted off a background, the zero is somewhat irrelevant, unless the curve is a special plot where the intercept is supposed to represent a parameter. In fact, with beer's law, I think you get some deviation from linearity at the extreme ends of the plots, better or worse depending on the instrumentation, techniques, and the conditions employed, it's important to realize that the points near the origin may not even be realistic, and "off the mark", irrelevant.
 
ShawnD said:
In conclusion, including the point 0,0 should almot never be done.
Haven't read the entire post, but I agree with that last bit based on definition : The calibration curve is (the suitably smoothed) the two-column data generated during an experimental run.

Extrapolation is a crime punishable by death (or grad school) ! :biggrin:
 
I came.across a headline and read some of the article, so I was curious. Scientists discover that gold is a 'reactive metal' by accidentally creating a new material in the lab https://www.earth.com/news/discovery-that-gold-is-reactive-metal-by-creating-gold-hydride-in-lab-experiment/ From SLAC - A SLAC team unexpectedly formed gold hydride in an experiment that could pave the way for studying materials under extreme conditions like those found inside certain planets and stars undergoing...

Similar threads

  • · Replies 6 ·
Replies
6
Views
7K
Replies
1
Views
815
  • · Replies 4 ·
Replies
4
Views
4K
  • · Replies 3 ·
Replies
3
Views
3K
  • · Replies 6 ·
Replies
6
Views
5K
  • · Replies 27 ·
Replies
27
Views
7K
  • · Replies 5 ·
Replies
5
Views
2K
  • · Replies 62 ·
3
Replies
62
Views
10K
  • · Replies 51 ·
2
Replies
51
Views
5K
  • · Replies 40 ·
2
Replies
40
Views
5K