Why Do Shelves 3 and 4 in My Plasma Oven Heat Up More Than Others?

Click For Summary
SUMMARY

The discussion centers on temperature discrepancies observed in shelves 3 and 4 of an Advance Plasma Systems industrial plasma oven, which operates at 1600 watts RF for 8 minutes. Temperature mapping revealed that these shelves increased from 20°C to 45°C, while others only rose by 3°C. A slight leak in the cooling water pipes at these levels is suspected to contribute to the temperature increase. The user intends to maintain the current setup for further trials before addressing the leak.

PREREQUISITES
  • Understanding of RF heating principles in plasma ovens
  • Familiarity with vacuum systems and pressure measurements (450 mTorr)
  • Knowledge of temperature mapping techniques
  • Basic understanding of fluid dynamics related to cooling systems
NEXT STEPS
  • Investigate the effects of cooling water leaks on temperature stability in plasma ovens
  • Learn about RF heating efficiency and its impact on material processing
  • Research methods for effective temperature mapping in industrial ovens
  • Explore troubleshooting techniques for vacuum systems in plasma applications
USEFUL FOR

Engineers and technicians working with plasma ovens, industrial equipment operators, and researchers studying thermal dynamics in vacuum systems will benefit from this discussion.

martine5
Messages
1
Reaction score
0
Ok small bit of a problem which I have been experiencing.
I am using an industrial size plasma oven by advance plasma sytems. This plasma oven operates at an RF of 1600 watts for approx 8 minutes for the duration of the plasma cycle.
It pulls down to a vacuum of 450mTorr and use oygen as a medium for firing the plasma.
There are 6 shelves within the plasma oven.
After temperature mapping the mapping the oven tehre was 1 significant find, shelves 3 & 4 once the RF kicked in the temperature rose from 20C to roughly 45C, on the other shelve the temp increase was only about 3C,(10 to 23 )
What I am trying to figure out is what would cause the increase on these 2 shelves? On further inspection there has been a slight leak noted upon the pipes for the cooling water on these 2 shelf level. As the shelves are paired in a +/- config between shelving units is it a case taht the water is vaporising under for all intensive purposes what could be classified as a farraday cage? Is this water leak the most likely cause of the temp increase.
Common logic dictates that eliminating this leak and remapping the temperature is the way to go, but at present I would prefer to keep in the current state to run a body of trials and get greater understanding around this temperature increase.
Any help/suggestions with would be greatly appreciated.
Thanks in Advance
EM
 