  • #1
Salman Khan
19
1
If pulse source output is recorded by a detector on oscilloscope, is it possible to calculate dose from this output voltage pulse?
 
Engineering news on Phys.org
  • #2
That depends on the detector and any amplifier. For an ion chamber it would be the total charge transferred, which would be the area under the curve on an oscilloscope trace under some circumstances, or the height of the pulse in others.

A Geiger counter tube would not work. Some other forms of detector might work, but interpreting the result might be challenging.
 
  • #3
Alex A said:
That depends on the detector and any amplifier. For an ion chamber it would be the total charge transferred, which would be the area under the curve on an oscilloscope trace under some circumstances, or the height of the pulse in others.

A Geiger counter tube would not work. Some other forms of detector might work, but interpreting the result might be challenging.
In case of PMT or other solid state detector?
 
  • #4
A commercial PMT/Scintillator would probably fail to work. Counters are usually designed with high value resistor chains to save power and with small value capacitors close to the anode. These saturate easily. Spectroscopy grade combinations would probably damage themselves at fairly low exposures.

If you are measuring the dose within the beam I would expect a strong pulse of light from the scintillator. Much more than would be a good fit for a PMT.

A reverse biased diode would probably give a signal, but at these power levels there may be a lot of nonlinear effects. A scintillator connected to a photodiode by a light pipe or fiber optic cable might work well.

I'm bothered that you seem to be trying to reinvent the wheel for instrumentation for a very dangerous device.
 
  • Like
Likes Salman Khan
  • #5
Dear Alex, very thanks for sharing your experience.
 
  • #6
Why use an oscilloscope? An ion chamber and a DVM works and is standard for dose measurement of x/gamma radiation.
 
  • Like
Likes Salman Khan
  • #7
So far as I know ion chamber is gaseous detector having response tim in micro second, how is it possible to measure a pulse dose of nano second pulse using ion chamber ??
 
  • #8
You didn't say nanoseconds. It works for microseconds for sure. The problem with nsec. pulses is ion recombination. A solid-state detector can be used, I think, instead of an IC, but can be calibrated with a longer pulse width beam relative to an IC.

Just found this for ICs https://www.drct.com/Ion_Chambers/TBM-IC-Pulse-X.html
 
  • Like
Likes Salman Khan

1. How do you calculate the voltage and time scale on an oscilloscope?

The voltage scale on an oscilloscope is typically measured in volts per division. To calculate the voltage scale, divide the total range of the vertical axis (in volts) by the number of divisions on the screen. For example, if the vertical range is 10 volts and there are 5 divisions on the screen, the voltage scale would be 2 volts per division. The time scale is typically measured in seconds per division. To calculate the time scale, divide the total time of the horizontal axis (in seconds) by the number of divisions on the screen.

2. How do you determine the sampling rate of an oscilloscope?

The sampling rate of an oscilloscope refers to how many times the waveform is sampled per second. To determine the sampling rate, divide the total time of the horizontal axis (in seconds) by the number of samples displayed on the screen. For example, if the horizontal axis is 10 milliseconds and there are 100 samples displayed on the screen, the sampling rate would be 10,000 samples per second.

3. What is the difference between real-time and equivalent-time sampling on an oscilloscope?

Real-time sampling is when the oscilloscope captures and displays the actual waveform in real-time. This is useful for observing fast-changing signals. Equivalent-time sampling is when the oscilloscope takes multiple samples of the same signal over a longer period of time and then combines them to display the waveform. This is useful for capturing and analyzing signals that occur less frequently.

4. How do you calculate the rise time of a signal on an oscilloscope?

The rise time of a signal is the time it takes for the signal to rise from 10% to 90% of its maximum amplitude. To calculate the rise time, measure the time it takes for the signal to rise from 10% to 90% on the oscilloscope's time scale and multiply it by the time scale setting. For example, if it takes 2 milliseconds for the signal to rise from 10% to 90% and the time scale is set to 0.5 milliseconds per division, the rise time would be 4 milliseconds.

5. How do you adjust the trigger level on an oscilloscope?

The trigger level on an oscilloscope is used to stabilize the waveform on the screen. To adjust the trigger level, first select the channel you want to trigger on. Then, adjust the trigger level knob until the waveform is stable on the screen. You can also adjust the trigger level by using the up and down arrows on the trigger menu and selecting a specific voltage level.

Similar threads

  • Nuclear Engineering
Replies
1
Views
1K
  • Nuclear Engineering
Replies
9
Views
1K
  • Nuclear Engineering
Replies
4
Views
1K
  • Nuclear Engineering
Replies
1
Views
1K
  • Nuclear Engineering
Replies
25
Views
4K
  • Nuclear Engineering
Replies
2
Views
2K
  • Nuclear Engineering
Replies
1
Views
1K
  • Nuclear Engineering
Replies
5
Views
404
  • Nuclear Engineering
Replies
2
Views
2K
  • Nuclear Engineering
Replies
2
Views
1K
Back
Top