Alright so I have been given a table for a transducer as follows:
Operation
Input range 0–1000 cm H2O
Excitation 15 V DC
Output range 0–5 V
Performance
Linearity error 0.5% FSO
Hysteresis error Less than 0.15% FSO
Sensitivity error 0.25% of reading
Thermal sensitivity error 0.02%/C of reading
Thermal zero drift 0.02%/C FSO
Temperature range 0–50 C
I was told that my Nomianl pressure is 500cm H20 and ambient temp is betwenn 18 and 25 degrees C.
I'm being asked to calculate the magnitude of each elemental error.
I'm not asking for all of the answers of course but maybe just walk me through let's say Hysteresis, I have no idea how to use the table with the given values to find the elemental error...