Can surrounding temperature affect IR camera results analysis?

Click For Summary
SUMMARY

The discussion centers on the impact of surrounding temperature on infrared (IR) camera results when measuring the emissivity of steel. The user, fictionx, is utilizing an energy balance equation, σ(Tc)^4 = εσ(Ta)^4, to calculate emissivity, but questions whether ambient temperature should be factored into the analysis. A participant from the forum, who employs a FLIR camera, confirms that ambient temperature, incident angle, distance, and air currents significantly affect IR measurements. Their testing procedure has been adapted to control these variables for accurate results.

PREREQUISITES
  • Understanding of emissivity and blackbody radiation principles
  • Familiarity with infrared camera technology, specifically FLIR cameras
  • Knowledge of energy balance equations in thermal analysis
  • Experience with thermocouple calibration methods
NEXT STEPS
  • Research the effects of ambient temperature on IR camera measurements
  • Learn about the calibration process for infrared cameras
  • Investigate methods to control environmental variables in thermal testing
  • Explore the use of thermocouples for accurate temperature correlation with IR readings
USEFUL FOR

This discussion is beneficial for thermal engineers, materials scientists, and professionals using infrared cameras for temperature measurement and emissivity analysis.

fictionx
Messages
4
Reaction score
0
Hello all,

I'm currently using an infrared camera to determine the emissivity of a steel. The camera "sees" whatever radiation entering the objective, and calculates the temperature of the steel by assuming it's emissivity is 1. I figured I could use an energy balance to find the emissivity:

σ(Tc)^4 = εσ(Ta)^4

where Tc is the temperature that the camera says and Ta is the actual temperature. Because the camera assumes blackbody behavior, I do not include the ε on the left hand side.

My question is, do I need to somehow incorporate the temperature of the surroundings in my equation? I wasn't sure if I am supposed to add it to the right hand side and use an emissivity of 1. It changes the outcome drastically.

Thanks,
fictionx
 
Engineering news on Phys.org
We regularly use a FLIR camera for testing at my company. The emissivity value for the test product was determined experimentally. Multiple thermocouples were placed on the actual part and the emissivity value of the IR camera was "calibrated" until acceptable results were obtained that correlated to the thermocouple values.

We also discovered that the IR camera was very sensitive to
  • incident angle of camera-to-part
  • distance of camera-to-part
  • ambient temperature
  • ambient air currents
  • elevation of camera-to-part
  • probably some other things I can't remember

Our testing procedure now incorporates setups to control all of those various sensitivities.
 
That's pretty intense thank you for this information. Eventually I'm going to be trying to use our camera to determine the temperature of a moving granulate salt in a screw heat exchanger which has IR optical glass on top. Another problem I'm running into is that when the glass gets warmer it also emits and it screws up the results.
 

Similar threads

  • · Replies 152 ·
6
Replies
152
Views
12K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 14 ·
Replies
14
Views
5K
  • · Replies 3 ·
Replies
3
Views
8K
  • · Replies 1 ·
Replies
1
Views
13K
Replies
1
Views
4K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 23 ·
Replies
23
Views
37K
  • · Replies 11 ·
Replies
11
Views
5K
  • · Replies 1 ·
Replies
1
Views
3K