Design a system to measure the electroretinogram (ERG)

8 replies · 2K views · last reply
benderlee
Messages
5
Reaction score
1
Homework Statement
Design a system to measure the electroretinogram (ERG). Use a differential amplifier to maximize the CMRR, and condition the signal appropriately for a 16-bit A/D system with input range ±10 Volts. Give all component values.

Suggest an appropriate sampling rate, assuming that all noise has been sufficiently attenuated by the time the signal reaches the digitizer, and give the resolution of the signal in V/bit at the digitizer.


ERG bandwidth= 0-300 Hz

ERG dynamic range = ± 1 mV
Relevant Equations
(See attached image)
243239
 
Welcome to the PF. :smile:

Thanks for posting some of the relevant equations. Now please show us your work on the solution. Also, what do the sensors for an ERG look like, and where are they placed?
 
Last edited:
  • Like
Likes   Reactions: benderlee
benderlee said:
I am truly stumped by this one, looking for more information online myself and was seeing if anyone in this great community knows anything about this, so far I found this:

https://webvision.med.utah.edu/book/electrophysiology/the-electroretinogram-erg/
That's a helpful link. So it looks like one ERG electrode is placed on the cornea, and the other is attached to the patient's body for reference.

So start with the desired sampling rate (how is it based on the signal bandwidth?), and design the differential amplifier circuit and other gain elements and filter elements next. Are you familiar with designing digitizing systems, including anti-alias filters? That is all part of this problem you are working on.

You can also probably do some reading about ECG and EEG measurement systems, since they will share a lot of the same design components (they will have different bandwidth and pickup voltages, but the rest is pretty similar).
 
  • Like
Likes   Reactions: benderlee
So this is how the actual circuit would be set up 😊
 
Attachments
  • Cross-section diagram of the eye showing a contact lens electrode on the cornea, connected to a differential amplifier with a reference electrode and ground, outputting ERG.
    erg.JPG
    33.2 KB · Views: 446
benderlee said:
So this is how the actual circuit would be set up 😊
Did you draw that, or find it somewhere? A differential amp has two signal inputs, not three.
 
  • Like
Likes   Reactions: benderlee
This is a diagram from a textbook - attached is another diagram from another text. that is more or less the same with a ground, reference, and corneal electrode:
 
Attachments
  • Textbook page with two paragraphs on the electroretinogram above, then Figure 1.2: block diagram of an ERG recording setup with a light stimulator, differential amplifier, transparent corneal electrode, reference and ground, beside a cross-section of the eye.
    erg2.JPG
    45.8 KB · Views: 463
Also FYI: in the ERG, the differential amp records the difference between the active electrode (attached to the cornea) and the reference electrode (attached to the outer canthus). The ground electrode is attached to the forehead.
 
benderlee said:
This is a diagram from a textbook
That's not a very accurate or informative diagram, unfortunately. Hopefully you are learning about differential amplifier design for biomedical applications from a better and more detailed textbook or application notes.
benderlee said:
Also FYI: in the ERG, the differential amp records the difference between the active electrode (attached to the cornea) and the reference electrode (attached to the outer canthus). The ground electrode is attached to the forehead.
So the diff amp inputs should come from closely spaced electrodes at the eye, and the "ground"/reference connection comes from some other large part of the body nearby. The "ground"/reference connection goes into the overall amplifier circuit, not into the diff amp itself.

So show us what you have figured out so far. What resistor values have you chosen for your diffamp and why? What filter topology and component values have you chosen for your anti-alias filter and why?

What sample rate have you chosen for your ADC and why?
 
  • Like