What is the best way to analyze the XRD data for phase identification and particle size calculation? so far I am using debye-scherrer's formula. Is there any advanced formula?

thankyou
 
  • Like
Likes Adashi
Physics news on Phys.org
Can you please provide more complete details?

What system are you studying?What phases do you expect? How many phases (down to what trace levels) do you wish to identify and to what accuracy? What instrumentation (just a DS camera, or maybe a diffractometer) are you using/do you have at your disposal?

What software is avaliable at your XRD lab? Do you have the PDF database on a computer? Do you also have any search/match software? Do you have a Rietveld Analysis (the only way I know to determine particle size) program ?
 
Thankyou gokul for your willingness...!

I use PANalytical XRay Diffractometer. I am comfortable with phase identification. But I wanted to know a way to calculate particle size. Yes we have PDF database and Rietveld Analysis program. So far I used to determine particle size using debye-scherrer formula

d = 0.9*lambda/delta*cos(theta) ,

where
lambda --- wavelength of x-rays
delta-FWHM of diffraction peak
theta--- angle corresponding to the peak
 
The one drawback of the above simple method is that it works only if stress-related and instrument-related broadening are negligible in comparison to particle size effects. This condition is often met with particle sizes that are in the 10 - 100 nm range. At um particle sizes, you will have to be careful to separate broadening effects from particle size (~ 1/cos \theta), strains fields (~ tan \theta) and all instrumental effects (~ some constant additive number). By plotting FWHM vs angle, and fitting the peak width vs Bragg angle curve to parameters A,B and C in :

FWHM^2 = \left( \frac{A}{cos \theta} \right) ^2 + (Btan \theta)^2 + C^2

From the best fit, you can extract A and calculate the particle size from the DS formula above.

PANalytical itself makes pretty good particle size analysis software (it's part of the software package they try to sell with their diffractometers) and it may just be that your XRD lab already has this software.
 
Hi,

I am using the Scherrer's formula for calculating the grain size of powders, I was not getting proper results. Is there any way that i can know what is going wrong in my calculations.

Thank you
priya

anil.bose said:
Hi friends,

What is the best way to analyze the XRD data for phase identification and particle size calculation? so far I am using debye-scherrer's formula. Is there any advanced formula?

thankyou
 
We might be able to help if you can give us more details:
1. What is the form of your data (x, y, ranges)?
2. How exactly did you do the calculation?
3. Did you ignore stress-related broadening?
4. Is the powder analyzed directly after milling, or is it annealed before it is measured?
 
I am using Iron and nickel powder. The X axis (2- Theta) range for both of them is (40-100 degrees). The Y axis is the counts(which varies). I actually did the Xrd of my as received powder which is stress free. I did a (2- Theta) Scan of the powder. I can send you one of my sample calculation so that you have a look at it. The problem is my as received powder should give me a higher value but i am getting the grain size about 15 nm which i don't think is possible. I have checked the Xrd of both as received as well as 10 hrs ball milling. I have not annealed my powder.
Thnak you
krishna priya
 
Yes, if you include the calculation as an attachment in your next post (click on "Go Advanced" below the Reply box, and scroll down to "Manage Attachments") I can take a look at it. Also, if you can include the raw data in 2-column ascii (a .txt file will work) format and attach that too, it will be useful.
 
Hai Gokul,

I am sending you a word file as i cannot attach an Excel file. So as soon as you receive my attachment you have to change the extension to"Grain size.xls".I calculated the grain size of both as received and 10 hrs of ball milled powder. I hope you can understand my calculation. If not please let me know. I have also enclosed charts in them, you can have a look at them too.
Thank you
krishna priya
 

Attachments

  • #10
Just glanced at it now. My initial estimate agrees with yours ~ 30nm for the "as received" powder. I'll need to do a more careful calculation when I find a little more time - there may be a large instrument-related broadening that is making the particle size look small.

A few questions:
1. What is the expected range of the particle size? Have you tried resolving particles under an optical microscope?
2. This isn't stated explicitly, but it appears you are using Cu-K_alpha1. Is that right?
3. What is the model of diffractometer this was measured with?
 
  • #11
After spending a little more time with the data, I'm beginnig to think the scan was too fast to be able to do a reasonable particle-size calculation. The error bars on the data may be too large to do a meaningful analysis of the various contributions to the peak broadening (see post#4, above).

In any case, here's how I would do it from the given data:

1. For each of the 5 peaks, do a Gaussian fit to the data (limit it to about 3-sigma from the mean) and determine the width(\beta) and the peak value (\theta) from the Gaussian (not from the raw data), as well as the errors in these values \delta\beta~,~~ \delta \theta. For the baseline subtraction in half-maximum determination, I would use two sets of baseline data about 3 degrees on either side of each peak and use a mean of both sets (this is to minimize any angular dependence in the background).

2. Calculate the 5 values of \beta ^2 cos ^2\theta and plot these points (along with their error bars) against sin^2 \theta, for each of the two data sets (as received, and 10-hrs milled).

3. If you are lucky, and the error bars are small enough, you will be able to fit a straight line to both sets and extract a slope and intercept for each one.

4. What these 4 numbers represent:

(from post #4)

\beta^2 = \left( \frac{A}{cos \theta} \right) ^2 + (Btan \theta)^2 + C^2

\implies \beta^2 cos^2 \theta = A^2 + (Bsin \theta)^2 + (Ccos \theta )^2

= A^2 + (Bsin \theta)^2 + C^2(1-sin^2 \theta )

= A^2 + C^2 + sin^2 \theta(B^2 - C^2)

From the linear fit, you have slope=B^2 - C^2~, ~~intercept=A^2 + C^2.

(i) For the as received powder, B (the stress-related broadening coefficient) should be small and hence it might be reasonable to neglect it. Note that, this would mean the slope for this line should be negative. If you do not have a negative slope, you can not neglect the stress in the as received sample; this makes the calculation more tricky.

(ii) The broadening coefficient from instrumental factors should be sample independent, so C is the same number for both samples. If the intercepts for both samples are close to each other, the broadening is probably dominated by intrument-related factors.

That leaves you with 4 unknowns : A(as rec'd), A'(milled), B'(milled) and C(instr). From the two data sets you have 4 equations for these 4 unknowns (2 slopes and 2 intercepts).

5. Finally, using A=0.9 \lambda / d, you can determine the approximate particle size, d.
 
Last edited:
  • #12
Hai,

Thank you very much for being so patient and replying me.The diffractometer which i am using is Philips 1830, and i am using Cu-K_alpha1. I have done a detailed analysis of about 45 min for my powder. Anyways i will try your suggested method. Can you do me a small favour can u post me a sample calculation so that it might be helpful forme to proceed.
I have read in some books about the method which you posted (Suryanarayana). I also tried a Gaussian fit to my curves but i still end up with the same value. So, i will be very thankful if you will send me any sample claculation.
Thank you
krishna priya
 
  • #13
particle size calculation

To calculate particle size, I want to use the Scherrer equation. My problem is that i don't know what are the value of constant k that should be used. In the note that I read the value of k varies between 0.89 to 1.39 depending on crystallite size and shape.. For a small cubic crystal of uniform size normally used k=0.94. Now I'm calculating alumina so what value of k should be used??
 
  • #14
Dear frinds,
I have a problem in structure analysis. The PXRD (Philips Diffractometer) data are with me. Can anyone please help me in the analysis?
I can provide you *.uxd files, with other details.
I have used Mn2O3 and VO2 to form the compound.
Regards,
Girish
 
  • #15
XRD of microsized TiO2

Hello to everybody,

I performed some XRD patterns of two sample of titanium dioxide, one with micron particles and the other one with nano particles. While with the nanosized one I didn't have any problems to apply the Scherrer equation to evaluate the particle size, with the microsized I couldn't. I know that Scherrer equation is valid in a range 10nm - 100nm and my microsized sample should have bigger particles, between 100nm and 200nm.

Could anyone help me with this matter?

Thanks a lot in advance

Shaka:rolleyes:
 
  • #16
Hi guys and gals, I have a question. As the axis is in 2θ degrees, do we have to half the FWHM values as we do for the cosθ.
 
  • #17
Hai,
I want to calculate the particle size, by Scherrer equation. My problem is that i don't know what are the value of constant k that should be used. In the note that I read the value of k varies between 0.89 to 1.39 depending on crystallite size and shape.. For a small cubic crystal of uniform size normally used k=0.94. Now I'm calculating nano Zro2 so what value of k should be used??

Thankyou

anbu
 
  • #18
How to estimate particle size...

Hi friends,
I'm studying dusty plasmas. I made in situ FTIR measurements...is it possible to estimate particle radius and density if absorption and diffusion (diffusion means scattering) are present on infrared spectra? How?
 
  • #19
How do I use the Debye Scherrer Equation

Hi! I plan to determine the size of my deposits using XRD and the debye scherrer equation. Could anyone tell me how this is done? Thanks!

- Buknoy
 
  • #20
@ kpullab
Thank You very much for the .xls file really appreciate it.
 
  • #21
also I need to know the value of K in Scherrer formula when the system is tetragonal chalcopyrite semiconductor.plaese send the answer on my e-mail
 
  • #22
I have XRD measurements data of (004),(111) and (115). How to calculate the FWHM value and dislocation density of my samples by Williamson-Hall plot. Please guide me how to calculate.
 
  • #23
I am trying to use the information from your post #11 to use with my data. In following the derivation in your post, you quote the final formula as A = 0.9 Lamda/d.

I'm wondering if you are implyin that A = Beta/cos(theta) as that would match your derivation, or maybe you are showing a simplification here.

I have particles on a surface that are between 200 - 700 nm, but do not know how they may recombine under the surface. I originally was using the
A = 0.9 Lamda*cos(theta) /d
formula that I have seen others use, but am now concerned since my particles are clearly greater than 1000 Angstroms... can you please advise?
 
  • #24


anil.bose said:
Hi friends,

What is the best way to analyze the XRD data for phase identification and particle size calculation? so far I am using debye-scherrer's formula. Is there any advanced formula?

thankyou

how to determine particle size by scherrer formula from xrd data?
 
  • #25
Dear gokul,
we have got a program named xpert plus, which gives us fwhm value for the selected peak. is the program feasible. can we get a program which could directly find the particle size.
 
  • #26
Dear gokul
i am fascinated by your helping nature.
i have got a program named xpert plus which finds fwhm for the xray data with extension .raw. is the program feasible? can i have a program which could give the particle size directly.
Nigarbage
 
  • #27
Plot the data (2θ vs. Counts)
Determine the Bragg Angles for the peaks
Calculate d and a for each peak
e.g. if λ = 1.54 Ǻ, θB=38.3

d = λ / (2 Sin θB)
= 1.54 Ǻ / ( 2 * Sin ( 38.3 / 2 ) )
= 2.35 Ǻ
 
  • #28
Hi all,

Kudos to the good work. I have a glitch with my XRD data analysis. Hope one of you have a way to work it out. I am trying to know the grain size of my NiMnGa alloy thin films. They are on Si and partially oxidized silicon substrates. About 500nm thick. The problem however is, my XRD only shows one peak.(Please note that they XRD was done on powder diffractometer that is usually used for powders.) How can I calculate the particle size from a single peak.

If I am successful in doing that using Scherrer, how accurate would it be. The peak position is at 43.9 deg and intensity about 1200 counts.

Thank you for your help.
 
  • #29
to determine the particle [grain] size (d) you need to use the Scherrer formula;

d=0.9λ/(Bcosθ),

where λ is the X-ray wavelength, B is the corrected peak width and θ is the Bragg
diffraction angle
 
  • #30
Thanks Drizzle.

I figured that out from a XRD textbook. My concern however is the accuracy of such calculated value. I know that scherres calculations are +/- 10% accurate. But in my case, due to just one peak, I do not have anything to compare the calculation with. I don't have a way to eliminate the broadening due to stress. Do you know how I can solve this?

Also, do you know if I could use a single crystal silicon wafer, to calculate the instrumental broadening?

Thank you.
 
  • #31
you do realize there’s something known as the full width at half maximum (FWHM);

B2=B2obs – B2m
where Bobs is the measured peak width and Bm is the beak broadening due to the machine.
[it also can be determined by ω scan* which could vary with increasing temperature depending on your sample]. as for your question about how accurate is it? it’s an accurate method to determine the size of the grain, I don’t think you need to compare it to any. in fact, it can measure up to nano scale and can specify if the sample powder is nano particles or not, as far as I know this is the most effective measuring method and the most used. hope I’m helping


* the crystal quality of a sample is confirmed by a ω scan (out-of-plane rocking curve θ[degrees] vs. intensity[a.u.]), which could measure for a specific peak. and by saying half maximum it means the half-way of the peak's intensity. this shows the variation of the FWHM of the XRD ω scan for the peaks as a function of the film growth temperature [since it has been asked previously].
 
  • #32
kpullab said:
Hai Gokul,

I am sending you a word file as i cannot attach an Excel file. So as soon as you receive my attachment you have to change the extension to"Grain size.xls".I calculated the grain size of both as received and 10 hrs of ball milled powder. I hope you can understand my calculation. If not please let me know. I have also enclosed charts in them, you can have a look at them too.
Thank you
krishna priya

Nice calculations, thanks.
As everyone else in the field, you used 0.9 as the shape factor, though.
Now I know the shape of my crystals from TEM work. When would you use numbers other than 0.9, and do you have a table of them (with reference)?
Thanks
 
  • #33
anbulagam said:
Hai,
I want to calculate the particle size, by Scherrer equation. My problem is that i don't know what are the value of constant k that should be used. In the note that I read the value of k varies between 0.89 to 1.39 depending on crystallite size and shape.. For a small cubic crystal of uniform size normally used k=0.94. Now I'm calculating nano Zro2 so what value of k should be used??

Thankyou

anbu

Could you kindly supply us all with the references from which you derived those numbers?
thanks!
 
  • #34
kpullab
Thank you very much for the calculation details. They were very useful
 
  • #35
Hi guys, i am referring to some of my old notes when I had done grazing incidence studied on some Silver thin films and calculated grain size using the d.s. equation.
using cu kalpha source . I have noted the grain size to be 79.504 / (beta.cos(theta))
e.g. 2theta for the 111 peak is 38.116 deg and the fwhm is 0.4724 . I have noted the grain size according to the above formula to be 178.05 . Myquestion is I don't recall where that 79.504 came from? Can somebody please help me figure it out if at all it is correct?
 
  • #36
I think i used another multiplier other than 0.9 and converted the fwhm in degrees to radians. thus the 79.504 . does that make sense?


arairkar said:
Hi guys, i am referring to some of my old notes when I had done grazing incidence studied on some Silver thin films and calculated grain size using the d.s. equation.
using cu kalpha source . I have noted the grain size to be 79.504 / (beta.cos(theta))
e.g. 2theta for the 111 peak is 38.116 deg and the fwhm is 0.4724 . I have noted the grain size according to the above formula to be 178.05 . Myquestion is I don't recall where that 79.504 came from? Can somebody please help me figure it out if at all it is correct?
 
  • #37
I would highly recommend picking up a copy of Cullity's https://www.amazon.com/dp/0201610914/?tag=pfamazon01-20. It covers all aspects of x-ray diffraction analysis and is a fantastic reference. Try to find a copy of the second edition or earlier, if you can. The third edition is all screwed up.
 
Last edited by a moderator:
  • #38
Do anyone know GSAS or Fullproof?
 
  • #39
Hi

I'm looking to determine crystallite size from micro-diffraction data I have collected using synchrotron radiation. In the Scherrer equation do I still need to take into effect instrument broadening with synchrotron radiation? Would using a silicon standard allow me to take into effect the broadening?

Cheers
 
  • #40
Dear sir,

I need one detail in XRD analysis. If possible, please provide the following details.

In a Research paper Preparation of Silver Nanoparticles and Their Characterization - Journal of Nanotechnology online. DOI: 10.2240/azojono0129.
It has been written in XRD Study about standard diffraction angle of Silver Nano particle [2 Theta in degrees] is 44.3. I like to know, from where I can get this data? Can you intimate standard diffraction angle of Copper Nanopowder [2 Theta in degrees]

I expect your reply shortly.

Mrs.T.Theivasanthi,
Lecturer in Physics,
PACR Polytechnic College ,
Rajapalayam – 626108, Tamilnadu , India .
Email: sankarg4@yahoo.com Mobile : 9245175532
 
  • #41
Hello Theivasanthi,
I am not sure whether this will be helpful..but at least you get some insight!
I had a look at that paper [DOI is mis-typed in your post].
First find d from the usual formula:
<br /> d=\frac{a}{\sqrt {h^2+k^2+l^2}}<br />
I notice in that paper it is a [2 0 0] scattering. So h=2, and k=l=0.
For a:
As silver is a FCC structure we have
<br /> 2r\sqrt 2=a<br />
Here atomic radius for silver r=144 pm (you can refer some books or in internet). From this i get a=407.2935 pm. If you substitute a in d's formula by taking \lambda=0.1541~nm (wave length value taken from that paper)..
you get d=0.203646 nm.
Finally you know the famous bragg law:
<br /> 2d\sin\theta=n\lambda<br />
take n=1 and you get \theta=22.2316\;° and 2\theta=44.46\;°.
-Rajini
PS: So for copper i hope you can easily determine 2\theta!
 
Last edited:
  • #42
Dear sir,
Please refer to the TABLE -1 of said paper," Preparation of Silver Nanoparticles and Their Characterization" . 2 Types of Diffraction angles are there. ( Experimental and standard diffraction angles of SILVER specimen ).

I like to know, from where I can get this standard diffraction angle of Silver Nano particle [2 Theta in degrees] 44.3 ?

Can you intimate standard diffraction angle of Copper Nanopowder [2 Theta in degrees]



Rajini said:
Hello Theivasanthi,
I am not sure whether this will be helpful..but at least you get some insight!
I had a look at that paper [DOI is mis-typed in your post].
First find d from the usual formula:
<br /> d=\frac{a}{\sqrt {h^2+k^2+l^2}}<br />
I notice in that paper it is a [2 0 0] scattering. So h=2, and k=l=0.
For a:
As silver is a FCC structure we have
<br /> 2r\sqrt 2=a<br />
Here atomic radius for silver r=144 pm (you can refer some books or in internet). From this i get a=407.2935 pm. If you substitute a in d's formula by taking \lambda=0.1541~nm (wave length value taken from that paper)..
you get d=0.203646 nm.
Finally you know the famous bragg law:
<br /> 2d\sin\theta=n\lambda<br />
take n=1 and you get \theta=22.2316\;° and 2\theta=44.46\;°.
-Rajini
PS: So for copper i hope you can easily determine 2\theta!
 
  • #43
I like to know about " d spacing " in XRD data analysis.
 
  • #44
Hello,
I assume authors might have theoretically calculated the standard diffraction angle (please see my previous post for the theoretical calculation).
And the d spacing in table 2 might be the experimentally obtained one based on 45 degree.
d spacing is the spacing between two successive latices.
For copper you can easily calculate the 2\theta by taking h\,k\,l=[2 0 0] and
\lambda as 0.1541 nm.
Hope this helps
rajini
 
  • #45
For Copper
r=128 pm and it is also a FCC structure!
 
  • #46
How to calculate h,k,l for every peak in a diffraction graph?
Also, I like to know about " d spacing " in XRD data analysis. What is the uses of calculating " d spacing " ?
 
  • #47
Aha, it is just some work but not difficult..by using brag law and mathematics ..usually this is done from some assumption and using square bragg law..
Actually i did this long back in my practical courses..(tomorrow i'll send you by email)..
Also there is good book for this indexing diffractogram they also showed a example..(i also forgot the name of it) but starts with a textbook for materials science...however will try to recollect the name and let you know.
D-spacing or lattice parameter gives important information on the structure of material..
good luck
 
  • #48
Did you study "X-Ray Diffraction Studies of Copper Nanopowder" in
http://arxiv.org/a/thirugnanasambandan_t_1 . please reply.

Rajini said:
Aha, it is just some work but not difficult..by using brag law and mathematics ..usually this is done from some assumption and using square bragg law..
Actually i did this long back in my practical courses..(tomorrow i'll send you by email)..
Also there is good book for this indexing diffractogram they also showed a example..(i also forgot the name of it) but starts with a textbook for materials science...however will try to recollect the name and let you know.
D-spacing or lattice parameter gives important information on the structure of material..
good luck
 
  • #49
Hi,
I am a research scholar in chemistry. I am stuck with phase identification of my powder xrd anaysis through Fulproof. I don't have the PDF data base so its difiicult for me to identify the phases. Moreover the crystall size i calculate from scherrers formula is also not coming correct.
please suggest me to solve these problems. I am attaching my zip file of my system.Kindly help me.

Awaiting reply
hanif
 

Attachments

  • #50
I have some questions sort of related to this.

I have glancing angle XRD of some silver films with varying thickness. 1. The peaks that show up are <111>, <200> and <220>. Are the <200> and <220> same as <100> and <110> respectively. Why is the reason that <110> and <100> not show up?

I tried to calculate grain size for each phase for the films, but I think it is in accurate since it does not seem to change with film thickness.

Thirdly, how do I calculate texture fractions for each of the phases from this data without having to do pole figures.
 
Back
Top