Is this tantalum capacitor damaged?

  • Thread starter Thread starter gnurf
  • Start date Start date
  • Tags Tags
    Capacitor
AI Thread Summary
The discussion revolves around troubleshooting a 1995 LeCroy oscilloscope exhibiting irregular behavior, potentially linked to a tantalum capacitor. The capacitor in question shows a bump that raises concerns about damage, though its capacitance measured within spec after removal. Users suggest comparing it to similar capacitors on the board and consulting manufacturer resources for visual references. While the capacitor could be a contributing factor, the scope's issues may stem from other components, necessitating thorough troubleshooting. Ultimately, the scope has been reassembled and appears to function correctly, highlighting its value for educational purposes.
gnurf
Messages
370
Reaction score
8
I'm troubleshooting an old (1995) LeCroy scope that has shown some irregular behavior lately ("multiple calibration errors" warnings, display freezes for a while when changing the volt/div and time/div settings, etc), especially after running for a while. I won't bother you with all that—I'm just wondering if the cap in the picture is damaged? The picture quality is not great, but it's the best I could do. I thought it looked worse in a 4X microscope where the bump seemed darker/grey/black in the middle. Any thoughts?
 

Attachments

  • _failed_tantalum.jpg
    _failed_tantalum.jpg
    56.3 KB · Views: 534
Engineering news on Phys.org
I have no idea, though I can see why you might wonder. Are there others on the board that you can compare with? If you can see a manufacturer's name, maybe try their website there may be photos of their products there.

Good luck! Very disheartening to have to bin a piece of equipment for being unable to trace a faulty component worth a few cents.
 
how does it comapre to the one in the upper right of the pic
the one partly hidden by your enlargement inset ?

Dave
 
Vishay has tantalum series (e.g. 595D) that look exactly like that:

R100163-06.jpg


I'm still not sure if it's a sign of damage or not, but the other similar looking caps on the board do not have this bump, and Vishay, of course, wouldn't put one of their ugliest ducks in a marketing photo. Anyway, I've replaced it now and after a complete disassemble/reassemble it seemingly works again. Time will tell, I guess.

What is certain is that the scope (which I got for free) is not going in the bin any time soon. It's a 500MHz-2Gs/s monster compared to what I had before, and the fact that you can download the complete 400 page manual complete with schematics, BOM list, etc, blows my mind a little bit—very educational! You old heads had it too easy. :)

Edit: Btw, I measured the capacitance once the cap was out of circuit and it was within spec. Google: Is there some definitive capacitor test I could do?
 
Last edited:


Based on the information provided, it is difficult to determine if the tantalum capacitor is damaged without further inspection. The irregular behavior of the LeCroy scope could be caused by a number of factors, not just a damaged capacitor. It would be best to thoroughly troubleshoot the scope and rule out any other potential issues before concluding that the capacitor is the cause. If the capacitor does appear to be damaged, it is possible that it could be contributing to the issues with the scope. However, without a closer examination and analysis, it is impossible to determine the extent of the damage and its impact on the scope's performance. It is recommended to consult with a professional technician or engineer for a more accurate assessment of the capacitor's condition.
 
Thread 'Electromagnet magnetic field issue'
Hi Guys We are a bunch a mechanical engineers trying to build a simple electromagnet. Our design is based on a very similar magnet. However, our version is about 10 times less magnetic and we are wondering why. Our coil has exactly same length, same number of layers and turns. What is possibly wrong? PIN and bracket are made of iron and are in electrical contact, exactly like the reference design. Any help will be appreciated. Thanks. edit: even same wire diameter and coil was wounded by a...
Very basic question. Consider a 3-terminal device with terminals say A,B,C. Kirchhoff Current Law (KCL) and Kirchhoff Voltage Law (KVL) establish two relationships between the 3 currents entering the terminals and the 3 terminal's voltage pairs respectively. So we have 2 equations in 6 unknowns. To proceed further we need two more (independent) equations in order to solve the circuit the 3-terminal device is connected to (basically one treats such a device as an unbalanced two-port...
I have recently moved into a new (rather ancient) house and had a few trips of my Residual Current breaker. I dug out my old Socket tester which tell me the three pins are correct. But then the Red warning light tells me my socket(s) fail the loop test. I never had this before but my last house had an overhead supply with no Earth from the company. The tester said "get this checked" and the man said the (high but not ridiculous) earth resistance was acceptable. I stuck a new copper earth...
Back
Top