Measuring Open Circuit Voltages with Sensing Resistors and a CRO

  • Thread starter Thread starter houlahound
  • Start date Start date
  • Tags Tags
    Resistors
Click For Summary

Discussion Overview

The discussion revolves around measuring open circuit voltages from various power supplies using a Cathode Ray Oscilloscope (CRO). Participants explore the behavior of these voltages under different conditions, particularly focusing on the effects of load resistors and the nature of the power supplies.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested

Main Points Raised

  • One participant observed spurious readings when measuring open circuit voltages, questioning why these voltages could not be measured without a load resistor.
  • Another participant explained that high open circuit voltages are common, particularly in low power AC and DC supplies, due to internal resistance and capacitor behavior.
  • A participant noted that adding a resistor across the output cleaned up the signal, suggesting that the measurement was affected by noise when no load was present.
  • One reply proposed that the lack of load could disconnect the output from the internal components of the power supply, leading to stray pickup from nearby power lines.
  • Another participant mentioned that they had previously obtained expected voltages using a digital multimeter, implying a difference in measurement techniques.
  • Some participants speculated that the internal design of the power supply might include a resistance across the output to maintain a slight load.
  • One participant expressed embarrassment over possibly overlooking a simple explanation related to the CRO leads' connection.
  • A later reply encouraged experimentation to verify the observations, indicating a desire for further exploration of the issue.

Areas of Agreement / Disagreement

Participants generally agree that the behavior of the open circuit voltage is influenced by the characteristics of the power supplies and the measurement techniques used. However, there is no consensus on the exact reasons for the observed phenomena, and multiple competing views remain regarding the underlying causes.

Contextual Notes

The discussion highlights potential limitations in measurement techniques, such as the connection method of the CRO leads and the internal design of the power supplies, which may affect the readings. However, these aspects remain unresolved.

Who May Find This Useful

Individuals interested in electrical engineering, power supply design, or measurement techniques may find this discussion relevant.

houlahound
Messages
907
Reaction score
223
I measured a bunch of open circuit voltages across a variety of power supplies using a CRO. the signal is just spurious.

Adding a resistor across the output of the same power supplies gives the expected trace. I used high and low impedance option which was 10 megaOhms on the CRO leads.

What is going on, in theory the open circuit voltage in this case max 12 volts should be measurable without a load resistor to draw a current??

Apologies if a dumb question.
 
Engineering news on Phys.org
A high open circuit voltage from a power supply is common. The reason depends on the type of supply.

If it is a low power AC 'wall wart' style these are usually just a small transformer and when you draw current theire is a voltage drop due to the internal resistance of the transformer.

If it is a low power DC 'wall wart' style these are usually just a small transformer, a diode, and a capacitor for filtering. they have the same problem as the AC 'wall wart' supply, internal resistance and additionally, at no load, the capacitor charges up to the transformer peak voltage. When a load is supplied, the capacitor voltage drops until the next power line voltage peak occurs to recharge it.

These small supplies are rated for output voltage at their rated load current.

(EDIT) added last sentence
 
The power supplies were in fact inexpensive AC and DC constructed as you said. I also used a high quality 40 Amp, regulated 12 volt supply from my radio station.

The spurious reading was kind of noise on the mains hum. It was not really the output signal at all. Any random resistor placed across the output and the CRO put in parralel with the resistor fully cleaned up the signal and made the measurement sensible.

I am just wondering why it does that.
 
Hard to say for sure. What comes to mind as most likely is with no load on that 40A supply the output was pretty much disonnected from the internals. This would make your measurement about the same as putting the CRO probe on a length of wire that's not connected to anything, you would see stray pickup from the power lines.
 
That's exactly what it looked like.

If I recall I have done this test with a digital multimeter and get the precise voltages expected.
 
Due to the simple and quick nature of this little experiment I would like to see if anyone else can reproduce it.
 
It's probably dependent on the internal details of the particular power supply. Many have a resistance across the output so always have a slight load on them.
 
You know I think that you solved it. The multimeter plugs into the output ports. The CRO leads kind of contact the outer conductor which may not be connected.

I am embarrassed if it is that simple.
 
All I can say is "Welcome to the crowd."
(It happens to all of us on occasion)
 
  • #10
Will attach my CRO leads to some banana plug leads that push an inch into the output port of the power supply to verify.

Really hoping on one hand it is a deeper more mysterious problem...but on the other hand.
 

Similar threads

  • · Replies 5 ·
Replies
5
Views
2K
  • · Replies 13 ·
Replies
13
Views
3K
Replies
12
Views
3K
Replies
2
Views
2K
  • · Replies 9 ·
Replies
9
Views
5K
  • · Replies 5 ·
Replies
5
Views
2K
  • · Replies 12 ·
Replies
12
Views
3K
Replies
8
Views
2K
Replies
2
Views
2K
  • · Replies 10 ·
Replies
10
Views
2K