Microscope condensers and resolution kohler illumination

  • Context: Graduate 
  • Thread starter Thread starter u0362565
  • Start date Start date
  • Tags Tags
    Microscope Resolution
Click For Summary

Discussion Overview

The discussion revolves around the principles of Kohler illumination in brightfield microscopy, specifically focusing on the optimization of condenser alignment and its impact on resolution. Participants explore the relationship between numerical aperture (NA), light cones, and the roles of the condenser and field diaphragm in microscopy.

Discussion Character

  • Technical explanation
  • Conceptual clarification
  • Debate/contested

Main Points Raised

  • One participant describes resolution in microscopy as the minimum detectable distance between two objects, emphasizing the role of light rays at obtuse angles in achieving higher resolution.
  • Another participant clarifies the distinction between Kohler illumination and critical illumination, noting that Kohler illumination images the Fourier transform of the light source onto the sample plane.
  • Concerns are raised about the effects of stopping down the condenser, with one participant suggesting that it can improve image contrast and depth of focus.
  • There is a discussion on the necessity of the field diaphragm, with one participant questioning its role in focusing the source at the condenser's front focal plane versus the specimen plane.
  • Another participant mentions that stopping down the condenser can reduce lens aberrations, which may enhance image quality.

Areas of Agreement / Disagreement

Participants express differing views on the roles of the condenser and field diaphragm, particularly regarding their impact on resolution and image quality. There is no clear consensus on the necessity and function of the field diaphragm in relation to the condenser.

Contextual Notes

Participants acknowledge that the relationship between the numerical apertures of the condenser and objective lenses is complex and that adjustments to the condenser aperture can influence resolution and image characteristics. The discussion includes assumptions about the behavior of light and lens optics that may not be universally agreed upon.

u0362565
Messages
51
Reaction score
0
Hi all,

I have a couple of questions about the process of kohler illumination (optimised condenser alignment) in brightfield microscopy and how this affects the final resolution attained.

Resolution concerns the minimum detectable distance between two objects and the closer to structures are, then a greater degree of diffraction occurs. Ultimately this means that higher resolution information is encoded by light rays that are at a more obtuse angle if the optical axis is zero degrees, hence the wider the acceptance angle of the lens (NA) the higher the resolution.

The light leaving the specimen can be seen as a mirror image of the light cone emanating from the condenser. If the condenser is closed down too far then the angles over which the objective lens can receive light is reduced and thus the full NA of the lens is not realized. Is that an acceptable way of thinking about this?

I'm slightly confused though by the opposite case, if the condenser aperture is opened too far. The acceptance angle range of the lens is fixed, so why does a wider light cone mean that the NA of the lens is not realized? Does closing the aperture force light into the lens, an over simplification i know..

The first image on this page gives a nice illustration of how I'm visualising the light cones emanating from the condenser and specimen.
http://www.olympusmicro.com/primer/anatomy/condensers.html

My second question concerns the field diaphragm. Why do you need a physical barrier to prevent some light from reaching the condenser? The crucial step is controlling the angles of light allowed to pass to the objective lens so why not let the condenser do all the work?

Thanks for your help!

Matt
 
Science news on Phys.org
There are a few issues here to discuss. First, a clean definition of Kohler, and critical, illumination: critical illumination images the source onto the sample plane, while Kohler illumination images the (spatial) Fourier transform of the source onto the sample plane. Both illumination conditions maximize the illumination efficiency, both provide independent control of the field and aperture diaphragms, and in both cases aberrations in the illumination optics do not reduce the optical performance of the objective lens.

Because the condenser NA is independent of the objective NA, two main effects occur, corresponding to NA_cond << NA_obj and NA_cond > NA_obj. Overall, resolution is given by

When NA_cond > NA_obj, image contrast is often lowered due to the 'mixing' of unscattered high NA rays and light from low NA rays scattered by then sample into higher NA. By slightly stopping down the condenser, image contrast is often increased, as is the depth of focus.

When the NA_cond << NA_obj, the sample is essentially illuminated by a plane wave, which can be used to emphasize diffraction effects. For example, if you are looking at a grating, stopping the condenser all the way down will provide improved images of the diffracted orders.

As for *why* the condenser should be stopped down at all, the answer has to do with aberrations- lens aberrations increase as the diameter of the stop increases- stopping down *any* lens will reduce aberrations (except for distortion).

Does this help?
 
Hi Andy,

Yes thanks you made some good points there. Ultimately when I'm describing use of the condenser i'll make it clear that the achievable resolution is equal to the Na of the objective + NA condenser and NA condenser is changed by adjusting the condenser aperture. Would you say that the field diaphragm is purely used to ensure that the source is focussed at the front focal plane of the condenser rather than at the specimen plane?

Thanks again,

Matt
 
The field diaphragm defines the illuminated region of the sample. I only stop down the condenser field diaphragm for two reasons- during initial alignment, to ensure the condenser and objective are focused to the same sample plane, and if I am concerned about sample bleaching (fluorescence imaging)- I don't want to illuminate portions of the sample I am not looking at.

With Kohler illumination, focusing the source requires the ability to view pupil planes, either with a Bertrand lens or some other method. This is not always easy to do, which is why many lamps have a diffuser plate.
 

Similar threads

  • · Replies 1 ·
Replies
1
Views
1K
Replies
1
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 13 ·
Replies
13
Views
3K
  • · Replies 5 ·
Replies
5
Views
2K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 10 ·
Replies
10
Views
7K
Replies
4
Views
4K