The discussion revolves around the design and implementation of an EMG (electromyography) data acquisition system. Participants explore various aspects of the design, including the choice of instrumentation amplifier, gain settings, and filter types suitable for obtaining clean EMG signals. The conversation includes both theoretical considerations and practical advice for building the system.

Discussion Character

  • Exploratory
  • Technical explanation
  • Conceptual clarification
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • One participant inquires about the appropriate gain for the instrumentation amplifier and the suitable type of band-pass filter to use for clean EMG signals, suggesting either Sallen-Key or Butterworth filters.
  • Another participant emphasizes the need to calculate the gain based on the peak voltage and input specifications of the ADC, along with defining filter requirements such as ripple and phase specifications.
  • Some participants suggest that trial and error may be necessary when selecting sensors and designing the system, highlighting the importance of practical experimentation.
  • A participant shares insights from personal experience, recommending the use of a differential input amplifier to reduce electrical noise and discussing the characteristics of EMG signals.
  • Concerns are raised about the limitations of schematic design tools, with suggestions to seek guidance from advisors or mentors for more direct assistance.
  • Several participants encourage building the circuit to test its functionality and to use an oscilloscope for troubleshooting rather than adding complexity with an ADC at the initial stages.

Areas of Agreement / Disagreement

Participants express a range of views on the design considerations, with no consensus reached on specific gain values or filter types. The discussion reflects multiple competing approaches and acknowledges the necessity of practical experimentation.

Contextual Notes

Participants mention various factors that could influence the design, such as noise from electrical sources, the need for adjustable gain, and the importance of using low-noise cables. There are also references to specific components and design strategies that may or may not be suitable depending on the intended application.

Who May Find This Useful

This discussion may be useful for students, hobbyists, or professionals interested in EMG signal acquisition, circuit design, and practical experimentation in electronics.

  • #31
Mo_Tuk said:
Should I add isolation circuit or not since I supply my circuit with two 9v batteries?
I don't think you need it . You have no home-made equipment in between you and wall power that's likely to cause a short circuit. Your oscilloscope chassis should be tied to the the Earth grounding plug of your power cord. I would suggest check your outlet with an inexpensive receptacle checker, here's what they look like in USA :
.
upload_2018-10-16_13-30-8.png


Remaini aware you are relying on the internal insulation of any line connected equipment to protect you, so check it carefully for leakage current to Earth ground.

Mo_Tuk said:
and then connected the electrodes to my bicep muscle and try to detect some signal as shown in attached screenshots on the oscilloscope.
Try touching a moistened finger on your other hand to oscilloscope common (outside of the BNC connector if it's metal).
If that reduces noise then you know you need a reference connection to your body.

Mo_Tuk said:
do you mean that I can just connect the circuit as shown in the screenshot below.
Not quite.
That screenshot shows a wire shorting out your signal. Take it out of your sketch and connect REF to power supply common.

old jim
 

Attachments

  • upload_2018-10-16_13-30-8.png
    upload_2018-10-16_13-30-8.png
    34.9 KB · Views: 615
Engineering news on Phys.org
  • #32
Mo_Tuk said:
i connected the circuit (pre-amplification stage using INA114 + Filtering stage of BPF) and then connected the electrodes to my bicep muscle and try to detect some signal as shown in attached screenshots on the oscilloscope.
berkeman said:
Can you post a diagram of your complete circuit, including the battery power supply, amp circuit, patient connections, and Earth Ground scheme?
Mo_Tuk said:
I connected the circuit exactly as shown in screenshot:
Your circuit that you posted is very confusing for me. I asked to see the circuit that you said you connected to your bicep, but your schematic shows instruments, not a pair of patient connections as input. I also don't see Earth ground, and I see very different biasing compared to your latest differential amplifier questions.

Are you using a single 9V battery for power, or two to give you +/-9V? Please always show your power supply configuration in your schematics. What electrodes were you using to connect to your bicep? What brand, and how old are they? I assume they are single-use disposable, right?

When you flexed your bicep, did you see any response at all out of the amplifier string? It should have been much lower frequency than your 50/60Hz interference.

This is more like a circuit diagram that I would expect you to be posting...

https://greenhecktech.files.wordpress.com/2016/09/schematic4.png?w=748

schematic4.png
 

Attachments

  • schematic4.png
    schematic4.png
    20.6 KB · Views: 1,212
  • Like
Likes   Reactions: dlgoff and jim hardy
  • #33
berkeman said:
Your circuit that you posted is very confusing for me. I asked to see the circuit that you said you connected to your bicep, but your schematic shows instruments, not a pair of patient connections as input. I also don't see Earth ground, and I see very different biasing compared to your latest differential amplifier questions.

Are you using a single 9V battery for power, or two to give you +/-9V? Please always show your power supply configuration in your schematics. What electrodes were you using to connect to your bicep? What brand, and how old are they? I assume they are single-use disposable, right?

When you flexed your bicep, did you see any response at all out of the amplifier string? It should have been much lower frequency than your 50/60Hz interference.

This is more like a circuit diagram that I would expect you to be posting...

https://greenhecktech.files.wordpress.com/2016/09/schematic4.png?w=748

View attachment 232292
Yes, I'm using two +/-9V batteries to supply the circuit, and I used surface electrodes "BioProtech Electrodes" which shown in attached image.
Although the signal is 50Hz noise but when I flexed my bicep the amplitude increased but the signal still almost the same "50Hz noise"
 

Attachments

  • BioProtech electrode.png
    BioProtech electrode.png
    38 KB · Views: 399
  • Like
Likes   Reactions: berkeman
  • #34
Mo_Tuk said:
Although the signal is 50Hz noise but when I flexed my bicep the amplitude increased but the signal still almost the same "50Hz noise"
Okay, there are two things that you should approach this with, IMO:
  • You should look more at shielding and balancing your probe lines to cut down on the AC Mains noise pickup -- if you post your full circuit diagram including Earth ground connections and shielding, we should be able to help with that
  • You should consider digitizing the waveform output from your circuit and doing DSP processing to filter out the AC Mains noise. You can do much higher order filters and adaptive filtering with DSP routines. Do you have experience with programming microcontrollers (uCs) in C or Assembly code?
 
Last edited:
  • Like
Likes   Reactions: dlgoff, jim hardy and Mo_Tuk
  • #35
berkeman said:
Okay, there are two things that you should approach this with, IMO:
  • You should look more at shielding and balancing your probe lines to cut down on the AC Mains noise pickup -- if you post your full circuit diagram including Earth ground connections and shielding, we should be able to help with that
  • You should consider digitizing the waveform output from your circuit and doing DSP processing to filter out the AC Mains noise. You can do much higher order filters and adaptive filtering with DSP routines. Do you have experience with programming microcontrollers (uCs) in C or Assembly code?

Thanks @berkeman
I hope this screenshot of the circuit would be clear, and about the DSP part for now I just want to see clear EMG signal from my bicep muscle and then I will move to programming part using Arduino.
and about the electrode cables would you suggest any specific type of cable that less noise produced ??
 

Attachments

  • Circuit Diagram.PNG
    Circuit Diagram.PNG
    24 KB · Views: 398
  • #36
Hi Mo_Tuk,
Could you post a photograph of the assembled amplifier and another photograph of showing it connected to you? Please show where the leads connect to you and to the amplifier. With only the wiring diagram, it is very hard for us to guess what may help to get it working.

The waveform you posted earlier (https://www.physicsforums.com/attachments/tek0003-jpg.232273/) looks like there is a DC power supply nearby that is making some electrical noise, maybe a battery charger or computer, or maybe even an LED lamp!

Cheers,
Tom
edit: fixed typo
 
Last edited:
  • Like
Likes   Reactions: jim hardy and Mo_Tuk
  • #37
Mo_Tuk said:
I hope this screenshot of the circuit would be clear,

we trust you tied the junction of your batteries to circuit common

Mo_Tuk said:
for now I just want to see clear EMG signal from my bicep muscle
and shielded your signal wires

i added those features to your drawing

upload_2018-10-19_6-53-45.png
 

Attachments

  • upload_2018-10-19_6-53-45.png
    upload_2018-10-19_6-53-45.png
    24.8 KB · Views: 831
  • Like
Likes   Reactions: Mo_Tuk
  • #38
Tom.G said:
Hi Mo_Tuk,
Could you post a photograph of the assembled amplifier and another photograph of showing it connected to you? Please show where the leads connect to you and to the amplifier. With only the wiring diagram, it is very hard to for us to guess what may help to get it working.

The waveform you posted earlier (https://www.physicsforums.com/attachments/tek0003-jpg.232273/) looks like there is a DC power supply nearby that is making some electrical noise, maybe a battery charger or computer, or maybe even an LED lamp!

Cheers,
Tom

Thanks @Tom.G
do yo mean photograph of the circuit connected on the breadboard??
and about the electrical noise I think the laptop was near to my circuit and plugged into charger.
 
  • #39
jim hardy said:
we trust you tied the junction of your batteries to circuit commonand shielded your signal wires

i added those features to your drawing

View attachment 232407

Thanks @jim hardy
for the connection you made of electrodes A&B to reference, would that help to reduce noise or something else??
 
  • #40
Mo_Tuk said:
for the connection you made of electrodes A&B to reference, would that help to reduce noise or something else??
The red dashed lines do NOT represent connections to the electrodes,
instead
they represent shields around the wires bringing signal from the electrodes to your circuit board.

The purpose of a shield is to physically surround the signal wire so the can be no capacitve coupling into it.
Recall from basic physics that no E field can exist inside a closed conducting surface - google "Faraday Shield".

Using a twisted pair to route both signals together will reduce magnetic coupling.
That's why "twisted shielded pair" is so popular.

Your 50 hz noise is probably a combination of pickup that's
capacitively and inductively coupled into the input wires ( and gets amplified 1100X) ,
and
capacitively coupled across the power transformer in your oscilloscope's power supply.

Shielding the input wires will help .

old jim
 
Last edited:
  • Like
Likes   Reactions: Mo_Tuk
  • #41
Tom.G said:
looks like there is a DC power supply nearby that is making some electrical noise, maybe a battery charger or computer, or maybe even an LED lamp!

Emphasizing @Tom.G 's comment"Driven Shield" is a method that's used in precision equipment .
It's shown in the INA111 datasheet

upload_2018-10-19_8-5-7.png


the cylinder surrounding the signal wires represents a shield that is driven to nearly average voltage of the two signals
which minimizes voltage between the shield and signal wires
minimizing capacitive coupling of current into the signal wires.

INA111's input impedance is so high that your signal wires act like an antenna .
Shielding hides them from electric fields by creating a Faraday Cage.
Twisting them reduces their sensitivity to magnetic fields by reducing the area(hence the magnetic flux) they encircle.
That's why we EE's like twisted shielded pair for signal wires..

Try grounding shields first to see if it helps. Driven shield should give even better results.

old jim
 

Attachments

  • upload_2018-10-19_8-5-7.png
    upload_2018-10-19_8-5-7.png
    9.6 KB · Views: 553
Last edited:
  • Like
Likes   Reactions: Mo_Tuk and dlgoff
  • #42
Mo_Tuk said:
do yo mean photograph of the circuit connected on the breadboard??
Yes.
 
  • Like
Likes   Reactions: Mo_Tuk
  • #43
Mo_Tuk said:
about the electrode cables would you suggest any specific type of cable that less noise produced ?
Belden 2221 or 11700A would be good choices but will be hard to find and the 2221 costs about $2 to $3 USD per foot ($6 to $9.3 USD per meter). A little bit cheaper if you buy 1000 feet (330 meters).

I suggest you first try whatever shielded wire you can get and twist the wires together, about 3 or 4 twists per foot (10 per meter) should help.

A possible source is: https://www.nassauelectrical.com/pr...ingle-pair-bc-stranded-30x40-pvc-jacket-cable
 
  • Like
Likes   Reactions: Mo_Tuk
  • #44
berkeman said:
Okay, there are two things that you should approach this with, IMO:
  • You should look more at shielding and balancing your probe lines to cut down on the AC Mains noise pickup -- if you post your full circuit diagram including Earth ground connections and shielding, we should be able to help with that
  • You should consider digitizing the waveform output from your circuit and doing DSP processing to filter out the AC Mains noise. You can do much higher order filters and adaptive filtering with DSP routines. Do you have experience with programming microcontrollers (uCs) in C or Assembly code?
I have also another type of electrodes (Covidien Kendall Q-Trace Diagnostic Tab Electrodes Resting ECG Electrodes) , is that type better than (BioProtech)?
 

Attachments

  • Covidien Kendall Q-Trace Diagnostic Tab Electrodes.jpg
    Covidien Kendall Q-Trace Diagnostic Tab Electrodes.jpg
    17.2 KB · Views: 368
  • #45
Mo_Tuk said:
I have also another type of electrodes (Covidien Kendall Q-Trace Diagnostic Tab Electrodes Resting ECG Electrodes) , is that type better than (BioProtech)?
As long as you are using disposable medical grade electrodes and not something homemade, it should be fine.
 
  • #46
Screenshot of the circuit that I connected on Breadboard, I hope it's clear.
 

Attachments

  • circuit.PNG
    circuit.PNG
    34.2 KB · Views: 407
  • Like
Likes   Reactions: Tom.G
  • #47
This is screenshot of the activity of my bicep muscle (Resting and Contraction)
 

Attachments

  • bicep1.JPG
    bicep1.JPG
    46.5 KB · Views: 408
  • Like
Likes   Reactions: dlgoff and jim hardy
  • #48
216171-578fd36e16795062cf1e7f973f69ae21.jpg


WOW! You did it! Congratulations.

Can you post the details and schematic of what you did?

Cheers,
Tom
 

Attachments

  • 216171-578fd36e16795062cf1e7f973f69ae21.jpg
    216171-578fd36e16795062cf1e7f973f69ae21.jpg
    1.9 KB · Views: 757
  • Like
Likes   Reactions: dlgoff, Mo_Tuk and jim hardy
  • #49
Tom.G said:
View attachment 233125

WOW! You did it! Congratulations.

Can you post the details and schematic of what you did?

Cheers,
Tom

Thanks @Tom.G
I will post the schematic, now I need to know how to get data from the oscilloscope so that I can deal with it using Arduino, the oscilloscope used is "Tektronix TDS 2024B" and I just used Print button to capture the signal.
 
  • Like
Likes   Reactions: jim hardy
  • #50
The attached image is the schematic for EMG circuit that worked, and I have question about DC level shifter circuit should I add it ??
so that (e.g. the output will be from 0 to 5V max), because Analog to Digital Converter "ADC" of the arduino can only accept this range.
any Suggestion ...
 

Attachments

  • Capture.PNG
    Capture.PNG
    20.6 KB · Views: 388
  • #51
Mo_Tuk said:
The attached image is the schematic for EMG circuit that worked, and I have question about DC level shifter circuit should I add it ??
so that (e.g. the output will be from 0 to 5V max), because Analog to Digital Converter "ADC" of the arduino can only accept this range.
For your particular use of detecting different muscle activity you would not need a level shifter.

However the converter on the Arduino board may not like a negative-going signal. I would say try connecting to the Arduino to see if you get reasonable results, if the results are bad then add a level shifter. (Unless of course you want to add a level shifter as part of learning more -- often a good idea. :wink:)

Cheers,
Tom
 
  • Like
Likes   Reactions: Mo_Tuk, dlgoff and jim hardy
  • #52
Hello

I added the dc lavel shifter and the circuit works fine with arduino using ADC of the arduino, but now I have to do two things:

1) comparing the result with other built-in EMG shield by using correlation technique "Auto-Correlation"
2) add some realistic application to prove that the system is practical, any idea ??
 
Last edited:

Similar threads

Replies
6
Views
2K
Replies
8
Views
3K
  • · Replies 9 ·
Replies
9
Views
5K
  • · Replies 6 ·
Replies
6
Views
2K
  • · Replies 6 ·
Replies
6
Views
6K
  • · Replies 3 ·
Replies
3
Views
3K
Replies
1
Views
2K
  • · Replies 6 ·
Replies
6
Views
3K
  • · Replies 1 ·
Replies
1
Views
1K
  • · Replies 3 ·
Replies
3
Views
2K