Question about Designing EMG data acquisition
  • Thread starter Thread starter Mo_Tuk
    Data Designing
The discussion focuses on designing an EMG data acquisition system, particularly the selection of gain and filter types for signal processing. An instrumentation amplifier is recommended for amplifying EMG signals, with the gain determined by the peak voltage before amplification and the input specifications of the ADC. A band-pass filter is suggested to eliminate unwanted frequencies, with considerations for ripple, phase lag, and filter order influencing the choice of filter type, such as Sallen-Key or Butterworth. The importance of practical experimentation and adjustments is emphasized, along with the need for a differential input amplifier to reduce electrical noise. Overall, the design process involves iterative testing and refinement to achieve accurate classification of hand movements.
Mo_Tuk
Hello
I'm working on project called"design & implementation of EMG data acquisition",after I search on these systems i really confused about what the gain and the filter type that i have to use.
at the first stage i use instrumentation amplifier to amplify the signal, so how much the gain of this stage should be? and i use Band-Pass Filter in the second stage which is eliminating the unwanted frequency components, which type of filter is suitable to get a very clean EMG signal? Is sallen-key or butterworth good enough for that? and how much should the order of filter be?
the attached image for my design for now in Multisim, I will add proper ADC to it...thanks...
You work from both ends to the middle. You need the peak voltage before your input amplifier. You need the input voltage specification for your analog/digital converter. Divide. That gives you the gain you need. Then you need the source impedance of your input signal, from which you calculate the minimum input impedance of the input amplifier.
Then you define the requirements of your filter. Allowable ripple in the passband, maximum allowable gain in the stopband, minimum range of passband frequencies, maximum range of stopband frequencies, allowable phase lag, allowable phase variation. That will give you the minimum rolloff (attenuation rate), from which you can find the minimum filter order. The gain, ripple, and phase specifications will determine the best type of filter.
A Google search using search terms analog filter had many good hits. The hits from Sensors Online and Analogue Devices (analog.com) are a good place to start.
Likes davenn
I searched on this phrase
and found this surprisingly informative article by an outfit that manufactures such gear.
https://delsys.com/Attachments_pdf/download/tutorials/01_fundamental-concepts-in-emg-signal-acquisition.pdf
Another hit was this sensor
http://www.biometricsltd.com/surface-emg-sensor.htm
which has built in gain of 100 or 1000.
Does your instrumentation amplifier have to include that gain, or are you handing it signal from such a sensor ?
i think you'll have to resort to trial and error with whatever sensors you've picked and whatever is your subject...
One experiment is worth a thousand expert opinions.
If this is for a product development effort, check out http://www.ti.com/product/ADS1298
Likes Mo_Tuk, dlgoff, jrmichler and 1 other person
This applies at a hobbyist or "I'm curious" level. As such it is simplified and is incomplete for serious research usage.
I built one way back when I was in high school (grade 10 or 11). I strongly recommend to use a differential input amplifier to suppress the electrical noise from the electrical power lines and from fluorescent lights. EMG signals are essentially 0.5 to 1mS spikes of around 50μV to 1mV at the skin surface and can exceed 50mV with electrodes embedded in the muscle. The repetition rate and amplitude increase with increased muscle activity.
Depending on the usage environment, you may need:
  • Low frequency gain rolloff, or even a notch filter to suppress the power line hum.
  • To suppress the power line third harmonic frequency if it is a problem, probably with a notch filter as it is getting into the signal bandwidth.
  • A lower first stage gain to avoid overloading from the power line noise. Worst case would be a passive filter before the first amplifier, but that should be a last resort.
  • Low noise shielded cable from the electrodes. Most wires and shielded cables will generate a tiny voltage when moved/bent. Even Oscilloscope probes have this problem. Low noise microphone cable is available to minimize this. There may be other cable sources too.
What is the intended usage of your device?
Hope this helps.
Tom
Likes berkeman, dlgoff, jrmichler and 2 others
Hope this helps.
Tom
Thanks for your response, the goal of my project is that after complete designing this system, I have to classified four different hand movements like up,down,right,and left, so could this possible with and INA114 "instrumentation amplifier" and 1st order band-pass filter ?
please, if you could see the screenshot that I've attached of my design and tell me if this good enough to reach the goal of different hand movements?
Both @jim hardy and @Tom.G told you that you must think of things beyond the schematic on your screen. Not all problems can be solved using only a schematic design tool.
I think you need more direct help from an advisor or a mentor rather than written questions/answers on the Internet. Are there teachers or other students there who can help you?
Likes berkeman and Mo_Tuk
I never built one of those so have no idea.
I do know that arriving at a successful design is trial and error,
We learn by doing not by reading about doing, e.g. the Wright Brothers...
My advice is build it.
1. Make gain adjustable perhaps R1 a 5K rheostat?
That way if noise is overwhelming you can turn down the gain and look at it with the 'scope.
2. If noise is high try a driven shield as in fig 10 of that datasheet @Tom.G linked. (EDIT i thought he linked it - anyhow it's at http://www.ti.com/lit/ds/symlink/ina114.pdf )
Likes Tom.G, jrmichler, Mo_Tuk and 1 other person
Hi @Mo_Tuk,
I think what you have done so far is a fair start.
In the spirit of helping you to learn, you should build the circuit you designed and find out how well it works. If there are problems look at the ideas in my post #4 and in post #7 by @jim hardy to help you figure out what is needed. There is no need to add an ADC at this point, that will only make things harder to understand. Use an oscilloscope to look at the output of the circuit.
As @anorlunda said, an advisor, teacher, or other students there to actually see what you have built and what it is doing can help much faster and better than we can. Of course if you and your helpers need extra help, we are still here to advise.
Don't get discouraged. As Jim said:
Please let us know about your progress, it sounds like an interesting project.
Cheers,
Tom
Likes jim hardy and Mo_Tuk
Tom.G said:
Hi @Mo_Tuk,
Don't get discouraged. As Jim said:Please let us know about your progress, it sounds like an interesting project.
Cheers,
Tom
Thanks a lot, off course I will give you a feedback of any progress I do, now I'm interesting only on simulation because in this month I can't reach to the instruments such as oscilloscope but I will try to get one these days.
Likes Tom.G and jim hardy
mine is an antique Ebay find HP180. Real simple .
https://www.ebay.com/itm/HP-Hewlett-Packard-180-C-OSCILLOSCOPE/123304604812?hash=item1cb586f48c:g:UZ0AAOSwElZanHR0
Likes Mo_Tuk
I started to build the circuit and make some tests but Should I build protection circuit before connecting the electrodes to instrumentation amplifier ?
i don't think you'll need it. What killed early ECGs was defribillators.
Of course use good sense -
power it from batteries or a power supply with medical grade isolation , that's for the safety of your subject
and before hooking it up touch both it and whatever (whomever?) you're connecting the electrodes to so there's no static charge.
old jim
Likes berkeman, Mo_Tuk and dlgoff
Hi @jim hardy
I need to use only one battery to supply the INA114 with +4.5V as reference. in the design in multisim I connected +9V(Vdd) and +4.5V(Ref pin) of INA and Vss to ground. and the input signal is 10mV.
the gain is 101 and the output voltage should be 1.01V centered at +4.5V, But the result is just dc voltage at +4V (red line) and not sinusoidal. the question is: How to get a correct sinusoidal output centered at +4.5V ??
I hope you understand me.
Cheers.
Mo_Tuk said:
I need to use only one battery to supply the INA114 with +4.5V as reference. in the design in multisim I connected +9V(Vdd) and +4.5V(Ref pin) of INA and Vss to ground(? jh). and the input signal is 10mV.
Been aaway a few days, wasn't ignoring you.
it is not clear to me what you connected to the REF pin.
The REF pin should be connected to a constant source of about elf the difference between V+(Vdd) and V-(Vss) pins.
If you connected REF pin to +4.5 volts that is correct and the circuit ought to work.
If you connected the REF pin to V-(Vss pin) that won;t work.
The fellows who wrote the datasheet have a mindset that it's being used with a balanced dual supply, where GND is halfway between the V+ and V- "supply rails".
GND is NOT V- in their minds.
But that might not be your problem.
Did i ask my question clearly?
If indeed you connected REF to 4.5 volts,
then
try reducing GAIN to 1 and see if the circuit will reproduce a 1 volt differential signal. Output will appear between OUT and REF pins.
They're telling you right here that input voltage must remain a couple volts above V- and a couple volts below V+
Likes Mo_Tuk
Thanks @jim hardy
I really appreciate your effort, you helped me a lot.
I will supply the INA with + and - 9V and about the reference pin as in datasheet Is that mean I Should connect 5 ohm resistor to it ?
Good ! The junction of your +9 and - 9 volt supplies becomes "GND " ..
...and about the reference pin as in datasheet Is that mean I Should connect 5 ohm resistor to it ?
No, the datasheet is telling you that it takes only as little as 5 ohms to cause you trouble.
REF needs to be held at mid-supply voltage.
It is important that the two resistors i circled in red be very closely matched. Any resistance in series with REF pin adds to the bottom resistor making them unequal.
REF needs to be held near mid-supply voltage. Midpoint of your dual supply is the best place to connect it.
You would do well to master the "Difference Amplifier" for that's what A3 is. See http://hyperphysics.phy-astr.gsu.edu/hbase/Electronic/opampvar6.html#c3 , or any number of tutorials from a search.
and study figure 4 of the INA114 datasheet.
Keep up the good work -
old jim
Likes Mo_Tuk
Thanks @jim hardy
I will use battery to avoid any issues of power line (I mean 60 Hz noise)
but if I use function generator (+9,-9v) is there any possibility to harm my body or not ?
Wait, what? Where are you going to connect this to your body? I haven't been following this thread for a while, but just saw your post. Why are you trying to drive a signal into your body? That is not what EMG sensors do. Hopefully I've misunderstood your post. Thanks.
you misunderstood,
I mean I will use battery to supply my circuit so that when I connect the electrodes to my arm I hope no leakage current exists and I will be in safe side.
Likes berkeman
Not so long as the function generator's insulation system is in good shape,
and so long as the wall outlet that you plug it into is wired correctly, is properly earthed, and has nothing else plugged into it that has an internal fault.
So yes, there is a possibility. .
Likes Mo_Tuk
Has anyone defined what is meant by EMG, Electromyography? Are you sensing electrical muscle RMS noise directly, or are you able to monitor the electrical changes that result from a physically active muscle in an elastic-resistive glove.
Not much will have changed on the signal acquisition side since I was designing polygraph components in the early 1980s. The instrumentation amplifiers will use less power, have better balance with greater CMRR and much lower drift. Only the anti-aliasing filters should be integrated into the amplifier chain. Once the signal is digitised, the problem will become software estimates of signal power in a particular bandwidth determined by signal processing software. That needs to be in software to be flexible.
You may need to consider ways to measure all signals as differential against each other. That may help eliminate problems due to contact variation and reduce gain errors. Look for common electrode positions that can be used for sensing several muscles.
Safety regulations will probably require an isolation barrier between your electronics and the human body. Isolation amplifiers are now available for that job. We used back-to-back depletion mode FETs as current limiters.
Likes jim hardy
Hi @jim hardy
i connected the circuit (pre-amplification stage using INA114 + Filtering stage of BPF) and then connected the electrodes to my bicep muscle and try to detect some signal as shown in attached screenshots on the oscilloscope. I set the gain of INA to 2 "next time the gain will be 10" and the BPF has a gain of 100 and frequency band from about 20 to 700 Hz
I used the inputs of INA114 directly to the electrode with no isolation circuit ,Should I add isolation circuit or not since I supply my circuit with two 9v batteries?
Mo_Tuk said:
Hi @jim hardy
Looks mostly like 60Hz noise...
Likes Mo_Tuk
can I use notch filter to remove 60Hz noise?
It's better to design your circuit to reject that common-mode noise (hopefully it's not getting into the circuit in differential mode)...
Thanks, @berkeman
for the screenshot of INA, what the benefit of using two 15k resistors?
They look like they are just supplying the bias voltages for the two inputs.
Can you post a diagram of your complete circuit, including the battery power supply, amp circuit, patient connections, and Earth Ground scheme?
Likes Mo_Tuk
I connected the circuit exactly as shown in screenshot:
If we're talking about the image attached to that post,
the two 15K resistors establish a reference that's halfway between the amplifier's supply voltage pins.
It's known as "virtual ground"
and if you're powering from two 9v batteries you could instead use for reference their junction.
nitpicking here - the noise is line frequency which for us is 60 hz but apparently for you 50 hz,
old jim
Likes Mo_Tuk
old jim
Thanks @jim hardy
do you mean that I can just connect the circuit as shown in the screenshot below.
and about the 50Hz noise, How could I remove it ? and also do I have to add isolation circuit ?
  • #31
I don't think you need it . You have no home-made equipment in between you and wall power that's likely to cause a short circuit. Your oscilloscope chassis should be tied to the the Earth grounding plug of your power cord. I would suggest check your outlet with an inexpensive receptacle checker, here's what they look like in USA :
Remaini aware you are relying on the internal insulation of any line connected equipment to protect you, so check it carefully for leakage current to Earth ground.
Try touching a moistened finger on your other hand to oscilloscope common (outside of the BNC connector if it's metal).
If that reduces noise then you know you need a reference connection to your body.
Not quite.
That screenshot shows a wire shorting out your signal. Take it out of your sketch and connect REF to power supply common.
old jim
  • #32
Your circuit that you posted is very confusing for me. I asked to see the circuit that you said you connected to your bicep, but your schematic shows instruments, not a pair of patient connections as input. I also don't see Earth ground, and I see very different biasing compared to your latest differential amplifier questions.
Are you using a single 9V battery for power, or two to give you +/-9V? Please always show your power supply configuration in your schematics. What electrodes were you using to connect to your bicep? What brand, and how old are they? I assume they are single-use disposable, right?
When you flexed your bicep, did you see any response at all out of the amplifier string? It should have been much lower frequency than your 50/60Hz interference.
This is more like a circuit diagram that I would expect you to be posting...
https://greenhecktech.files.wordpress.com/2016/09/schematic4.png?w=748
Likes dlgoff and jim hardy
  • #33
Yes, I'm using two +/-9V batteries to supply the circuit, and I used surface electrodes "BioProtech Electrodes" which shown in attached image.
Although the signal is 50Hz noise but when I flexed my bicep the amplitude increased but the signal still almost the same "50Hz noise"
Likes berkeman
  • #34
Okay, there are two things that you should approach this with, IMO:
  • You should look more at shielding and balancing your probe lines to cut down on the AC Mains noise pickup -- if you post your full circuit diagram including Earth ground connections and shielding, we should be able to help with that
  • You should consider digitizing the waveform output from your circuit and doing DSP processing to filter out the AC Mains noise. You can do much higher order filters and adaptive filtering with DSP routines. Do you have experience with programming microcontrollers (uCs) in C or Assembly code?
Likes dlgoff, jim hardy and Mo_Tuk
  • #35
Thanks @berkeman
I hope this screenshot of the circuit would be clear, and about the DSP part for now I just want to see clear EMG signal from my bicep muscle and then I will move to programming part using Arduino.
and about the electrode cables would you suggest any specific type of cable that less noise produced ??
  • #36
Hi Mo_Tuk,
Could you post a photograph of the assembled amplifier and another photograph of showing it connected to you? Please show where the leads connect to you and to the amplifier. With only the wiring diagram, it is very hard for us to guess what may help to get it working.
The waveform you posted earlier (https://www.physicsforums.com/attachments/tek0003-jpg.232273/) looks like there is a DC power supply nearby that is making some electrical noise, maybe a battery charger or computer, or maybe even an LED lamp!
Cheers,
Tom
edit: fixed typo
  • #37
we trust you tied the junction of your batteries to circuit common
and shielded your signal wires
i added those features to your drawing
Likes Mo_Tuk
  • #38
Tom.G said:
Hi Mo_Tuk,
Could you post a photograph of the assembled amplifier and another photograph of showing it connected to you? Please show where the leads connect to you and to the amplifier. With only the wiring diagram, it is very hard to for us to guess what may help to get it working.
Cheers,
Tom
Thanks @Tom.G
do yo mean photograph of the circuit connected on the breadboard??
and about the electrical noise I think the laptop was near to my circuit and plugged into charger.
  • #39
jim hardy said:
we trust you tied the junction of your batteries to circuit commonand shielded your signal wires
Thanks @jim hardy
for the connection you made of electrodes A&B to reference, would that help to reduce noise or something else??
  • #40
The red dashed lines do NOT represent connections to the electrodes,
instead
they represent shields around the wires bringing signal from the electrodes to your circuit board.
The purpose of a shield is to physically surround the signal wire so the can be no capacitve coupling into it.
Recall from basic physics that no E field can exist inside a closed conducting surface - google "Faraday Shield".
Using a twisted pair to route both signals together will reduce magnetic coupling.
That's why "twisted shielded pair" is so popular.
Your 50 hz noise is probably a combination of pickup that's
capacitively and inductively coupled into the input wires ( and gets amplified 1100X) ,
and
capacitively coupled across the power transformer in your oscilloscope's power supply.
Shielding the input wires will help .
old jim
Likes Mo_Tuk
  • #41
Emphasizing @Tom.G 's comment"Driven Shield" is a method that's used in precision equipment .
It's shown in the INA111 datasheet
the cylinder surrounding the signal wires represents a shield that is driven to nearly average voltage of the two signals
which minimizes voltage between the shield and signal wires
minimizing capacitive coupling of current into the signal wires.
INA111's input impedance is so high that your signal wires act like an antenna .
Shielding hides them from electric fields by creating a Faraday Cage.
Twisting them reduces their sensitivity to magnetic fields by reducing the area(hence the magnetic flux) they encircle.
That's why we EE's like twisted shielded pair for signal wires..
Try grounding shields first to see if it helps. Driven shield should give even better results.
old jim
Likes Mo_Tuk and dlgoff
  • #42
Yes.
Likes Mo_Tuk
  • #43
Belden 2221 or 11700A would be good choices but will be hard to find and the 2221 costs about $2 to $3 USD per foot ($6 to $9.3 USD per meter). A little bit cheaper if you buy 1000 feet (330 meters).
I suggest you first try whatever shielded wire you can get and twist the wires together, about 3 or 4 twists per foot (10 per meter) should help.
A possible source is: https://www.nassauelectrical.com/pr...ingle-pair-bc-stranded-30x40-pvc-jacket-cable
Likes Mo_Tuk
  • #44
I have also another type of electrodes (Covidien Kendall Q-Trace Diagnostic Tab Electrodes Resting ECG Electrodes) , is that type better than (BioProtech)?
  • #45
As long as you are using disposable medical grade electrodes and not something homemade, it should be fine.
  • #46
Screenshot of the circuit that I connected on Breadboard, I hope it's clear.
Likes Tom.G
  • #47
This is screenshot of the activity of my bicep muscle (Resting and Contraction)
  • #48
WOW! You did it! Congratulations.
Can you post the details and schematic of what you did?
Cheers,
Tom
Likes dlgoff, Mo_Tuk and jim hardy
  • #49
Cheers,
Tom
Thanks @Tom.G
I will post the schematic, now I need to know how to get data from the oscilloscope so that I can deal with it using Arduino, the oscilloscope used is "Tektronix TDS 2024B" and I just used Print button to capture the signal.
Likes jim hardy
  • #50
The attached image is the schematic for EMG circuit that worked, and I have question about DC level shifter circuit should I add it ??
so that (e.g. the output will be from 0 to 5V max), because Analog to Digital Converter "ADC" of the arduino can only accept this range.
any Suggestion ...