SR830 Lock In Amplifier Not Reading Reference Signal
The SR830 lock-in amplifier is not detecting the reference signal, which is confirmed by the "Unlock" light indicating a 1 kHz frequency, while the oscilloscope shows the correct signal at 336 Hz. The reference input voltage was measured at only 140 mV, below the required minimum of 400 mV peak-to-peak, leading to a switch in the detector to achieve the necessary voltage. Suggestions for stabilizing the output voltage include reducing noise, increasing the input signal amplitude, adjusting the sensitivity range, increasing the time constant, and using input filters. Proper setup and signal conditions are crucial for the lock-in amplifier to function effectively. The discussion highlights the importance of ensuring adequate reference signal levels for successful operation.
dkuljis
So here is my dilemma:
We have our SR830 in a MOKE set-up. The modulation of the laser is with a chopper at approximataly 336 Hz. The reference signal is not being picked up by the SR830, but using the same cable the reference frequency is coming up on our oscilloscope. ie, plug the cable into the oscilloscope or DMM and we get the appropriate frequency. Plug it into the SR830, and the "Unlock" light is on with a frequency of exactly 1 kHz, which my adviser says it means it's not getting a signal. The rest of the SR830 seems to be in good working order. Both CH1 and CH2 are picking up signals from our photo detector. Any thoughts or suggestions?
Sorry if the post is skim on details, I'm new to all this.
Thanks for your time,
Danny
"The SR810 and SR830 Lock-In Amplifiers have differential inputs with 6 nV/√Hz input noise. The input impedance is 10 MΩ, and minimum full-scale input voltage sensitivity is 2 nV. "
WOW that's some instrument.
The reference input needs to be 400mV pp minimum. You didn't mention what the voltage is.
Has the setup ever worked?
Does the instrument work in other setups with an external reference?
Can you drive the external reference with a signal generator to test its acquisition range?
Just guessing.
Thanks meBigGuy. We measured the voltage on the reference signal and it was only around 140 mV so we switched our detector and got the reference frequency where it should be. Thanks a bunch for the quick response!Sent from my iPhone using Physics Forums
I am measuring voltage output from the pick-up coils of a vibrating sample magnetometer using the SR830 lock-in amplifier. The problem is that the channel 1 output voltage cannot be stabilized at all. I read the manual but that is not helpful either. What can I do to stabilize the reading ?
*Reduce the amount of noise in your measurement setup. Lock-ins can "cut through" quite a lot of noise, but there are always limits and this should always be the first step
*Increase the amplitude of the input signal
*Make sure you have the SR830 is on the optimal sensitivity range
*Increase the time constant
*Turn on the filters on the input.
Generally speaking, there will always be some trade-off between measurement speed and sensitivity; if your noise is mainly white you can detect some very small signals with a lock-in if you increase the time constant.
Likes Sakin_EEE and berkeman
While I was rolling out a shielded cable, a though came to my mind - what happens to the current flow in the cable if there came a short between the wire and the shield in both ends of the cable? For simplicity, lets assume a 1-wire copper wire wrapped in an aluminum shield. The wire and the shield has the same cross section area. There are insulating material between them, and in both ends there is a short between them. My first thought, the total resistance of the cable would be reduced...
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...