Why frequency-weight gauss meters for human-bias?

  • Thread starter Thread starter semiotically
  • Start date Start date
  • Tags Tags
    Gauss
Click For Summary

Discussion Overview

The discussion revolves around the comparison of frequency-weighted and flat gauss meters, particularly in relation to their effectiveness in measuring electromagnetic fields (EMFs) from everyday electrical appliances and their potential impact on the human body. Participants explore the implications of different frequency responses and their relevance to human exposure to AC magnetic fields.

Discussion Character

  • Technical explanation
  • Debate/contested
  • Conceptual clarification

Main Points Raised

  • Some participants suggest that the standard TriField Meter's weighted frequency response is designed to correspond to the current induced in the body, with higher frequencies resulting in greater induced current.
  • Others explain that the flat TriField meter measures energy levels independently of frequency, which may be more suitable for applications not related to human exposure.
  • A participant questions the implications of measuring energy independent of frequency, indicating a need for clarification on the relevance of frequency in these measurements.
  • Some participants express curiosity about the reasons behind the choice of 50/60 Hz for AC mains power and its potential effects on human exposure to EMFs.
  • Another participant seeks to contrast the qualities of frequency-weighted versus flat gauss meters specifically in the context of measuring EMFs from household appliances.

Areas of Agreement / Disagreement

The discussion contains multiple competing views regarding the effectiveness and implications of frequency-weighted versus flat gauss meters. Participants have not reached a consensus on the best approach or the significance of frequency in these measurements.

Contextual Notes

Participants have not provided detailed background information or clarified specific assumptions regarding their questions, which may affect the understanding of the topic. There are unresolved aspects related to the impact of different frequencies on induced currents in the human body.

semiotically
Messages
16
Reaction score
0
I have read two ideas:

1. is that the 50/ 60Hz everyday home power supplies encourage this
2. that the higher frequencies are disproportionally (greater) 'absorbed' than the lower...

Please point me down the right track

Thank you,

Sem
 
Engineering news on Phys.org
TriField Flat verses Weighted Frequency Response


Weighted: The standard TriField Meter has a weighted frequency response that corresponds to the
potential current that would be induced into a body. As the frequency doubles, the amount of current
induced onto the body, over the same period of time, also doubles. The center frequency is 50Hz or 60Hz
specified at the time of order to correspond to the line frequency in your area/country.


If the 60Hz meter is in the presence of a 60Hz AC magnetic field with the energy level of 10 milligauss.
The meter will register 10 milligauss. If the same trifled meter is in the presence of a 120Hz AC magnetic
field with an energy level of 10 milligauss, the meter will register 20 milligauss, indicating the relative
induced energy impact on the body.


Flat: The flat TriField meter measurers the energy level present independent of the frequency of the
energy, provided that the frequency of the energy is between 30Hz and 1000Hz. This meter is
recommended for applications that are not related to the potential induced energy impact on a human body.



Frequency of a AC
Magnetic Field with a
level of 10 milligauss

Standard TriField 60Hz
Meter Response

Flat TriField Meter
Response

30Hz 5 Milligauss 10 Milligauss
60Hz 10 Milligauss 10 Milligauss
120Hz 20 Milligauss 10 Milligauss
240Hz 40 Milligauss 10 Milligauss
480Hz 80 Milligauss 10 Milligauss



Frequency of a AC
Magnetic Field with a
level of 10 milligauss

Standard TriField 50Hz
Meter Response

Flat TriField Meter
Response

30Hz 6 Milligauss 10 Milligauss
50Hz 10 Milligauss 10 Milligauss
100Hz 20 Milligauss 10 Milligauss
200Hz 40 Milligauss 10 Milligauss
400Hz 80 Milligauss 10 Milligauss
 
...measure energy independent of its frequency..?
 
semiotically said:
I have read two ideas:

1. is that the 50/ 60Hz everyday home power supplies encourage this
2. that the higher frequencies are disproportionally (greater) 'absorbed' than the lower...

Please point me down the right track

Thank you,

Sem

Welcome to the PF. Could you please post more background information for your question? What are you asking? Are you asking why 50/60 Hz is chosen over some other frequency for AC Mains power distribution? Or are you asking about something having to do with shock currents in humans that come in contact with AC Mains voltages?
 
I'm trying to contrast frequency-weighted with flat gauss meters and assess their qualities in measuring everyday electrical appliance EMFs that can induce current within the body..?
 

Similar threads

Replies
4
Views
3K
Replies
2
Views
2K
  • · Replies 1 ·
Replies
1
Views
2K
Replies
24
Views
10K
  • · Replies 13 ·
Replies
13
Views
823
  • · Replies 14 ·
Replies
14
Views
2K
Replies
4
Views
3K
Replies
17
Views
6K
Replies
4
Views
7K
  • · Replies 11 ·
Replies
11
Views
6K