Thickness of lead required for 200 keV x-rays?

In summary, the "200 keV" refers to the energy level of x-rays, which is important in determining the appropriate thickness of lead for shielding purposes. The relationship between lead thickness and x-ray energy level is an inverse one, with higher energy levels requiring thicker lead. The thickness of lead needed for 200 keV x-rays can be calculated using the mass attenuation coefficient of lead at this energy level, which can be found in tables or calculated using specialized software. Other factors that can affect the required thickness of lead include the distance between the x-ray source and the lead barrier, the angle of incidence of the x-rays, and the type of lead used. Finally, the thickness of lead needed cannot be determined using a standard formula for all
  • #1
ireland01
17
0
is this correct?

mass attenuation coeff. = 8.636 cm^2/g

8.636 * (density of lead: 11.3 g/cm^3) = 975.868 per cm

thickness (mm):

t= 9.21 / 975.868

= 0.009 cm

= 0.09 mm
 
Physics news on Phys.org
  • #2
Where does the factor of 9.21 come from?
Assuming your coefficient is right, and you want 9.21 radiation lengths, the answer looks right.
 
  • #3
I found that factor online.It said to get intensity to < 1% but by my calculations, for 200 KeV beam using formula:
HVL = 0.693/u

need 7 HVL layers to get to 0.8% which is 4.48 mm.
where u = 10.15 cm^-1
 
Last edited:
  • #4
4.5mm looks better, and the new prefactor 7*ln(2) looks better as well.
 
  • #5


I cannot definitively say if this calculation is correct without knowing the specific context and assumptions being made. However, the formula used for calculating the thickness of lead for 200 keV x-rays appears to be based on the mass attenuation coefficient and density of lead, which are both commonly used in radiation shielding calculations. Therefore, the calculation may be a reasonable estimate, but it would be best to double check the formula and assumptions being used to ensure accuracy. Additionally, other factors such as the type and energy of the x-rays, the desired level of protection, and the specific application should also be considered when determining the appropriate thickness of lead for shielding against 200 keV x-rays.
 

1. What is the significance of the "200 keV" in this question?

The "200 keV" refers to the energy level of the x-rays. In order to determine the appropriate thickness of lead, we need to know the energy level of the x-rays because this affects their ability to penetrate materials.

2. What is the relationship between lead thickness and x-ray energy level?

The relationship between lead thickness and x-ray energy level is an inverse one. The higher the energy level of the x-rays, the thicker the lead needs to be in order to effectively shield against them.

3. How do you calculate the thickness of lead needed for 200 keV x-rays?

The thickness of lead needed for 200 keV x-rays can be calculated using the mass attenuation coefficient of lead at this energy level. This coefficient can be found in tables or calculated using specialized software. The thickness can then be determined using the equation: thickness = (ln10) / (μ/ρ) where μ/ρ is the mass attenuation coefficient and ln10 is the natural logarithm of 10.

4. Are there any other factors that can affect the required thickness of lead for 200 keV x-rays?

Yes, there are other factors that can affect the required thickness of lead for 200 keV x-rays. These include the distance between the x-ray source and the lead barrier, the angle of incidence of the x-rays, and the type of lead used (e.g. pure lead or lead alloys).

5. Can the thickness of lead be determined using a standard formula for all energy levels?

No, the thickness of lead required for x-rays of different energy levels cannot be determined using a standard formula. The mass attenuation coefficient of lead varies depending on the energy level of the x-rays, so the calculation for thickness will differ accordingly.

Similar threads

  • Advanced Physics Homework Help
Replies
5
Views
963
  • Introductory Physics Homework Help
Replies
3
Views
1K
  • Nuclear Engineering
Replies
7
Views
2K
  • High Energy, Nuclear, Particle Physics
Replies
7
Views
423
  • Nuclear Engineering
Replies
3
Views
1K
  • High Energy, Nuclear, Particle Physics
Replies
1
Views
747
  • Advanced Physics Homework Help
Replies
1
Views
1K
Replies
1
Views
1K
Replies
6
Views
332
  • Introductory Physics Homework Help
Replies
2
Views
747
Back
Top