PIN Photodiode, High Voc, Low Isc
  • Thread starter Thread starter Qianlong
    Photodiode Pin
The discussion centers on the performance of a silicon PIN photodiode in photovoltaic mode, focusing on the relationship between open-circuit voltage and short-circuit current under varying light intensities. Measurements indicate that while open-circuit voltages range from 40μV to 350mV, short-circuit currents only appear above 250mV, with a measured current of 1.4μA at that threshold. The participants explore the reasons for the low current output, suggesting that the internal resistance of the measuring multimeter may significantly affect readings. Additionally, the concept of barrier potential in semiconductor junctions is discussed, highlighting that a voltmeter cannot measure this potential due to current flow until equilibrium is reached. Overall, the findings raise questions about the expected current levels in ambient light conditions and the underlying physics of the photodiode's operation.
Qianlong
I am working with a silicon PIN-type photodiode in photovoltaic mode, conducting measurements of open-circuit voltage and short-circuit current for varying intensities of incident light. The light is being measured as a proxy for another quantity, and so I must apologise for I can give no quantitative values of optical power.
The photodiode has a claimed max “Operating Current” of 10mA, and an active area of 20mm^2.
The measuring apparatus is a fluke 187 digital multimeter, which can measure to precisions of tens of microvolts and tens of nanoamps, though in the setup there is some background noise (probably electrical) that renders the last digit unreliable. Attempts have also been made to measure current with a moving coil microammeter, which has a resistance of 150Ω.
As I understand it, with increasing optical power incident on the detector, the short-circuit current should increase linearly and be equal to the photocurrent. The open-circuit voltage should increase logarithmically, but for low optical powers, the increase is approximately linear*.
*The open circuit voltage is given by the equation V = KT/q ln[(I/Io +1)], which for values of I/Io<<1 the Mercator series can be used to yield V=KT/q I/Io
Open-circuit voltages have been detected at both at ultra-low light levels and in ambient environmental lighting. The open-circuit voltage s in these scenarios has ranged from 40μV to 350mV. Linearity is confirmed up until about 20mV. However, when the short-circuit current is measured, nothing is seen until the open-circuit voltage for the same light intensity exceeds 250mV. At this light intensity, the current is 1.4μA.
My question: is the current just that small? Other measurements (albeit for p-n junctions) I have seen reported in the literature record that light intensities that lead to 1mV open-circuit voltages lead to at least 1μA short-circuit currents. Certainly, one would have thought that a current much closer to 10mA would have been seen in ambient environmental lighting.
There is a second question I wished to ask, and it relates more to the theory. I understand that when p-type semiconductor is connected with n-type semiconductor, the majority carrier in each type diffuses along the concentration gradient. This leave behind ions, trapped in the crystal lattice, which generate an electrical field that opposes further diffusion. Forgive my ignorance, but why can this “barrier potential” not be read by a voltmeter placed in parallel around the junction?
Greatly obliged for your assistance.
The fluke multimeter has some internal resistance. I don't know how much (and it depends on the measurement range: smaller ranges correspond to larger internal resistances), but that could reduce the current significantly.
Unrelated to that, at the level of millivolts you can have temperature effects that lead to additional voltage differences.
As far as I know, 1.4µA is not at the lower end of the fluke precision - do you see a sudden jump in the current?
Current flows through the voltmeter (and the timescale is way too short to see that) until the potential is equal at both sides. That is true for every setup without a power supply.
Qianlong said:
the open-circuit voltage for the same light intensity exceeds 250mV. At this light intensity, the current is 1.4μA. My question: is the current just that small?
Yes, that actually sounds quite reasonable.
The key to understanding this is to realize just how small I_0 is. It's value is very temperature dependent (due to its dependence on N_i^2), but even at I=1.4uA the ratio I/I_0 is likely to be order of around magnitude 10^5 or so.
Put this value into the equation for voltage and you'll find that it does indeed correspond to somewhere around 250 mV.
Update: I tested that with a Fluke289, and at 1µA, the voltage is well below 1mV (measured: 0.15mV, but that is not reliable*). That is not an issue if the multimeters are not completely different.
*Edit: A more careful measurement showed 0.10mV.
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I have recently moved into a new (rather ancient) house and had a few trips of my Residual Current breaker. I dug out my old Socket tester which tell me the three pins are correct. But then the Red warning light tells me my socket(s) fail the loop test. I never had this before but my last house had an overhead supply with no Earth from the company. The tester said "get this checked" and the man said the (high but not ridiculous) earth resistance was acceptable. I stuck a new copper earth...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...