Accuracy of an Optical Spectrum Analyzer (OSA)

In summary, the datasheet reports a level accuracy of 0.4 dB but this seems quite strange to me. Indeed level measurements are done in dBm thus, as long as dB measures a relative power, I assume that even the 0.4 dB are referred to 1 mW as in the case of dBm. Am I correct? If yes what sounds strange to me is that if I measure a power near to 1 mW than the relative error will be quite high (I'm measuring something like 0 dBm) and as the power that I'm measuring decreases also the relative error decreases.
  • #1
zambadeos
3
0
Hey!
I've a simple (and maybe stupid) question about the accuracy of an OSA. The datasheet report a level accuracy of 0.4 dB but this seems quite strange to me. Indeed level measurements are done in dBm thus, as long as dB measures a relative power, I assume that even the 0.4 dB are referred to 1 mW as in the case of dBm. Am I correct?
If yes what sounds strange to me is that if I measure a power near to 1 mW than the relative error will be quite high (I'm measuring something like 0 dBm) and as the power that I'm measuring decreases also the relative error decreases.
An example to explain my problem:
1) P=0.4 dBm, the relative error is equal to 100%. Expressed in mW means that I'm measuring a power of 1.096 mW
2) P=-65 dBm (the minimum detectable level for my OSA), the relative error is equal to 0.6%. Expressed in watt: [itex]P= (0.316 \pm 0.002)~nW[/itex], that would mean an accuracy of 2 pW: quite amazing!

So where is the bug in my reasoning?

Thanks to everyone
 
Science news on Phys.org
  • #2
Hello
It might be a good idea to give a reference to that particular data sheet. I agree that what you seem to have read is a bit confusing. There should be some indication of the lowest power value it can register, at least.
 
  • #3
Ok, my OSA is MS9710C from Anritsu, you can find the datasheet here (appendix A).
In any case the minimum detectable power should be -65 dBm as i wrote above (for the sake of completeness i have to say that it changes accordingly to the wavelength range in which you are working, but we can focus on that value). Maybe my example is not correct because I'm working at the lower limit of my detector but the conceptual problem is still present even if we consider an higher value.
Moreover I've read datasheets from some other osa and i saw that is quite common to express accuracy in terms of dB, so there must be an error in my reasoning!
In any case thank you for the answer!
 
  • #4
It's been a while since I've used this type of instrument, but the level specification (page A-3) is given as +/- 0.4 dB with an input level of -23 dB. Since the linearity spec is +/- 0.05 dB with input levels between -50 and 0 dB, I expect the accuracy spec to be approximately linear with changing input levels- curse you, log scales!
 
  • Like
Likes zambadeos
  • #5
zambadeos said:
that is quite common to express accuracy in terms of dB,
...and why not? It's only an alternative to percentage for showing the ratio. On a dB scale, it'd really just what you want, I would have thought. You could argue that, in many measurements, the uncertainty is often the same, whatever scale you are using - so it would be more appropriate to use an absolute measure of accuracy. In the end, the choice would depend upon what is being measured and the method being used.
 
  • #6
Thank you very much for the answers.
Andy Resnick you get the solution to my problem, i didn't thought about a linear relation between the accuracy and the input level but it seems to be the right answer :wink::wink:
 

What is an Optical Spectrum Analyzer (OSA)?

An Optical Spectrum Analyzer (OSA) is a scientific instrument used to measure the spectrum of a light signal. It analyzes the intensity of light at different wavelengths, providing information about the optical properties of the source.

How does an OSA measure accuracy?

An OSA measures accuracy by comparing the measured spectrum to a known reference spectrum. This is typically done using a calibrated reference source or a built-in calibration function. The accuracy is then calculated as the difference between the measured and reference spectra.

What factors affect the accuracy of an OSA?

Several factors can affect the accuracy of an OSA, including the resolution of the instrument, the stability of the light source, and the alignment of the optical components. Environmental factors such as temperature and vibration can also impact the accuracy of an OSA.

How can I improve the accuracy of my OSA measurements?

To improve the accuracy of OSA measurements, it is important to use a stable and well-aligned light source. Regular calibrations and maintenance of the instrument can also help maintain accuracy. Minimizing environmental factors and using appropriate averaging techniques can also improve the accuracy of OSA measurements.

What is the typical accuracy of an OSA?

The typical accuracy of an OSA can vary depending on the model and manufacturer. Generally, the accuracy of an OSA is in the range of +/- 0.1-0.5 dB, with some high-end models offering even higher accuracy. It is important to consult the specifications of the specific OSA being used for accurate accuracy information.

Similar threads

  • Other Physics Topics
Replies
2
Views
3K
  • Engineering and Comp Sci Homework Help
Replies
1
Views
4K
  • Advanced Physics Homework Help
Replies
6
Views
2K
  • Electrical Engineering
Replies
4
Views
2K
Replies
46
Views
2K
  • Astronomy and Astrophysics
3
Replies
80
Views
24K
Replies
5
Views
3K
Replies
6
Views
5K
Replies
1
Views
2K
Back
Top