Hello everyone,
I am planning to build a device to test Beer's law for a specific solution and obtain the molar concentration of the solution. Beer's law presumes that the input and output light intensities are related by an exponential law.

I need to know:
  1. The sample path length ##\ell##. Easy to know.
  2. The substance extinction coefficient ##\epsilon##. This is unique for each substance. How can I find it out? What if I had a mixture like water and algae?
  3. The ratio between the input and output light intensities: ##I_{f} / I_{0}##
The molar concentration ##c## is what I am trying to determine.

  • What if the relation between the input and output intensities was not exponential?
  • Also, what about the "wavelength" of the light intensity? The ratio ##I_{f} / I_{0}## may be wavelength dependent while the molar concentration ##c## is the investigated solution is fixed. So which wavelength should we pick?
 

Answers and Replies

  • #2
Borek
Mentor
28,635
3,106
The substance extinction coefficient ϵϵ\epsilon. This is unique for each substance.
And depends on the wavelength. Typical approach is to register a curve of ϵ vs λ and choose the optimal value. Typically the one with the largest ϵ value, as this gives best sensitivity, but sometimes (when we deal with mixtures) we choose other wavelengths to avoid interference.

How can I find it out? What if I had a mixture like water and algae?
Algae is not a substance.

Can't help you with building the spectrophotometer, but it doesn't sound trivial. Light sources and light sensors are not necessarily linear so you either need to choose them very carefully or to compensate, which requires precise calibration.
 
  • #3
1,214
81
Thank you Borek.

Why do we choose the wavelength that is the most absorbed, hence has the largest extinction coefficient ##\epsilon##? Because that gives the most contrast if we change the mixture's concentration ##c##?
 
  • #4
Borek
Mentor
28,635
3,106
As I wrote - it gives best sensitivity. That means both relatively large differences in the signal when changing concentrations and signal becomes detectable at low concentrations, making it possible to analyze more diluted samples.
 
  • #5
mjc123
Science Advisor
Homework Helper
1,085
522
Also because that is where the variation of ε with λ is lowest, so it is less sensitive to errors in wavelength calibration (or small changes in the position of the absorption peak as a function of concentration) than a wavelength on the slope of a peak. (NB You should always run a full spectrum of each sample, to check for consistency - or the accidental presence of contaminants - not just measure ε at a particular wavelength. And of course you don't measure ε, you measure absorbance. Do you know how to translate that into ε for your samples?)
 
  • #6
1,214
81
Thank you mjc123. I see how the extinction factor ##\epsilon## can vary with wavelength. Why is the variation the smallest at the wavelength ##\lambda## with the largest absorption? because it is an absorption peak so in that neighborhood the slope, i.e. the rate of change, is more or less horizontal?
 
  • #7
Borek
Mentor
28,635
3,106
Yes, just a simple math.
 
  • #8
272
208
The Beer-Lambert part of this is simple. Building a fixed frequency device: not simple (but possible, if you can determine the proper freq and can find a compatible lamp/receiver combination) - you'll have to learn a lot about lamps, receivers, temperature regulation, electronics, noise... Building device upon which one can 'run a spectrum': You'll need a lab assistant named Sancho Panza.
 

Related Threads on Beer's law experiment - parameters

  • Last Post
Replies
6
Views
10K
  • Last Post
Replies
4
Views
3K
Replies
5
Views
1K
Replies
2
Views
3K
Replies
9
Views
3K
  • Last Post
Replies
9
Views
3K
  • Last Post
Replies
2
Views
15K
Replies
3
Views
2K
  • Last Post
Replies
4
Views
3K
Top