How can I accurately measure irradiance for my UV sanitization device design?

AI Thread Summary
To accurately measure irradiance for a UV sanitization device, understanding the inverse square law (1/d^2) is essential, as it relates irradiance to distance from a surface. Reliable resources include Wolfe's "Introduction to Radiometry" and various online articles. Measuring irradiance in practice requires knowing the power output of the bulbs within the desired wavelength range and the bulb's radiance characteristics. Irradiance detectors are standard tools, but achieving high precision (less than 10%) necessitates careful consideration of spectral characteristics and geometry. For optimal accuracy, using an integrating sphere with a stable detector can yield measurements with 1% precision.
tihort
Messages
6
Reaction score
0
hello.

I am working to understand and design a UV sanitization device.

What is a good resource to learn about irradiance and its relation to distance from a surface.

Secondly, in real life, how is irradiance measured? Given a surface area, I would need to specify bulbs of certain intensity to get the device to work.

Thank you.
 
Physics news on Phys.org
http://en.wikipedia.org/wiki/Irradiance has the basics. For a simple source (isotropic in all directions) it is an inverse square law (1/d^2, where d is distance) http://en.wikipedia.org/wiki/Inverse_square_law

You can get detectors that measure irradiance. Do you need more information than that?

For an actual calculation you need to know the power output of the bulbs you are getting within the wavelength band that you need (bulbs usually emit over a wide range of frequencies).
 
Last edited by a moderator:
The wiki article was ok, my gold-standard reference is Wolfe's "Introduction to Radiometry".

As to your second question, it can get a little tricky- you need to know details about the bulb's radiance: how much light goes in what direction. That plus the geometry will tell you how much radiation is incident on your surface.

Irradiance detectors are a fairly standard piece of equipment, but to get highly accurate and precise (<10%) measurements requires some care- spectral characteristics, polariation control, etc. An integrating sphere combined with a stable detector can give 1% precision without too much trouble.

NIST has a good bit of information about irradiance measurements:

http://physics.nist.gov/Divisions/Div844/facilities/fascal/fascal.html

http://ts.nist.gov/MeasurementServices/Calibrations/spectroradiometric.cfm
 
Last edited by a moderator:
I think it's easist first to watch a short vidio clip I find these videos very relaxing to watch .. I got to thinking is this being done in the most efficient way? The sand has to be suspended in the water to move it to the outlet ... The faster the water , the more turbulance and the sand stays suspended, so it seems to me the rule of thumb is the hose be aimed towards the outlet at all times .. Many times the workers hit the sand directly which will greatly reduce the water...