How can I prevent increasing capacitance on my homemade sensor?

AI Thread Summary
A homemade capacitive water level sensor is experiencing a gradual increase in measured capacitance from 53 pF to 54 pF while connected to a multimeter. The sensor is currently in air, and various factors such as nearby conductors, temperature, and humidity may influence the readings. Suggestions include experimenting with different environmental conditions, turning off the multimeter between measurements, and calibrating the sensor by measuring capacitance at varying heights above water. The accuracy of the multimeter should also be considered, as fluctuations can occur due to battery voltage, signal proximity to transition values, and thermal drift. Engaging in hands-on experimentation is encouraged to identify the root cause of the issue.
akaliuseheal
Messages
53
Reaction score
8
TL;DR Summary
Capacitance of sensor is not stable, why?
Hello,

I have made a capacitive water level sensor. It is a parallel plate capacitor.
While measuring the capacitance of my sensor, I measure 53 pF. I then leave it for couple of minutes, still conected to the meter, capacitance then rises to 54 pF. Capacitance is slowly increasing. Why is that and how do i prevent it?
 
Engineering news on Phys.org
Is this in water?
 
Exactly how are you measuring the capacitance?
 
hutchphd said:
Is this in water?

It is in air.

hutchphd said:
Exactly how are you measuring the capacitance?

I have the capacitor connected to the multimeter set to measure capacitance. I have it conected and turned on for couple of minutes.
 
Can you make it move by breathing on it? Also this sensor geometry (this is the parallel trace device?) will be affected by the presence of nearby conductors...see if you can make it change by threatening it with a wire or your hand.
 
The secret of dealing with this sort of problem is to play about with absolutely everything and see what has an effect. Try to make the effect worse by changing temperature and humidity (boil a kettle nearby for instance). Try turning off the DMM between measurements. Calibrate the depth - measure C vs height of plate above water. Min and max values as water level is varied. Are you sure that Capacitance is actually that of the plate to water? etc. etc. Measure the C of a pair of plates (not water) does the value agree (near enough) with what the sums tell you?

Most of those things may not give you a clue ( you may have done some of them) but one of them may give the crucial clue. There is a great temptation, when you make something and it doesn't behave as you expect, to put it in a cupboard and not chase the problem. Zero learning experience. A half day fiddling with it as above and you will learn so much and it may even end up working!
 
  • Like
Likes Tom.G and jrmichler
Read the accuracy specs of your meter, they will likely be of the form xx% ±1 digit. A 1 digit jitter in any digital readout is quite common and expected for several reasons; i.e. battery voltage dropping, the actual input signal being very close to the transition value between steps (with random electrical noise occassionally forcing a transition), internal thermal drift in the meter... well, you get the idea.

Also that's only a 2%-of-reading, not bad at all in the real world.

EDIT: that accuracy of xx% may be either % of Reading or % of Full Scale

Cheers,
Tom
 