How Can You Effectively Measure Reactive Power in Your Experiment?

Click For Summary

Discussion Overview

The discussion centers around measuring reactive power in an experimental setup using a meter. Participants explore various methods and components, including transformers, resistive loads, motors, and capacitors, to effectively generate and measure reactive power.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • David seeks to measure reactive power using a meter and considers using a transformer with a resistive load, questioning if this setup would register reactive power.
  • Some participants suggest that a 2-channel oscilloscope would be beneficial for observing reactive effects, although David does not have access to one.
  • David expresses the need for a load that consumes reactive power, indicating uncertainty about how to utilize the meter for this purpose.
  • There is a discussion about the nature of transformers and their ability to absorb reactive power, with David referencing external material on the topic.
  • One participant mentions that a VAR meter is necessary for measuring reactive power, while another suggests using a combination of voltmeters and ammeters to calculate reactive power indirectly.
  • Participants discuss the concept of using a "pure" reactive load, such as a motor run capacitor, to demonstrate the principles of reactive power measurement.
  • David clarifies that he is using a digital electrical meter capable of separating real and reactive power, but the reactive power reading is zero due to the lack of a suitable load.
  • There is a query about the implications of using a capacitor rated for higher voltage than the supply, with reassurance provided that it would function correctly at reduced voltage.

Areas of Agreement / Disagreement

Participants express various methods and components for measuring reactive power, but there is no consensus on the best approach or the effectiveness of the proposed setups. The discussion remains unresolved regarding the optimal configuration for David's experiment.

Contextual Notes

Participants note the importance of using a "pure" reactive load for clarity in measurements, and there are references to potential issues with power dissipation in resistors. The discussion includes various assumptions about the capabilities of the measuring equipment and the nature of reactive loads.

Davanger
Messages
4
Reaction score
0
Hi everyone! I always use this forum to consult but I had never had the need to post something.

I need to create an experiment in which I measure reactive power with a meter (see attachment).

I was thinking on using a transformer with a resistive load on it's secondary, but I'm not sure if the current generated by the resistor on the secondary would be perceived as reactive by the meter.

Would this work?

Besides connecting a motor to the line, and using a transformer, is there any other way I could generate reactive power for it to be perceived by the meter?

Thanks for your help.
David.

PD: My native language is Spanish, sorry if I made any mistakes while writing this.
 

Attachments

  • transf.PNG
    transf.PNG
    4.8 KB · Views: 535
Physics news on Phys.org
Davanger said:
Hi everyone! I always use this forum to consult but I had never had the need to post something.

I need to create an experiment in which I measure reactive power with a meter (see attachment).

I was thinking on using a transformer with a resistive load on it's secondary, but I'm not sure if the current generated by the resistor on the secondary would be perceived as reactive by the meter.

Would this work?

Besides connecting a motor to the line, and using a transformer, is there any other way I could generate reactive power for it to be perceived by the meter?

Thanks for your help.
David.

PD: My native language is Spanish, sorry if I made any mistakes while writing this.

Welcome to the PF.

To effectively observe reactive effects, you will need a 2-channel oscilloscope. Do you have access to such an instrument?
 
Thanks!

I don't have an oscilloscope at hand, however, the purpose of the experiment is to test the meter's capability to measure the reactive power consumed by a load connected to it, I just need to connect something that consumes reactive power as a load.
 
Davanger said:
Thanks!

I don't have an oscilloscope at hand, however, the purpose of the experiment is to test the meter's capability to measure the reactive power consumed by a load connected to it, I just need to connect something that consumes reactive power as a load.

A load that contains an inductor or capacitor will have a reactance, but I'm not sure how to use the meter to measure the reactive power.

BTW, in the circuit you posted, the first 100 Ohm resistor will be dissipating close to 500 Watts. I'm not sure that's such a good idea...
 
Yes, I actually have really big resistors that can dissipate that sort of power as heat, that is really not an issue, it's not very efficient but this is for test purposes only, not something permanent.

I read this:
Transformers - Transformers produce magnetic fields and therefore absorb reactive power.
The heavier the current loading the higher the absorption


http://www.nationalgrid.com/NR/rdonlyres/43892106-1CC7-4BEF-A434-7359F155092B/3543/Reactive_Introduction_oct01.pdf (Page 2). Is that correct? and if so, is the reactive power absorbed by it comparable to the reactive power absorbed by an electric motor?
 
If you want to use a single meter to measure reactive power, what you need is a "VAR meter"; see:

http://www.globalinstru.com/watt-meters.html

http://2.imimg.com/data2/UN/NG/MY-3232726/electronic-analog-type-w-var-pf-and-hz-meters.pdf

VAR meters are not commonly found outside of the electric power industry.

An alternative method of measuring reactive power would be to measure VA (volt-amperes) with two meters--an ammeter and a voltmeter, and also measure the real power with a wattmeter. From those two measurements you can calculate reactive power.

See these references:

http://en.wikipedia.org/wiki/Volt-amperes_reactive

http://en.wikipedia.org/wiki/Power_factor

http://en.wikipedia.org/wiki/Wattmeter
 
berkeman said:
To effectively observe reactive effects, you will need a 2-channel oscilloscope.
A little off-topic...but interesting.

A vintage way:

Wesleyan6a.JPG


An electrodynamometer is an instrument used for measuring the electric power. The basic principle was laid out in an 1848 paper by Wilhelm Weber (1804-1891): when the same current passes through two concentric coils placed at right angles to each other, the resulting torque depends on the square of the current.

http://physics.kenyon.edu/EarlyApparatus/Electrical_Measurements/Electrodynamometer/Electrodynamometer.html



A modern way:

2hptK.png


http://www.cirrus.com/en/products/cs5463.html
 
Davanger said:
......

I read this:
Transformers - Transformers produce magnetic fields and therefore absorb reactive power.
The heavier the current loading the higher the absorption


http://www.nationalgrid.com/NR/rdonlyres/43892106-1CC7-4BEF-A434-7359F155092B/3543/Reactive_Introduction_oct01.pdf (Page 2). Is that correct? and if so, is the reactive power absorbed by it comparable to the reactive power absorbed by an electric motor?

Your proposed circuit consumes both real and reactive power. So does a motor or transformer.
You haven't said what your "meter" consists of... is it able to separate real from reactive power?

In an experiment one usually tries to make it clearly demonstrate the principle being investigated. To that end you'd like to have a fairly 'pure' reactive load to start with.

A large transformer operated at say half voltage should be a fairly 'pure' inductive reactance, because the iron losses will be small at half voltage and the copper losses will be nil with only magnetizing current flowing. Your trouble will be finding one cheap..

A motor run capacitor should make a fairly pure capacitive reactance. One of those should cost around twelve bucks at an electrical supply house. Often you can pillage one from an abandoned appliance in a scrap pile - airconditioners and furnaces have one for the fan motor. Beware motor start capacitors, they are not suitable for continuous energization.
Motor run caps usually have a metal case and are labelled "motor run". Most start caps are plastic case.


Once you have a fairly pure reactive load, you can parallel it with resistive and observe the pythagorean relationship of VA to watts and vars.
That'd be a neat experiment.

good luck

old jim
 
I'm using a digital electrical meter, the kind of meter you would have at you house measuring your monthly consumption for the electrical company. It is able so separate real power from reactive power and shows them in different fields, the reactive one is just marking zero because I haven't been able to connect anything that absorbs reactive power as a load.

A motor run capacitor sounds like a great idea! I found these in my area:

http://articulo.mercadolibre.com.ve/MLV-406697771-capacitor-de-marcha-55-uf-440-v-_JM

http://articulo.mercadolibre.com.ve/MLV-406785547-capacitor-de-marcha-aires-acondicionado-_JM

(Is MFD another way of writing uF? for a moment I thought that was Mega and thought it was odd)

However, is there any consequence to undervolting one of these capacitors? The only ones I could find are for 440 VAC and I have a 220 VAC connection.

You have been of great help, thank you!
David.
 
  • #10
(Is MFD another way of writing uF? for a moment I thought that was Mega and thought it was odd)

yes, it's often used. I guess some labelmakers just don't have Greek letters.

And your capacitor will be happy at reduced voltage.
Feel it after a few minutes, it should not be heating up.
Capacitors do have small resistance because their plates are made from thin foil.
And the dielectric does absorb a little energy which comes out as heat.
When you study capacitors, that's what the tan δ parameter is - a measure of the deviation from pure capacitance.
A motor run cap should be pretty good, especially at reduced voltage where you aren't working the dielectric very hard.

Have you yet calculated expected current with that 40 uf? I don't know your meter, just plan ahead so you don't let the smoke out of your test equipment...

Here's some info on motor run caps
http://www.cde.com/catalogs/ACappGUIDE.pdf

i've found CDE's site to be an excellent source of technical information on capacitors. They kinda "wrote the book"...
http://www.cde.com/Have fun and learn a lot !old jim
 
  • #11
The other option - and good experiment, will be to connect a 220V AC induction motor and have a way to vary the mechanical load.

These have poor power factor when unloaded - when looking at Reactive power it will stay about the same as you increase the mechancal load, but the real power will increase. Then add a capacitor to the circuit and see how you can "cancel" out the reactive power ( from the sources perspective).

Your original circuit will have relatively low reactive power. Unless you replace the resistor with a suitable sized inductor.
 

Similar threads

  • · Replies 4 ·
Replies
4
Views
2K
  • · Replies 8 ·
Replies
8
Views
3K
Replies
2
Views
3K
Replies
4
Views
3K
  • · Replies 8 ·
Replies
8
Views
4K
Replies
9
Views
4K
  • · Replies 1 ·
Replies
1
Views
845
  • · Replies 20 ·
Replies
20
Views
5K
  • · Replies 38 ·
2
Replies
38
Views
6K
  • · Replies 1 ·
Replies
1
Views
2K