How do you calculate temp change when compressing water?

  • Thread starter Thread starter Bombicis
  • Start date Start date
  • Tags Tags
    Change Water
Click For Summary
The discussion focuses on the challenges of measuring pressure changes in a water-filled pressure chamber at 1000 psi, particularly how ambient temperature fluctuations can significantly affect pressure readings. The user has observed that small temperature changes can lead to notable pressure variations, complicating leak detection. They aim to differentiate between actual leaks and pressure changes caused by temperature fluctuations, which can mimic leak behavior. To improve accuracy, they plan to monitor both ambient and vessel temperatures and consider using more precise sensors. The conversation highlights the need for a better understanding of the thermodynamic relationships involved in water compression and temperature changes.
Bombicis
Messages
3
Reaction score
0
I built a pressure chamber to test small parts for leaks at 1000 psi. I can measure the pressure with an accuracy of ± 1 psi. I found that small changes in ambient temperature ± 2 °F can result in a slow but profound change in the pressure in the system, say 25 psi.

Insulating the chamber helped a little, but I really need to control for the change in temperature associated with changing pressure.

As a side note, with the ambient temperature constant, running the system from 0 to 1000 psi over 2 minutes and then monitoring the pressure always results in a slow and relatively consistent pressure drop that slowly tapers off. If the pressure is then dropped over 2 minutes to 200 psi and then monitored, the pressure slowly rises consistently over time again slowly tapering off. These results lead me to conclude that pressure induced temperature change is the cause.
 
Engineering news on Phys.org
Hello Bombi, :welcome:

According to the ideal gas law you have ##pV = nRT## (in decent units). So your 2 F can explain the 25 psi.
Assuming you have air in the chamber (or is it water :smile:, you don't say).
Not clear to me why you would need to control things: leaks are faults, isn't it ?
Or do you neeed leak reates at precisely 1000 psi ?

[edit] ah, wait, there's the thread title. Could you describe the setup a little better ?
 
Hey BvU,

Thanks for the reply!

The pressure chamber is filled with water, no air. I expect that any small bubbles left in cracks would dissolve into the water relatively in the early in the pressurization process.

The pressure vessel is 1/2" thick walled stainless steel and pressure is regulated via a piston. The total volume if the pressure vessel is about 3 liters and there is a reservoir on the side for filling and emptying the vessel.

Leaks are indeed faults, but they can occur very slowly so I need to detect the difference between a leak and temperature induced pressure fluctuations. Because this is a manufacturing environment, the quicker I can determine if a part leaks, the better. I have found that in certain cases, leaks can look similar to thermally induced pressure changes and then it can take an hour or more to tell the difference between the two.

My hope is that by monitoring both ambient and vessel temperature, I can more quickly determine the difference between a leak and thermally induced pressure change.
 
I looked at the compressibility of water (2.2.1 there) and found about 0.36% difference in density going from 1 atm to 70 atm at 20 C. The work done would increase the temperature, but we need an steam tables expert to estimate the temperature rise from that. @Chestermiller or @SteamKing ?
 
The relationship between volume, pressure, and temperature for water is$$\Delta \ln V=\alpha \Delta T-\beta\Delta P$$
where ##\alpha## is the coefficient of volumetric expansion (0.000214/C) and beta is the bulk compressibility (##4.6\times 10^{-10}/Pa##). You probably also need to include the changes in volume of the container as a function of temperature and pressure.
 
Thanks so so much both BvU and Chestermiller.

I have been wrestling with this for a while and now I think I can get my arms around the problem. I think I will need a more accurate pressure sensor as well as a very accurate thermal probe, but now I am ready to take a stab at predicting outcomes rather than finding an explanation for them.
 
Had my central air system checked when it sortta wasn't working. I guess I hadn't replaced the filter. Guy suggested I might want to get a UV filter accessory. He said it would "kill bugs and particulates". I know UV can kill the former, not sure how he thinks it's gonna murder the latter. Now I'm finding out there's more than one type of UV filter: one for the air flow and one for the coil. He was suggesting we might get one for the air flow, but now we'll have to change the bulb...

Similar threads

  • · Replies 3 ·
Replies
3
Views
3K
  • · Replies 45 ·
2
Replies
45
Views
6K
  • · Replies 8 ·
Replies
8
Views
3K
Replies
1
Views
2K
Replies
7
Views
4K
  • · Replies 20 ·
Replies
20
Views
10K
  • · Replies 7 ·
Replies
7
Views
2K
  • · Replies 8 ·
Replies
8
Views
5K
Replies
4
Views
3K
  • · Replies 47 ·
2
Replies
47
Views
9K