IR transmission through a very thin liquid-filled cell

Click For Summary

Discussion Overview

The discussion revolves around the feasibility of measuring the transmittance of infrared (IR) radiation through a very thin liquid-filled cell, specifically with a thickness of about 10 microns. Participants explore the implications of sample thickness on absorption measurements and the potential challenges posed by interference effects and solvent interactions.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested
  • Mathematical reasoning

Main Points Raised

  • One participant inquires about the possibility of using a Nicolet is50 FTIR spectrometer to measure transmittance through a thin liquid sample, noting that thicker samples may lead to absorption coefficients that cannot be calculated using the Lambert-Beer law.
  • Another participant suggests diluting the sample with an inert solvent, although this may disturb hydrogen bonding in certain samples, such as alcohols.
  • It is proposed that FTIR or diffraction grating spectrometers could measure the transmission spectrum, but interference fringes may affect the results due to the sample thickness being comparable to infrared wavelengths.
  • A participant references historical measurements from the 1960s and discusses the need for precise thickness measurements to apply the Beer-Lambert law effectively.
  • Questions arise regarding whether specific liquids or any liquids can be used, with a focus on flammable hydrocarbons and alcohols relevant to radiative transfer in burning solvent pools.
  • Some participants note that differences between pure liquids and solutions may be negligible for modeling purposes, although the non-linearity of Beer-Lambert's law for concentrated solutions is acknowledged.
  • Concerns are raised about the complexity of calculating absorption for ethanol or alcohols diluted with water due to the importance of hydrogen bonding.
  • One participant mentions that finding spectra for pure ethanol is feasible and questions the necessity of using thin layers for recording spectra.
  • Another participant provides references to articles discussing absorbance in diluted alcohols, indicating that significant absorption can occur even at larger path lengths.

Areas of Agreement / Disagreement

Participants express a variety of viewpoints regarding the feasibility and methods for measuring IR transmittance through thin liquid samples. There is no consensus on the best approach, and multiple competing views remain regarding the implications of sample thickness, solvent interactions, and the applicability of Beer-Lambert law.

Contextual Notes

Participants highlight limitations related to the assumptions of Beer-Lambert law, the potential impact of interference fringes, and the need for precise thickness measurements. The discussion also acknowledges the complexity of interactions in specific liquid mixtures.

hilbert2
Science Advisor
Insights Author
Messages
1,611
Reaction score
611
Hi,

does anyone know if it's possible, with a Nicolet is50 FTIR spectrometer or similar, to measure the transmittance of IR radiation through a cell filled with a liquid sample with thickness in the range of about 10 microns and accurately adjustable? The reason for doing this would be because some substances may absorb so effectively at certain wavelengths that the transmittance would be practically zero for a larger sample thickness and one couldn't calculate the absorption coefficient for the whole wavelength range from the Lambert-Beer law.

The sample would probably have to be contained in a space between some kind of IR-transparent plates, with plate distance accurately adjustable.

Thanks,
Hilbert2
 
Chemistry news on Phys.org
Can't you dilute the sample with an inert solvent?
 
  • Like
Likes   Reactions: hilbert2
Good idea, I'll find out if that's possible. However, even an inert solvent will disturb hydrogen bonding in e.g. alcohol sample and affect the spectrum...
 
An FTIR spectrometer or even a diffraction grating spectrometer should be able to measure the transmission spectrum.
 
Because 10 µm is of the order of infrared wavelengths (10 µm = 1000 cm-1), your spectrum may be affected by interference fringes.
Here is a spectrum of a polyethylene film ca. 30 µ thick. At 10 µm the fringe spacing would be 3 x as great. The intensity would also depend on the refractive index mismatch at the interface.
 

Attachments

  • Like
Likes   Reactions: hilbert2 and Charles Link
Thanks, I saw those interference fringes being mentioned in some publication. The first information about this kind of measurements seems to have been produced in the 1960s: http://www.nrcresearchpress.com/doi/pdf/10.1139/v69-572 .

One approach may be to produce a liquid film with a thickness that's somewhere in the acceptable micron range, and then just measure that thickness precisely enough so that Beer-Lambert law can be used to find the absorption coefficient from the IR transmittance data.

The reason for doing these measurements is because I have to form some kind of law of how deeply blackbody-like radiation can go under a liquid surface. The radiation is produced in the gas phase above the liquid and I'm interested in something like how many millimeters of liquid are enough to absorb 90% of the total radiative energy flux - this of course depends on the blackbody temperature and the absorption spectrum of the liquid.
 
Specific liquid or any liquid?
 
If you are not interested in modeling of a particular substance, differences between measurements of a pure liquid and a solution can be negligible - that is, they can give values good enough to construct a model.

Then, Beer-Lambert's law is not linear for concentrated solutions, but I think you should be able to find reasonably good models that could help predict real values (at least I remember being told something like that). Or - if not - perhaps you can initially choose a rather "transparent" liquid for getting some reasonable data (transparent as in "with absorption low enough to allow measurements of more thick layers").

Just thinking aloud.
 
  • #10
Yeah, it's probably possible to find some relatively simple approximation that works well enough for complex hydrocarbon mixtures (like gasoline or motor oil) in most cases. A more problematic case would be a calculation for ethanol or some other alcohol diluted with water, as there the hydrogen bonding between the species will be important.
 
  • #11
hilbert2 said:
A more problematic case would be a calculation for ethanol or some other alcohol diluted with water, as there the hydrogen bonding between the species will be important.

Probably, but finding spectra of pure ethanol is not that difficult. I can be missing something, won't be for the first time, but I am not convinced you really need to use thin layers for recording them.

http://sdbs.db.aist.go.jp/sdbs/cgi-bin/direct_frame_disp.cgi?sdbsno=1300

(choose IR on the left)
 