fsonnichsen said:
I have taken a time dependent fluorescence series of spectra for an ethylene glycol solution using a 532nm laser as the excitation source. I have attached the spectra to this post. Nothing looks unusual-there is a sharp notch near the 532nm line--this is due to a filter I have used to protect the spectrometer from the laser.
However--note that the spectrum has points to the left of the laser line-e.g. at wavelengths lower (energies higher) than the laser. How does this happen? I don't believe this could be attributed to Raman /anti-stokes processes due to the large separation--usually these shifts are smaller yes? Even though we are dealing with a complex solution here, we cannot expect electrons to become promoted to levels higher than the excitation photons to this extent.
thanks
Fritz
You have to eliminate the possibility that the antiStokes line is an artifact caused by spontaneous fluorescence of the lasing material. The laser that you are using to excite the sample may not be monochromatic.
I recommend that you use a "sanity check". Replace your sample with either an empty cuvette or a scatter plate. A frosty piece of glass (i.e., scatter plate) should have no fluorescence lines. If you still still see the same anitStokes fluorescence line, then you have an artifact.
You may not have a frosty piece of glass. However, you could use almost any white thing as a sanity check. You could even use a white piece of paper. Warning, white paper does have a broad fluorescence spectrum. However, white paper probably doesn't have the same antiStokes fluorescence line.
One should always use at least one blank for comparison. I recommend a few blanks with different scattering properties. Analytical chemists are trained to always use a blank.
One thing you could do is place a narrow band filter at 532 nm right in front of the laser. You can use a multilayer filter, a diffraction grating or even a prism in front of the laser. If you use a diffraction grating or a prism, you also need an aperture to eliminate the stray reflections at oblique angles. If the antiStokes line disappears, then it was probably an artifact.
Spontaneous emission of the lasing material is a chronic problem with gas lasers. Gas lasers often have strong emission bands with wavelengths shorter than the main band. I made that mistake often when I was doing my thesis. The problem is not as severe with solid state lasers. However, even solid state lasers have some spontaneous emission at wavelengths shorter than the lasing wavelength.
I would even suggest that you always use a dispersive device in front of your laser to eliminate or reduce emission at wavelengths other than the desired laser line. Lasers are more monochromatic than other other sources of light, but they still have extraneous emission at different wavelengths. Fluorescence and Raman signals are very weak compared to the laser line. Therefore, even a small emission from the laser at the wrong wavelength can be mistaken for a fluorescence of Raman signal.
Never believe those people who tell you that the laser is monochromatic. Not even the manufacturer. Not even your thesis adviser. Trust no one! Since your antiStokes shift is larger than the typical antiStokes Raman line, it should be easy to find a dispersive device that can eliminate the spontaneous emission from the laser medium.
It may be useful for you to use a scatter plate to examine the spectrum of your laser. A l laser can have more than one lasing line in addition to spontaneous emission. If your "antiStokes" line matches one of the lines of your laser, then maybe you are looking at a laser artifact. If your anti-Stokes signal is actually from the sample (i.e., "real"), then the real fun begins. If you have done a number of tests to eliminate the possibility of an error, then you should do a number of other tests to determine what that line really is.
You still won't know if your antiStokes band is from scattering or fluorescence. Use different laser wavelengths to see how that line changes. Lines due to scattering have a constant shift, lines due to fluorescence have a constant wavelength.Use neutral density filters in front of the laser to determine how the intensity of the band changes with laser intensity. Check polarization of the antiStokes band.
Once you have reduced the possibility of artifacts, you should examine the weird possibilities.