Mechanical pump oil discharge through inlet

Click For Summary
SUMMARY

The discussion centers on an issue with a Leybold Trivac D16 mechanical vacuum pump discharging excessive oil through its inlet, potentially due to overfilling, worn components, or valve malfunctions. Users suggest checking the oil level, inspecting for leaking valves or broken piston rings, and ensuring the inlet line is sloped to facilitate oil return. Additionally, the importance of using a pre-vacuum filter to prevent oil vapor backflow is emphasized, along with the need to verify the condition of the diffusion pump.

PREREQUISITES
  • Understanding of Leybold Trivac D16 mechanical vacuum pump operation
  • Knowledge of vacuum pump maintenance and troubleshooting
  • Familiarity with oil-sealed vacuum pump principles
  • Experience with pre-vacuum filters and their applications
NEXT STEPS
  • Research Leybold Trivac D16 maintenance procedures
  • Learn about vacuum pump oil properties and replacement guidelines
  • Investigate the function and installation of pre-vacuum filters
  • Explore troubleshooting techniques for mechanical vacuum pumps
USEFUL FOR

Vacuum pump technicians, mechanical engineers, and maintenance personnel working with Leybold Trivac pumps or similar oil-sealed vacuum systems.

zaillian
Messages
10
Reaction score
1
I have an old mechanical vacuum pump that is discharging a lot of oil (half a gallon was found after some use) through its inlet, with enough force to move at least a few inches vertically. The pump is used in conjunction with a diffusion pump and the discharge ruined its oil. The pump also creates lot of oil mist, but this is not a problem.

I would like to hear ideas what causes the discharge and how to prevent it.

Possibly the issue is that the pump has been shut off, which after the air has been let into the chamber with the inlet open sucking oil from the pump. Then the issue is solved by making sure the valves are always shut off when filling the chamber with air. Anyone else having had this problem?
 
Engineering news on Phys.org
What make/model is this pump? Can you link to the datasheet for it? Thanks.
 
Check the oil is not too high. Someone may have overfilled the crankcase, so it is whipping oil up into a foam with the big end, then throwing it on the piston and rings from below.

Check for leaking valves or broken piston rings.

Maybe slope the inlet line so oil on the inlet side, runs with the flow, back into the mechanical pump.
 
zaillian said:
I would like to hear ideas what causes the discharge and how to prevent it.
Copied from
https://www.leyboldproducts.us/medi...tals_of_Vacuum_Technology_EN58774555441f3.pdf

"The TRIVAC rotary vane pumps are produced as single-stage (TRIVAC S) and two-stage (TRIVAC D) pumps."

Could it be that one or both stages of the pump are worn out, by-passing gas and oil through worn vanes and cylinders?
Perhaps old oil that has lost original properties?
 
  • Like
Likes   Reactions: zaillian
Hello! Perhaps a vacuum pump should be used together with a pre-vacuum filter? It is a device designed to capture or absorb oil vapor molecules coming from the pump inlet. In addition, it seems to me it makes sense to check whether the diffusion pump is the source of the oil.
1738669512010.png


An example of a pre-vacuum filter.
Pre-vacuum traps are used to protect against the reverse flow of oil from the pump to the working volumes.
They are used in combination with a mechanical pre-vacuum pump that works with a diffusion pump.
 
Last edited:
  • Like
Likes   Reactions: Lord Jestocost and Lnewqban
Lnewqban said:
Copied from
https://www.leyboldproducts.us/medi...tals_of_Vacuum_Technology_EN58774555441f3.pdf

"The TRIVAC rotary vane pumps are produced as single-stage (TRIVAC S) and two-stage (TRIVAC D) pumps."

Could it be that one or both stages of the pump are worn out, by-passing gas and oil through worn vanes and cylinders?
Perhaps old oil that has lost original properties?

Interesting and possible. Though the pump was able to achieve an ultimate pressure of approximately 10^-1 TOR in the chamber. A tech stated it is impossible to determine which way the wanes are supposed to go after taking it apart and assembling previously, so a possible fault.
 
  • Like
Likes   Reactions: Lnewqban
1738701064322.png

I suggest you check the valve indicated by the green arrow. If it does not work properly, the working chamber of the pump may overflow with oil, which will be intensively discharged through the inlet pipe of the pump. I also suggest you check the oil separator indicated by the purple arrow. If it does not pass air well, excessive pressure may be created inside the oil chamber, which will eject oil through the inlet pipe.
 
  • Like
Likes   Reactions: zaillian and Lnewqban

Similar threads

Replies
1
Views
873
Replies
18
Views
3K
  • · Replies 13 ·
Replies
13
Views
4K
  • · Replies 1 ·
Replies
1
Views
3K
  • · Replies 19 ·
Replies
19
Views
6K
  • · Replies 3 ·
Replies
3
Views
3K
  • · Replies 9 ·
Replies
9
Views
5K
Replies
5
Views
8K
  • · Replies 35 ·
2
Replies
35
Views
3K
  • · Replies 11 ·
Replies
11
Views
3K