Optimizing Averaging Algorithms for Reliable Pressure Measurements

Click For Summary
SUMMARY

The discussion focuses on optimizing averaging algorithms for reliable pressure measurements using the dsa3307 Differential Pressure Scanner. Key insights include the importance of ensuring that the pressure measurement reaches a quasi-steady state before averaging, which may require consideration of the tubing geometry and flow dynamics. The AVG command allows users to set averaging intervals, such as AVG 600 or AVG 3000, but the effectiveness of the averaging process depends on capturing sufficient data points to achieve statistical confidence. Participants emphasize the need for experimentation to determine the appropriate time for achieving steady pressure values.

PREREQUISITES
  • Understanding of dsa3307 Differential Pressure Scanner functionality
  • Knowledge of averaging algorithms and statistical confidence
  • Familiarity with pressure measurement techniques and quasi-steady state concepts
  • Basic principles of fluid dynamics and tubing effects on pressure readings
NEXT STEPS
  • Research the implementation of averaging algorithms in pressure measurement systems
  • Explore the effects of tubing geometry on pressure signal stabilization
  • Learn about statistical methods for determining sample size and confidence intervals
  • Investigate experimental setups for measuring quasi-steady pressure states
USEFUL FOR

Engineers, researchers, and technicians involved in pressure measurement and control, particularly those working with the dsa3307 Differential Pressure Scanner and related averaging algorithms.

can12345
Messages
21
Reaction score
0
< Mentor Note -- thread moved to HH from the technical engineering forums, so no HH Template is shown >[/color]

Hello everyone.
I have such a task to fullfill. I am using dsa3307 Differential pressure scanner. Can someone give me an idea? How to achieve a result for this topic.."Implementation and testing of averaging algorithms to assure sufficient averaging time"

Sincerely yours.
 
Last edited by a moderator:
Physics news on Phys.org
I'm sure you have all the proper manuals which seem to be found below?

http://www.scanivalve.com/products/pressure-measurement/ethernet-intelligent-pressure-scanners/dsa3307-pressure-scanner-liquid-measurement/

After reading the manuals the people who make the dsa3307 might reply to your questions?

http://www.scanivalve.com/contact-us
 
Last edited by a moderator:
They say you can change the variable of avarage by the using AVG command example AVG 600 or AVG 3000, however my question is to understand how generally averaging is being done? I must think about time? How long time is enough so on...
 
Well there are several issues. First, you would need to make sure that the pressure being measured has reached a quasi-steady state. Specifically, if you have a length of tubing attached to the transducer, it will take a finite time after placing it in a flow for the correct pressure to propagate to the sensing element.

Second, once you have that, the sample being averaged itself only has to be long enough that you capture enough points that you are satisfied with the statistical confidence in the measured average. Deciding that is up to you. If you are looking for spectral information, then there are additional concerns about the number of samples and frequency resolution.
 
Thank you so much. Do you know any sources or implementation for it? Your answer is exactly what i was looking for also.
 
Sources or implementation for what?
 
of averaging of a specific task to get some impression. Or some ways to determine time for quasi steady pressure time with some experiment so on.
 
I mean, you can figure out if it has reached its steady value just by looking at the pressure signal and seeing when it stops increasing or decreasing and seems to just hover around a single value. The exact amount of time will depend on the geometry of your tubing. Averaging is just averaging unless you have some reason to use a different sort of average (and it doesn't appear that you are). You just take the mean over whatever sample of values you have after it has reached its "steady" value.
 

Similar threads

Replies
3
Views
1K
  • · Replies 12 ·
Replies
12
Views
5K
  • · Replies 4 ·
Replies
4
Views
1K
  • · Replies 6 ·
Replies
6
Views
10K
Replies
3
Views
2K
  • · Replies 3 ·
Replies
3
Views
2K
Replies
14
Views
3K
  • · Replies 20 ·
Replies
20
Views
3K
  • · Replies 8 ·
Replies
8
Views
4K
  • · Replies 7 ·
Replies
7
Views
2K