Subtract Radiated from Background Spectrum to Analyze Absorption Spectrum

Click For Summary
SUMMARY

The discussion centers on the correct method for analyzing an absorption spectrum by subtracting a background spectrum using Excel. Participants emphasize that the background should be subtracted from the irradiated signal, not the other way around, to avoid issues with noise-to-signal ratios. The importance of using absorbance mode instead of transmittance mode is highlighted, along with the need for consistent x-axis units when plotting spectra. Users also note that statistical fluctuations can affect the clarity of the resulting spectrum.

PREREQUISITES
  • Understanding of absorption spectroscopy principles
  • Familiarity with Excel for data plotting
  • Knowledge of absorbance vs. transmittance modes
  • Basic concepts of noise-to-signal ratio in spectral analysis
NEXT STEPS
  • Learn how to perform background subtraction in absorbance mode
  • Research proper data plotting techniques in Excel for spectral data
  • Explore methods to improve signal-to-noise ratios in spectral measurements
  • Investigate the effects of cuvette size on spectral analysis
USEFUL FOR

Researchers and analysts in the field of spectroscopy, chemists conducting absorption experiments, and anyone involved in data analysis of spectral data will benefit from this discussion.

Jhon81
Messages
11
Reaction score
0
Hi everyone,
I have an absorption spectrum that I obtained from a sample after radiation and I need to use the background sample as a reference, how can I do that using excel?

what I did is subtract the radiated from the background spectrum then plot it, is that right?

Thanks.
 
Physics news on Phys.org
I would subtract the background from the signal not the opposite way, but anyway, that is just a sign.

If you don't run into a problem with statistical fluctuations that approach should be fine. More context would be useful, however.
 
To do a correct subtraction you should do it in absorbance mode, not transmittance; but perhaps you already knew that.
 
Thanks, mfb and mjc123 for your comments.
I subtracted the background from the signal but after plotting it seems that the points have spread! why is that?
subtraction.png

then I tried to add them and the shape of the plot was looking more nicer, could someone explain what happened to me?
adding.png
 

Attachments

  • adding.png
    adding.png
    9.5 KB · Views: 846
  • subtraction.png
    subtraction.png
    7.4 KB · Views: 800
Why did you add them?
Can you show your sample and background spectra? It looks as if they are very similar, so the subtracted signal is much weaker than either spectrum. That's why the noise-to-signal ratio is much greater on your subtracted spectrum (I assume that's what you mean by the points "spreading").
What are your y-axis units? If they are mAU, then your spectrum looks a bit too strong for accurate quantitative analysis. Can you repeat the experiment at a lower concentration?
 
mjc123 said:
Why did you add them?
Can you show your sample and background spectra? It looks as if they are very similar, so the subtracted signal is much weaker than either spectrum. That's why the noise-to-signal ratio is much greater on your subtracted spectrum (I assume that's what you mean by the points "spreading").
What are your y-axis units? If they are mAU, then your spectrum looks a bit too strong for accurate quantitative analysis. Can you repeat the experiment at a lower concentration?
I added them to see the shape of the spectrum.
This is the spectra of the background and the first irradiated sample, respectively:
background.png
1st irradiated.png

As for the unit, I am not sure as it was my first time using that spectrometer, I am going to check and let you know.
If you mean the concentration of irradiated sample it is very law ( I am using FXO with Fe+2 (0.3mM) & XO (0.05mM)) all of them dissolved in 50 mM of H2SO4.
 

Attachments

  • background.png
    background.png
    6.9 KB · Views: 684
  • 1st irradiated.png
    1st irradiated.png
    7.2 KB · Views: 714
Last edited:
What are your x-axis units here? You have very similar spectra, but one with the maximum at ca. 650 (nm? that would make sense) and the other at ca. 1800 (what? is it simply data point number in excel? I doubt very much if you scanned down to 0 nm!)
If you ran your spectra in a 10 mm cuvette, try using a 1 mm cuvette.
But I think you'll find it difficult getting a good subtraction spectrum. The difference between the spectra is of the same order of magnitude as the noise - as you found!
 
mjc123 said:
What are your x-axis units here? You have very similar spectra, but one with the maximum at ca. 650 (nm? that would make sense) and the other at ca. 1800 (what? is it simply data point number in excel? I doubt very much if you scanned down to 0 nm!)
If you ran your spectra in a 10 mm cuvette, try using a 1 mm cuvette.
But I think you'll find it difficult getting a good subtraction spectrum. The difference between the spectra is of the same order of magnitude as the noise - as you found!

The x-axis unite is nm. I tried to compare the 4th irradiated with the background
background spectra
background.png

4th irradiated spectra
4th irradiated.png

subtraction of the above spectrums
subtraction2.png

Is it good that the spectra are in the minus y-axis or not, but I noticed that the points are not very spread which makes the noise less right?
 

Attachments

  • background.png
    background.png
    6.8 KB · Views: 666
  • 4th irradiated.png
    4th irradiated.png
    10.8 KB · Views: 697
  • subtraction2.png
    subtraction2.png
    6.3 KB · Views: 659
This is a bit better. You can see that the peak at 600 nm is reduced in the irradiated spectrum. This presumably corresponds to your negative peak at "1500" in the subtraction spectrum - get your x-axis units right. Yes, you should subtract the background from the irradiated spectrum, so your peak is negative.. Again, what do you mean by the points being "not very spread"?
 
  • #10
mjc123 said:
This is a bit better. You can see that the peak at 600 nm is reduced in the irradiated spectrum. This presumably corresponds to your negative peak at "1500" in the subtraction spectrum - get your x-axis units right. Yes, you should subtract the background from the irradiated spectrum, so your peak is negative.. Again, what do you mean by the points being "not very spread"?
what do you mean to get my x-axis unit right?
In this picture, you can see that the points are very spread ( the spectra is not smooth)
subtraction.png


but in the second one, it is smoother.

subtraction2.png
 

Attachments

  • subtraction.png
    subtraction.png
    7.4 KB · Views: 616
  • subtraction2.png
    subtraction2.png
    6.3 KB · Views: 672
  • #11
Jhon81 said:
what do you mean to get my x-axis unit right?
Your measured spectra are from 200 to 1100 nm. Why does the difference have x values of 0 to 3500?
Jhon81 said:
In this picture, you can see that the points are very spread ( the spectra is not smooth)
That tells you that there is some noise in the measurement. Not surprising.
Jhon81 said:
but in the second one, it is smoother.
The signal (the true difference) is stronger, so the signal to noise ratio is better.
 
  • #12
@Jhon81 In one case it appears you are graphing wavelength on the x-axis in nm and in the other case, you are apparently doing wavenumber=## \frac{1}{\lambda} ## (cm^-1). It is difficult to try to analyze the results unless this is consistent. You must at least state that you made the conversion. ## \\ ## Edit: But that doesn't appear to be the case either. It leaves us guessing...
 
  • #13
It can't be that because the spectra aren't reversed. It seems to be that sometimes you are plotting an x-y chart with wavelength as the x-value and absorbance as the y-value; other times you are just plotting a y chart, with the absorbance as the y value and the data point number (looks like you have about 3700 data points)appears as the x value. You want to get some more practice plotting charts in excel (or whatever you use).
 

Similar threads

  • · Replies 2 ·
Replies
2
Views
3K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 6 ·
Replies
6
Views
2K
  • · Replies 4 ·
Replies
4
Views
3K
  • · Replies 46 ·
2
Replies
46
Views
4K
  • · Replies 17 ·
Replies
17
Views
8K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 29 ·
Replies
29
Views
4K
Replies
7
Views
2K
  • · Replies 1 ·
Replies
1
Views
4K